[Equipment] Sleep Report on ResMed AirCurve 10 S
#1
Sleep Report on ResMed AirCurve 10 S
My ResMed AirCurve 10 S would shut down after about 30 mins of usage and I reset my machine thru the Settings menu which resolved the problem. However now, my Sleep Report is not fully displayed on the device and the Events/hour is only viewable online at myair.resmed.com


Any suggestions? I tried removing and re-inserting the SD card, however, the problem persists. The SD card is not defective when checked on my laptop.

Thanks
Alex

   
   
Post Reply Post Reply
#2
RE: Sleep Report on ResMed AirCurve 10 S
Go into the clinical menu and find essentials in the options menu and change its setting from on to plus

That should give you an enhanced sleep report
Post Reply Post Reply
#3
RE: Sleep Report on ResMed AirCurve 10 S
(01-25-2020, 01:39 AM)jaswilliams Wrote: Go into the clinical menu and find essentials in the options menu and change its setting from on to plus

That should give you an enhanced sleep report

Thank you, I just discovered that this morning.

Regarding the earlier issue of my machine shutting down after about 30 mins, and now functioning after a reset, any thoughts on this?

Thanks
Alex
Post Reply Post Reply
#4
RE: Sleep Report on ResMed AirCurve 10 S
I wonder if auto stop/start was enabled and your leaks were excessive and the machine thought you no longer had the mask on so turned it’s self off. If you have a PC and an SD card in your Aircurve then I recommend you download OSCAR details in my signature with the detailed graphs it may give us a clue why your machine turned off without that information we would just be guessing
Post Reply Post Reply
#5
RE: Sleep Report on ResMed AirCurve 10 S
(01-25-2020, 02:07 PM)jaswilliams Wrote: I wonder if auto stop/start was enabled and your leaks were excessive and the machine thought you no longer had the mask on so turned it’s self off. If you have a PC and an SD card in your Aircurve then I recommend you download OSCAR details in my signature with the detailed graphs it may give us a clue why your machine turned off without that information we would just be guessing

This was the first suggestion from my Aerocare consultant. Yes Auto start was on and I retried after turning it off. On this try, the CPAP shutdown as earlier in about 30 mins. I have downloaded OSCAR and will try diagnosing the information on the shutdown days. 

Its working fine after resetting the CPAP to factory defaults, and I have adjusted the pressure, however, I am not sure if any other settings need to be adjusted. My AHI was 5.6 so things seem to be good. 

Again thank you so much for your support.

Alex
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  airsense/aircurve interchangeability info eok361 2 63 5 hours ago
Last Post: eok361
  AirCurve power question Shamus54 0 44 Today, 05:04 AM
Last Post: Shamus54
  AirCurve power problem Shamus54 0 56 Yesterday, 08:24 PM
Last Post: Shamus54
Question [Diagnosis] Comparing AHI Accuracy: ResMed AirSense 11 vs. In-Lab Sleep Study NightmaresOfTheBottom 2 111 11-28-2024, 07:08 PM
Last Post: Rickyricardo
  Can Mild Sleep Apnea With AHI 14 Be Fatal ? Can A Sleep Apnea Sufferer Die In Sleep ? EastWeekender 11 499 11-24-2024, 08:00 AM
Last Post: SarcasticDave94
  AirSense 10 "Sleep Report" SleepyTimeTime 9 5,959 11-22-2024, 04:20 PM
Last Post: SleepyBull
  airsense 10/11 aircurve 10 heated tube only eok361 0 125 11-19-2024, 11:30 PM
Last Post: eok361


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.