Support Apnea Board & OSCAR  

Difference between revisions of "Oxygen Bleed with CPAP"

From Apnea Board Wiki
Jump to: navigation, search

(Created page with "Many CPAP users also require supplemental oxygen. Oxygen is provided into the CPAP respiratory circuit using an "oxygen bleed adapter". Concentrated oxygen is generally del...")
 
Line 1: Line 1:
 
Many CPAP users also require supplemental oxygen.   
 
Many CPAP users also require supplemental oxygen.   
Oxygen is provided into the CPAP respiratory circuit using an "oxygen bleed adapter".  
+
=Oxygen Bleed Adapter=
 +
The Oxygen Bleed Adapter is inserted into the CPAP/BPAP circuit to allow oxygen to be mixed into the respiratory circuit. Oxygen is always provided at higher pressure at a set flow-rate, so the dose can be calculated. Several bleed adapters are available, and they are inexpensive.  The images below show oxygen bleed adapters for both standard and heated tubing.
 +
 
 +
[[file:oxygen bleed adapter Standard.png]]
 +
 
 +
[[file:oxygen bleed adapter heated hose.png]]
 +
 
 +
=Physical Characteristics of an Oxygen Bleed=
 
Concentrated oxygen is generally delivered at pressures in excess of 15 psi, and is metered by Flow.
 
Concentrated oxygen is generally delivered at pressures in excess of 15 psi, and is metered by Flow.
 
CPAP air pressure is delivered at pressures less than 30 cm H2O.  
 
CPAP air pressure is delivered at pressures less than 30 cm H2O.  
Line 7: Line 14:
 
The typical CPAP or Bilevel respiratory circuit flows 40 to 80 L/min of air volume in the absence of large leaks. (our example will assume 60 L/min)
 
The typical CPAP or Bilevel respiratory circuit flows 40 to 80 L/min of air volume in the absence of large leaks. (our example will assume 60 L/min)
  
What is the oxygen concentration in the diluted CPAP circuit?
+
=What is the oxygen concentration in the diluted CPAP circuit?=
  
 
(oxygen flow rate (litres) at 100% + air volume (litres) at 21% O2) / Total air volume = oxygen concentration %
 
(oxygen flow rate (litres) at 100% + air volume (litres) at 21% O2) / Total air volume = oxygen concentration %

Revision as of 01:36, 11 March 2018

Many CPAP users also require supplemental oxygen.

Oxygen Bleed Adapter

The Oxygen Bleed Adapter is inserted into the CPAP/BPAP circuit to allow oxygen to be mixed into the respiratory circuit. Oxygen is always provided at higher pressure at a set flow-rate, so the dose can be calculated. Several bleed adapters are available, and they are inexpensive. The images below show oxygen bleed adapters for both standard and heated tubing.

Oxygen bleed adapter Standard.png

Oxygen bleed adapter heated hose.png

Physical Characteristics of an Oxygen Bleed

Concentrated oxygen is generally delivered at pressures in excess of 15 psi, and is metered by Flow. CPAP air pressure is delivered at pressures less than 30 cm H2O. 1-PSI is 70 cm-H2O, so the oxygen concentrator flow is fully delivered to the CPAP circuit. A typical oxygen bleed circuit delivers between 1-liter/minute (L/min) to 3-L/min. The typical CPAP or Bilevel respiratory circuit flows 40 to 80 L/min of air volume in the absence of large leaks. (our example will assume 60 L/min)

What is the oxygen concentration in the diluted CPAP circuit?

(oxygen flow rate (litres) at 100% + air volume (litres) at 21% O2) / Total air volume = oxygen concentration % therefore, at 1 litres/minute oxygen addition and assuming a total CPAP flow of 60 L/min: (1L x 100) + (60L x 21) = 1360 /60 L/min = 22.% Oxygen. at 1 litre/minute oxygen (3L x 100) + (60L x 21) = 1560 / 60 L/min = 26% oxygen.

Any variable can be changed based on available data. If the flow rate is 40 L/min from the CPAP, the oxygen concentrations will be higher. If leaks are present or pressure results in a higher average vent-rate, the oxygen concentrations will be lower.




Donate to Apnea Board