Changes


Oxygen Bleed with CPAP

64 bytes added, 14:48, 11 March 2018
What is the oxygen concentration in the diluted CPAP circuit?
The equation to calculate concentration from flow is:<br
(oxygen flow rate (litres) at 100% O2) + (air volume (litres) at 21% O2)) / Total air volume = oxygen concentration Percent in circuit <br>
therefore, :<br>at 1 litres/minute oxygen addition at 100% and assuming a total CPAP flow of 30 L/min:<br>(1-L/min x 100) + (30-L/min x 21) = 730/31-L/min = 23.55% Oxygen.<br><br> at 3 litre/minute oxygen at 100% and assuming a total CPAP flow of 30 L/min:<br>
(3-L/min x 100) + (30-L/min x 21) = 930/33 L/min = 28.18% oxygen.
370
edits



Donate to Apnea Board