[CPAP] Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
#11
RE: Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
(01-04-2025, 12:18 PM)Sleeprider Wrote: Your therapy was better with the CPAP with EPR. Can you go back to that?  The Lumis VPAP 150 ST does not have EasyBreathe available to make the IPAP/EPAP transitions more comfortable.  It also has a backup rate that can trigger IPAP on a timed basis.  Instead of EasyBreathe, we only have time of inspiration and rise-time controls.  To help with this particular device, we really need to know all of the settings.   Note this device is unlike the Vauto or VPAP S devices that lack backup rate but have Easybreathe and other controls.

my main machine is autoset 10

the lumis 150 is borrowed


the problem is i dont feel rested even when below 2 ahi


some examples from October attached.


Attached Files Thumbnail(s)
           
Post Reply Post Reply
#12
RE: Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
Your path to optimized settings may be bilevel, but it's not he Lumis 150 ST.  This is a device with a different intent. If you can get your hands on an Aircurve 10 Vauto or even the VPAP S, it will produce pressure very similar to what you experienced with the Airsense 10 with EPR, but with more pressure support and controlls for trigger, cycle and time of inspiration.  Take a look at the mask pressure chart for the VPAP ST and you will see it is a square-wave pressure change.  IPAP goes on and off very rapidly, which is why I need to know your current settings for things like rise-time. Look that this flow rate and mask pressure graph. Your flow rate is extremely flow limited because your ST pressure support comes on and stays constant until the cycle to EPAP.  Your flow rate (respiration) is a mirror of that.  This machine frankly sucks, and this problem is not unique to you.  ST is intended for use with patients with COPD and other pulmonary issues. That aside, your obstructive issues are clustered and it looks like a combination of upper airway resistance and positional apnea.

[Image: attachment.php?aid=73754]
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#13
RE: Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
Sleeprider is right.  The only reason I posted was because you suddenly changed machines from the Resmed S10 Autoset to the Lumis 150 ST; and I didn't know why this change occurred.
Download OSCAR
OSCAR Chart Organization
Attaching Files

Advisory Members serve as an "Advisory Committee" to help shape Apnea Board's rules & policies. 

Membership in the Advisory Members group does not imply medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.  
Post Reply Post Reply
#14
RE: Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
The lumis 150 ST
Has these modes:
Cpap
S
ST
T
IVAPS?
PAC?

I am using the S mode
with these settings
ipap 12
epap 8


ti max 2.0
ti min 0.3
rise time = min
trigger = med
cycler = med
Post Reply Post Reply
#15
RE: Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
As suggested above, it would be advisable to rent an Aircureve 10 Vauto like mine or an Aircureve 11 Vauto for a trial period. With these units, you can further reduce your flow limit and feel better. I apologize for being vague and simply recommending a bi-level device. I should have been more specific. 
Until then, you should return to your Airsense 10.
Post Reply Post Reply
#16
RE: Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
You will need to change Ti Max to 2.4, rise time to medium or max and trigger to high. If we see continued UA events, then use ST mode with a breath rate of 16.

The 150 ST is a very sophisticated and expensive ventilator class PAP. It is certainly not your best choice.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#17
RE: Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
Back to AutoSet at 10.

Currently using CPAP with a pressure of 11 and EPR set to 3, along with 100mg of 5-HTP and a nasal strip.

While there hasn’t been significant improvement in how I feel, the AHI numbers are at least showing better results—though I’m wondering if it’s just a fluke.
Should I consider increasing the pressure to 12 or 13?

It also seems like OSCAR might be reporting some false negatives.
I do have a high arched palate.


Attached Files Thumbnail(s)
           
Post Reply Post Reply
#18
RE: Experimented with CPAP Pressures, Minimal Improvement (Possible UARS)
Higher pressure is not going to hurt anything. Trialing higher pressure and keeping notes on how you feel may help to optimize the best pressure for you. With a Vauto, my starting pressure is 9.6 plus 4.4 PS, so 14/9.6 starting pressure. I don't fine increased pressure to be a problem, and it's what works best for me, however it's not for everyone. The only way to know is to try it and evaluate your options.

Sorry about the VPAP 150 ST fiasco, but your CPAP looks a lot better in both the overall therapy and zoomed views. If you ever get a chance to try an Aircurve 10/11 Vauto or S, it will be a better experience.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
Arrow AHI < 1.0 now, but still tired? Is it UARS, RERA, arousals or FL in your FR peaks? 2SleepBetta 66 17,918 Yesterday, 12:03 PM
Last Post: SeePak
Question Help interpreting OSCAR Flow Rate Data w/ UARS Primarysecondary465 0 97 01-04-2025, 11:33 AM
Last Post: Primarysecondary465
  Advice for dialing in pressures? sleephead777 8 440 01-04-2025, 11:21 AM
Last Post: Fdxmike
  Small change nice improvement. super7pilot 5 267 12-29-2024, 06:04 PM
Last Post: ejbpesca
  Aerophagia at my ideal pressures macaronno336 3 222 12-18-2024, 01:04 PM
Last Post: SarcasticDave94
  UARS or PLMD lexit957 2 196 12-16-2024, 09:24 AM
Last Post: DaveSkvn
  Zeev - Therapy Improvement? Zeev 11 541 12-01-2024, 04:54 PM
Last Post: Zeev


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.