[CPAP] Increased EPR reduces flow limits even when EPAP is simultaneously decreased
#1
Exclaimation 
Increased EPR reduces flow limits even when EPAP is simultaneously decreased
I decided to start a separate thread with this specific title so it is easier to find and one can refer to it directly when advising members. 

There has been controversy over the correct setting up of EPR on this forum. 

One argument emphasizes that expiratory pressure keeps the airways open; hence, once some EPR is added to an already established constant-pressure CPAP therapy, the expiratory pressure should be increased by the same amount as the EPR drops the pressure during expiration. 

The other line of arguments, based on experience, emphasized that adding EPR in the above situation might benefit and reduce flow limitation even without increasing the expiratory pressure.

Below, I provide some measurements that support the latter view.

I run my AirCuve 10 with the following settings for 4 days: 
EPAP=5.4, IPAP=7.4, PS/EPR=2.0
With the following outcomes

FL(95%)=0.05; 0.01; 0.02; 0.03
Average FL (95%)=0.028
The standard deviation of FL (95%)= 0.017

FL(99.5%)= 0.26; 0.08; 0.10; 0.12
Average FL (99.5%)=0.14
The standard deviation of  FL (95%)= 0.082

I run my AirCuve 10 with the following settings for 4 days: 
EPAP=4.4, IPAP=7.4, PS/EPR=3.0
With the following outcomes

FL(95%)= 0.00; 0.00; 0.00; 0.00
Average FL (99.5%)=0.00
The standard deviation of FL (99.5%)= 0.00

FL(99.5%)= 0.08; 0.05; 0.05; 0.06
Average FL (99.5%)=0.06
The standard deviation of (95%)= 0.014

Outcome: 
When the EPR was increased from 2 to 3 by decreasing EPAP, the 95% flow limit was also reduced from 0.028 to 0.000. A similar FL decrease (from 0.14 to 0.06) can be observed at 99.5%. 

Conclusion: 
Adding EPR to the treatment plan is beneficial unless it permanently increases CAs. It is not necessary to increase the EPAP simultaneously. Nevertheless, the best practice is adding the EPR in the APAP regime when the instrument automatically increases pressure if, and only if,  it is needed. 

An added benefit of EPR is better exhalation efficiency, which reduces some inhalation limitations by more efficiently removing the stuck air from the lung.         


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#2
RE: Increased EPR reduces flow limits even when IPAP is NOT increased simultaneously.
I agree with what you're saying here, however it is possible to lower EPAP enough to lose airway patentcy. Also, the way EPR is set, it is possible to set minimum pressure lower than the available CPAP range, resulting in EPAP being stuck at 4.0 without the ability to respond to a need for higher pressure, so we always suggest minimum pressure is at least set to 4.0 cm plus EPR equals minimum pressure. The response of flow limits to increases in EPR and pressure support has been something I have advocated for a very long time, and is is also demonstrable that flow limits will not respond to positive air pressure increases in the same way they respond to pressure support increases. In nearly every case where flow limitation is reduced, individuals report better sleep integrity and better rest.

There has been a trend among the YouTube experts to advocate for fixed pressure, no EPR and the use of VCOM. None of these makes much sense to me based on results we have seen on the forum with bilevel pressure support or EPR. One of the guiding principles I have tried to live by here is that there is no single rule that applies to all. There is a great deal of individual response to any changes we make and those need to be observed and accommodated by being flexible with our suggestions. Positive Air Pressure therapy is inherently a trial and error experiment, from the titration protocols to more refined setting adjustments to improve an individual's comfort and efficacy. Another member Dormeo put is most simply, "EPR reduces pressure when you exhale, but by the same token, it increases pressure when you inhale. It's the increased pressure for inhalation that helps reduce flow limitation".
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: Increased EPR reduces flow limits even when EPAP is simultaneously decreased
   

Szabo.

I have decided to try EPR again.
Right off the bat, here's what has happened for 2 days..

Flow limitation went up, dramatically !?
Nov 12   7cm EPR 2      FL  0.47
Nov 13   7.4cm EPR 1    FL  0.43

I did not expect this to happen, but i have seen it in previous charts.
Like Sleeprider says, everyone is different etc. etc.

More to say probably, and test longer, but here's the twist......

I feel better today than i have in awhile !!

The test goes on...........will report results......

like
Post Reply Post Reply
#4
RE: Increased EPR reduces flow limits even when EPAP is simultaneously decreased
Thank you, SeePak, for reporting your results, 

However, please clearly state the settings you used for the two cases to benefit this community and myself.  It is not entirely clear. 

Best
G.
Post Reply Post Reply
#5
RE: Increased EPR reduces flow limits even when EPAP is simultaneously decreased
Szabo, yes i will elaborate, sorry....

Nov 10 and 11    CPAP pressure was 6.4 cm ,  i was NOT using EPR, pillow masks, 
Oscar results as per previously attached charts
Post Reply Post Reply
#6
RE: Increased EPR reduces flow limits even when EPAP is simultaneously decreased
   

Had to stop my new EPR trial and CPAP therapy using this machine.

Getting 'campfire' smell last 2 days, only 45 min 1st day of therapy and 3hrs i include here.

However, flow limit did drop a bit second night, but nothing conclusive here.

But i have felt great all day, which may seem impossible looking at this chart, but this is indeed happening !
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Stuffy Nose - Flow Limitation? mark.apnea 3 263 Yesterday, 05:16 PM
Last Post: radarrab
  Expiratory Flow Limitation? deebob 4 163 Yesterday, 03:35 PM
Last Post: Dormeo
Sad [Treatment] Weird Flow Rate/Minute Vent OK Laptop 1 623 11-28-2024, 09:08 AM
Last Post: ChadBSr
  Help Needed with CPAP Flow Chart Interpretation HijolG 19 1,056 11-23-2024, 03:25 PM
Last Post: HijolG
  Flow Limitations? gr3gg0 13 684 11-23-2024, 11:12 AM
Last Post: SeePak
  Worrying session last night - hypopneas increased tenfold! BigWing 15 1,093 11-21-2024, 11:28 PM
Last Post: notausername
  EPR 1, hight flow limit, but lots of CA throwaway1242 13 881 11-15-2024, 07:30 PM
Last Post: OpalRose


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.