Hi Apneaboard,

Attached is a PDF document directly from Philips that explains the various humidification options available in Dreamstation machines including temperature and humidity set points.

I have not been able to find this information in any Philips manuals or their website.

Of note is that both Fixed and Adaptive humidification modes do not appear to work when a heated tube is used.  A separate Heated Tube Mode is used instead with different temperature/RH setpoints.  This is not explained anywhere in the manual to my knowledge.

Hope this of some help to some of you.


Attached Files
.pdf   DreamStation Sleep Therapy System Humidification Options.pdf (Size: 73.32 KB / Downloads: 12)
Post Reply Post Reply
#2
RE: DreamStation Sleep Therapy Humidification Options (PDF)
Basically you can perceive heated tube mode as adaptive+heated tube.
And the temp and humidity sensor is near the air inlet so it should only measures the room temp and RH, the RH given in adaptive and HT mode should be an estimation, and I think it's quite roughly.
Post Reply Post Reply
#3
RE: DreamStation Sleep Therapy Humidification Options (PDF)
(09-06-2022, 09:23 PM)zellem Wrote: Basically you can perceive heated tube mode as adaptive+heated tube.
And the temp and humidity sensor is near the air inlet so it should only measures the room temp and RH, the RH given in adaptive and HT mode should be an estimation, and I think it's quite roughly.

That is a good way of putting it.  It does explain my (poor) experience with the humidification system ever since I first got my Dreamstation.  I always disliked the variability in the Adapative heating and would rather prefer the constant heating that Fixed mode provides.

However, even after changing the setting to Fixed I noticed that it didn't make much of a difference.  Now I know why!  I was using a heated tube the entire time.

Its dumb that the machine still lets you select Adapative and Fixed modes when using a heated tube.  They should have disabled those settings when a heated tube is attached so as to not confuse users.
Post Reply Post Reply
#4
RE: DreamStation Sleep Therapy Humidification Options (PDF)
Ironically if you use the Encore to view the device setting, you'll see the full content of that setting -- 'humidification mode on heated tube disconnect'.
Unlike the system one you can use a heated tube and disable the HT function, DS will mandatory switch to the heated tube mode.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Can Mild Sleep Apnea With AHI 14 Be Fatal ? Can A Sleep Apnea Sufferer Die In Sleep ? EastWeekender 11 491 11-24-2024, 08:00 AM
Last Post: SarcasticDave94
  Best Backup CPAP Machine Options? Lucid 14 645 11-14-2024, 10:21 AM
Last Post: jake79
  [Pressure] effect of humidification on felt CPAP pressure? Zombie on my Lawn 2 256 10-22-2024, 01:47 AM
Last Post: Zombie on my Lawn
Sad [Treatment] Therapy is helping, but only a little - good AHI numbers, bad sleep tiredsleeper 19 1,669 10-13-2024, 11:48 AM
Last Post: Dormeo
  HELP: Sudden drastic worsening of Sleep Apnea - Can't sleep at all, CPAP completely u Unkikonki 10 853 10-13-2024, 11:41 AM
Last Post: Dormeo
  Sleep study- Very high AHI with Central Sleep Apnea and multiple hours of CSR. Help. JoshO71 2 430 08-27-2024, 06:09 PM
Last Post: JoshO71
  Options to stop sleep apnea Timothy918 7 463 08-15-2024, 11:47 PM
Last Post: HalfAsleep


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.