RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
(11-16-2019, 06:01 AM)mper6794 Wrote: Hi, Joey
this is complex ! I would rather go for things I think would more simple. It won't cost me nothing, this opinion hereinbelow; it might help.

-Your charts suggest a great deal of long duration unflagged RERA's, too many which could be improved by properly adjusting your EPAPmin and PS.

_too many arousals/awakenings….hard to ask, but do you have any issue with RLS/PLMS issues? hope not....

Good luck
No RLS/PLMS.

(11-16-2019, 06:18 AM)bonjour Wrote: EERS is only to manage centrals. The xPAP is used to manage the obstructive events. And I missed your comment on increasing EPR needing a compensation in EPAP to compensate.  You are correct and that is the "secret"  to using a ResMed CPAP, Elite, or AutoSet as a limited BiLevel .

So yes, both need to be manipulated to achieve optimum results. Increasing PS and EPAP both have the capability of reducing your pCO2 in your blood and causing additional centrals.  Then the EERS needs to be adjusted to bring the centrals back down.

The centrals went up because the EPR increased thus decreasing the pCO2 in your blood which caused your Real Centrals.
Thank you, that's helpful.

(11-16-2019, 08:26 AM)Sleeprider Wrote: Results on the VPAP Auto look very good, and it's surprising not to see more emergence of centrals with the higher pressure support.  I have personally found that on the Vauto, my CA results are consistently better with a trigger sensitivity set to high rather than medium.  These are my results and  yours may vary.  If you are more comfortable with bilevel, stick with it, and add EERS if the CA re-emerges.  The flow rates look pretty good where leaks are controlled, and there is less flow limitation.  Your best tidal volume and respiration stability continues to be with the use of EERS, and it would be good to see you try that system with the VPAP.
It's surprising to me as well. It would be interesting to try that.


I slept two nights with another 6" of EERS but the first night I didn't have the SD card in. This is from last night, I woke up feeling good. I woke up after the last event. I turned the min pressure up to 12 because 11 and 10 didn't feel like enough (effect of more co2 increasing respiratory drive?). I could probably turn it up to 13 or even 14.


[Image: YcxqtEV.jpg][Image: I2DFbwD.jpg][Image: K0NmGpP.jpg][Image: 5CzvZbM.jpg][Image: m9vMvye.jpg][Image: He2xVmE.jpg][Image: RgLIAwr.jpg][Image: IFNo3ZN.jpg][Image: Lukzsq6.jpg][Image: akUVfLW.jpg][Image: 2VIrBsf.jpg][Image: zYMbak9.jpg][Image: eXc1q2v.jpg]
Nothing I post is medical advice and should not be taken as such, always consult a medical professional for guidance.
Post Reply Post Reply
#82
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
CA and flow limit looks excellent ant your tidal volume at 620 is pretty phenomenal. There is almost no change in pressure through the night, and that is going to contribute to very low disruption. If this is comfortable, stick with it.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#83
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
(11-18-2019, 08:49 AM)Sleeprider Wrote: CA and flow limit looks excellent ant your tidal volume at 620 is pretty phenomenal.  There is almost no change in pressure through the night, and that is going to contribute to very low disruption.  If this is comfortable, stick with it.

What do you make of the fact that his I:E ratio is inverted?
Post Reply Post Reply
#84
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
I noticed it, and don't know what to think of it. I would argue that none of the flow rate close-ups confirm that statistic. Also, all of the OA events are breath-holds, not preceded by an expiration. So most of those events seem to be arousal or movement.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#85
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
(11-18-2019, 09:33 AM)slowriter Wrote:
(11-18-2019, 08:49 AM)Sleeprider Wrote: CA and flow limit looks excellent ant your tidal volume at 620 is pretty phenomenal.  There is almost no change in pressure through the night, and that is going to contribute to very low disruption.  If this is comfortable, stick with it.

What do you make of the fact that his I:E ratio is inverted?

It looks a good question, slowriter. That had called my attention too. 

One of these days, I did a brief analysis of this E:I reations while moving from APAP to BPAP. It is here, on my thread, and other posts.
http://www.apneaboard.com/forums/Thread-...wer?page=4

all the best
Post Reply Post Reply
#86
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
(11-16-2019, 08:26 AM)Sleeprider Wrote: ... I have personally found that on the Vauto, my CA results are consistently better with a trigger sensitivity set to high rather than medium.  

I am curious.  

For those xPAP that can do it, I often see mention of setting trigger sensitivity to High to help reduce CAs but have not seen much discussion of adjustments to Cycle. 

If we set Cycle to Low would that shorten Expiration a little, and would shorting Expiration reduce CO2 washout?  So would setting Cycle to Low be a way to reduce CO2 washout and then in turn help reduce CAs?   

What have you guys learned about Cycle with regard to CAs?

Note:  I asked the question Cycle here on this thread because if the Inverse I:E ratio does become a concern adjusting Cycle might be one of the tools to help influence the I:E ratio.  

WillSleep

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#87
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
(11-18-2019, 08:49 AM)Sleeprider Wrote: CA and flow limit looks excellent ant your tidal volume at 620 is pretty phenomenal.  There is almost no change in pressure through the night, and that is going to contribute to very low disruption.  If this is comfortable, stick with it.
Thank you  Smile

(11-18-2019, 09:33 AM)slowriter Wrote:
(11-18-2019, 08:49 AM)Sleeprider Wrote: CA and flow limit looks excellent ant your tidal volume at 620 is pretty phenomenal.  There is almost no change in pressure through the night, and that is going to contribute to very low disruption.  If this is comfortable, stick with it.

What do you make of the fact that his I:E ratio is inverted?

(11-18-2019, 10:02 AM)Sleeprider Wrote: I noticed it, and don't know what to think of it. I would argue that none of the flow rate close-ups confirm that statistic.  Also, all of the OA events are breath-holds, not preceded by an expiration.  So most of those events seem to be arousal or movement.
Thinking-about Don't know either

I feel okay today, charts look bad (because of leaks I think). Didn't secure CPAP mask good.
Full data (password apneaboard) https://www.mediafire.com/file/6s0y6ovug...S.zip/file

Last night charts
[Image: B5WurYX.jpg][Image: ZwD7AsJ.jpg][Image: t0y2UJy.jpg][Image: jjG05pY.jpg][Image: cs1TZAn.jpg][Image: 26Ittpr.jpg]
Nothing I post is medical advice and should not be taken as such, always consult a medical professional for guidance.
Post Reply Post Reply
#88
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
Should I try BiPAP again? EERS has significantly improved both tidal volume and amount of central apneas but the BiPAP waveforms look significantly better to me. What do people think?

Auto CPAP EPR 3 with extra EERS (data from two nights)
[Image: 3tRAOVZ.jpg][Image: j7UHbKn.jpg][Image: PmR8Zn3.jpg][Image: V0gFyg3.jpg][Image: E3HbVZG.jpg][Image: hI19P1l.jpg][Image: Sm8ewtx.jpg]

Auto BiPAP pressure support 4 no EERS (data from two nights)
[Image: eSYEoON.jpg][Image: ID7uYxO.jpg][Image: ZLk07jf.jpg][Image: qavvesS.jpg][Image: dZNsPti.jpg][Image: 4fimxj7.jpg][Image: DbFOn2I.jpg]
Nothing I post is medical advice and should not be taken as such, always consult a medical professional for guidance.
Post Reply Post Reply
#89
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
More waveforms, they're interesting

Auto CPAP EPR 1 no EERS (data from one night) data from here #13
[Image: lxooLdj.png][Image: ykiKHp4.png][Image: eTpNXrg.png][Image: kqdN1Ef.png][Image: jYAMzYt.png][Image: 71QSxUe.png][Image: XFYOJtr.png]

Auto CPAP EPR 3 no EERS (data from one night) data from here #7
[Image: GuTy3Qs.png][Image: JRBcT07.png][Image: JB1clEi.png][Image: G7oCbVQ.png][Image: nR13JnM.png][Image: QbAnL66.png][Image: 9wk60fn.png]
Nothing I post is medical advice and should not be taken as such, always consult a medical professional for guidance.
Post Reply Post Reply
#90
RE: Interpreting sleep study results, EERS enhanced expiratory rebreathing space
Auto CPAP EPR 1 Standard EERS (data from one night) data from here #58
[Image: ewiLXvD.png][Image: h2gz8tS.png][Image: Yvw1Ead.png][Image: CCMZIjG.png][Image: tkQMBKI.png][Image: EGQMorp.png][Image: sdYNH7G.png]

Auto CPAP EPR 3 Standard EERS (data from one night) data from here #69
[Image: co5R1Sh.png][Image: owaylUY.png][Image: 2X3WyIw.png][Image: 3OuX2f2.png][Image: fGrUSa9.png][Image: 84oSXff.png][Image: WhFBIDd.png]
Nothing I post is medical advice and should not be taken as such, always consult a medical professional for guidance.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Expiratory Flow Limitation? deebob 4 156 Yesterday, 03:35 PM
Last Post: Dormeo
  Help interpreting results AlwaysProper 0 70 Yesterday, 10:22 AM
Last Post: AlwaysProper
Question [Diagnosis] Comparing AHI Accuracy: ResMed AirSense 11 vs. In-Lab Sleep Study NightmaresOfTheBottom 2 109 11-28-2024, 07:08 PM
Last Post: Rickyricardo
  Can Mild Sleep Apnea With AHI 14 Be Fatal ? Can A Sleep Apnea Sufferer Die In Sleep ? EastWeekender 11 494 11-24-2024, 08:00 AM
Last Post: SarcasticDave94
  PLS - Need some help with interpreting the sleep data from Oscar Laurenthepro 6 245 11-21-2024, 01:23 PM
Last Post: Deborah K.
  Given a sleep study, how to get a script? silverchief2 3 167 11-19-2024, 12:52 PM
Last Post: PeaceLoveAndPizza
  danger of rebreathing at lower pressures? Zombie on my Lawn 1 188 11-14-2024, 09:40 PM
Last Post: Sleepster


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.