Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Could My Aircurve 10 ASV Be Faulty From New?
#1
Could My Aircurve 10 ASV Be Faulty From New?
Three sleeps studies, 2 in 2016 and 1 in 2017.

Main problem = central apneas.

Provided with Aircurve 10 ASV. Third sleep study recommended EPAP 4-15, PS 0-20, Max Pressure 25cm (but I cannot see this setting as adjustable anywhere). Feb 2017

Subsequent testing for heart health specialist (because heart is supposedly stressed with ASV) saw a prescription of EPAP 5-10, PS 5-15.  Aug 2017

Original sleep doctor issued another prescription in Oct 2017 for EPAP 8-15, PS 0-20

Compliance report Dec 2017 showed settings at EPAP 8-15, PS 0-17. Oscar cannot read settings for this period up until I stopped using the machine in late 2021, however, when I accessed the clinician's menu in December last year those were the settings.

AHIs are frequently less than 1 on these settings.

PROBLEMS:

1. Don't really feel better after using it, especially when I wake, so stopped using it during an overseas trip when I could not get distilled water.

I started using it again early this year after noticing on my Samsung watch that my sleeping blood oxygen readings were often in the low 80s.

2. From early use found that on many occasions the machine would start "spurting" - my name for repetititve mask leaks.

Sometimes adjusting mask fit by hand pressure will reduce, but mostly not.

3. When repetitive mask leaks start due to high pressure they do not adjust when I breathe strongly, or change my breathing pattern - this flies in the face of claims that these are servo adjusting devices.

Switching machine off and then restarting by breathing normally (gently) corrects whatever the problem is.  Nothing else works.

4. I have not managed to find the methodology explanation for how these machines work (other than the very basic overview given in the ResMed manual) nor any information regarding most of the parameters shown by OSCAR and how they interact. As an electronics and computing diagnostic specialist for many years I would love to be able to intelligently make adjustments with some understanding of what I am doing.  :-)

Here are expanded views showing the build up in pressure and the pause when the machine is switched off and restarted by breathing, and the normal pressure prevailing following this.


Attached Files Thumbnail(s)
               
Post Reply Post Reply
#2
RE: Could My Aircurve 10 ASV Be Faulty From New?
I am not an expert by any means, but you should be able to test your machine works by holding your breath.

Go to your bedroom or where you have the machine. Put your mask on, start the machine and every 20 seconds hold your breath for 10 seconds or more (assuming you have no heart conditions), you will then feel the machine ramp up waves of pressure to try and induce breathing again. Once you have five minutes worth, pull out the SD card and load into Oscar.

I don't really have many events on my Oscar, so this is not an ideal example (its some sort of hypopnea),  but you should see something like my attachment, but even clearer

The flow rate will flat line and pressure will increase to tackle the event


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#3
RE: Could My Aircurve 10 ASV Be Faulty From New?
Thanks, but the problem is not that it doesn't increase air flow, but that it goes high and stays high unless I switch it off.
Post Reply Post Reply
#4
RE: Could My Aircurve 10 ASV Be Faulty From New?
Gotcha,

Do you know about the blowback technique?

https://www.apneaboard.com/forums/Thread...-ASV-reset
Post Reply Post Reply
#5
RE: Could My Aircurve 10 ASV Be Faulty From New?
Trevor, this is a known problem with the ResMed AirCurve 10 ASV, but people do overcome the problem after getting used to the device and making appropriate setting adjustments.    

Read from post #26 and on in my thread:

https://www.apneaboard.com/forums/Thread...ree?page=3
Post Reply Post Reply
#6
RE: Could My Aircurve 10 ASV Be Faulty From New?
(05-14-2023, 12:56 AM)stanleyjack Wrote: Gotcha,

Do you know about the blowback technique?

No, but I had tried doing that anyway. Thanks for the link.  :-)
Post Reply Post Reply
#7
RE: Could My Aircurve 10 ASV Be Faulty From New?
Trevor G I would prefer to see a couple changes to settings and get some feedback if it helps. You are using ASVauto mode with EPAP min, EPAP max 15 and PS 1.0-15.0. I'm certain you need a higher PS min and potentially lower PS max. Let's try EPAP min 8.0, EPAP max 11.0, PS min 3.0, PS max 12.0 and get some feedback on comfort and see a chart.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#8
RE: Could My Aircurve 10 ASV Be Faulty From New?
(05-14-2023, 05:00 PM)Sleeprider Wrote: Trevor G I would prefer to see a couple changes to settings and get some feedback if it helps.  You are using ASVauto mode with EPAP min, EPAP max 15 and PS 1.0-15.0.  I'm certain you need a higher PS min and potentially lower PS max.  Let's try EPAP min 8.0, EPAP max 11.0, PS min 3.0, PS max 12.0 and get some feedback on comfort and see a chart.

Thanks so much for taking the time - I'll try it tonight!
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Aircurve VAuto PS...pros & cons WW4B 13 515 10 hours ago
Last Post: Sleeprider
  Algorithm change in the ResMed AirCurve 10 ASV terp1984 9 778 Today, 03:04 AM
Last Post: alimama
  Aircurve 10 VAuto settings question dec3169 6 192 12-24-2024, 01:25 PM
Last Post: Deborah K.
  AirCurve ASV vs Vauto (is ASV king of the hill?) Carnal 3 200 12-23-2024, 12:27 PM
Last Post: Deborah K.
  Air leaks with Aircurve 11 ASV for CA chrishark 13 596 12-20-2024, 08:25 AM
Last Post: Sleeprider
  [Treatment] BIPAP Settings - ResMed AirCurve 11 Vauto dcbauman 20 899 12-17-2024, 03:42 PM
Last Post: Sleeprider
  Resmed Aircurve 10 ASV Rhodesie 2 183 12-16-2024, 10:43 PM
Last Post: Rhodesie


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.