Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Intermittent desaturations during sleep
#1
Intermittent desaturations during sleep
Just started using a new pulse ox device that vibrates when I desaturate <90. I've been getting 4-5 episodes nightly, moreso in the morning. My AHIs have averaged in the mid 2s. Most prevalent when I lay on my back as opposed to side sleeping. Dead ringer for upper airway collapse I would think. Using a chinstrap but am noticing an increase in dry mouth in the early mornings before waking. Should I bump up my pressure? What else can I do?
Post Reply Post Reply
#2
RE: Intermittent desaturations during sleep
Let's see your charts including oximetry.

with a dreamstation you have two options

1. more pressure
2. Increase supplemental oxygen
ok there is a third option.
3. Get a BiLevel which will improve your exchange

BUT these are general answers because you have not presented your charts.
Post Reply Post Reply
#3
RE: Intermittent desaturations during sleep
(04-22-2021, 06:08 PM)Gideon Wrote: Let's see your charts including oximetry.

with a dreamstation you have two options

1. more pressure
2. Increase supplemental oxygen
ok there is a third option.  
3. Get a BiLevel which will improve your exchange

BUT these are general answers because you have not presented your charts.

Last 2 nights:

[Image: RVKpfia.jpg]
[Image: NCWmUye.jpg]

[Image: kJ8S9Ku.png]
[Image: yixh593.png]
Post Reply Post Reply
#4
RE: Intermittent desaturations during sleep
In part, I think you are falling into the spike trap. Honestly, the spikes, which is what I see, don't matter. What matters is duration. How long were you under 93%, 90%, 88%, 85%, 80%, 75%? These values are there because the top four are used by some doctors as critical values. The lower two are there to indicate the really bad values. Do you have access to these types of values? The longer the duration and the lower the percentage the worse it is.

IMHO turn the alarm off and check your durations. Without question, the alarm is disturbing your sleep.

On your charts turn off the Piechart and calendar and we will see your settings. Take your screenshot with F12 as that will remove them take the shot then restore them after.
Post Reply Post Reply
#5
RE: Intermittent desaturations during sleep
(04-22-2021, 08:12 PM)Gideon Wrote: In part, I think you are falling into the spike trap.  Honestly, the spikes, which is what I see, don't matter.  What matters is duration.  How long were you under 93%, 90%, 88%, 85%, 80%, 75%?  These values are there because the top four are used by some doctors as critical values.  The lower two are there to indicate the really bad values.  Do you have access to these types of values? The longer the duration and the lower the percentage the worse it is.

IMHO turn the alarm off and check your durations.  Without question, the alarm is disturbing your sleep.

On your charts turn off the Piechart and calendar and we will see your settings.  Take your screenshot with F12 as that will remove them take the shot then restore them after.

[Image: rmttkom.png]
[Image: E1Lfvbl.png]
Post Reply Post Reply
#6
RE: Intermittent desaturations during sleep
Anyone help in interpreting my ODI (oxygen desaturation index)? 02 score doesn't look too bad (10 = zero hypoxia).


[Image: mqzgN9k.jpg]
[Image: 7AcPcga.jpg]
[Image: n4JtZyJ.jpg]
[Image: XLAOOmb.jpg]
Post Reply Post Reply
#7
RE: Intermittent desaturations during sleep
I'm only seeing brief downward spikes on the SpO2 charts. It didn't appear these last long enough to be a serious issue.
Mask Primer

Positional Apnea

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#8
RE: Intermittent desaturations during sleep
This is Viatom/Wellue data yeah? You can also import this into OSCAR, if you want to (as an aside).

I actually asked Viatom about this: ODI is the number of oxygen drops that you have per hour at a specific percent. So ODI 3% is the number of drops in an hour of at least 3 percent from your baseline SpO2 measurement. Most often, this drop is required to happen within a specific time frame (such as 8 seconds), but Viatom's software seems to measure all of them, no matter how long they last.

I would definitely recommend turning OFF the vibration warning for now - at least to get an idea of if the drops would actually turn into anything problematic, or if they are just the occasional spike. With pretty much 100% of your time spent between 90 and 100%, and your average being at 94, that's not particularly worrisome.
Post Reply Post Reply
#9
RE: Intermittent desaturations during sleep
That 85 SpO2 value appears to be an artifact. You probably bumped the sensor. To me, your overall low is around 88%.
Crimson Nape
Apnea Board Moderator
Project Manager for OSCAR - Open Source CPAP Analysis Reporter
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#10
RE: Intermittent desaturations during sleep
Note if it's at or below 88% for over 5 minutes, they're putting you on supplemental oxygen. I haven't seen it that low consistently on these charts.
Mask Primer

Positional Apnea

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Sleep Pillow (Hack for less leaks and a good sleep) Canuck 2 3 458 12-19-2024, 07:09 PM
Last Post: Canuck 2
  SP02 Baseline and desaturations calculation gdenton2000 1 293 10-25-2024, 03:33 PM
Last Post: PebbelsinCH
  HELP: Sudden drastic worsening of Sleep Apnea - Can't sleep at all, CPAP completely u Unkikonki 10 946 10-13-2024, 11:41 AM
Last Post: Dormeo
  Sleep study- Very high AHI with Central Sleep Apnea and multiple hours of CSR. Help. JoshO71 2 464 08-27-2024, 06:09 PM
Last Post: JoshO71
  Sleep position affecting sleep test AHI/OSA classification Fox66 7 577 06-21-2024, 12:02 PM
Last Post: Fox66
  Desaturations on APAP Songbird 4 482 06-11-2024, 03:39 PM
Last Post: UnicornRider
  Just had a sleep study - Central Sleep Apnea with Cheyne Stokes srpmccarthy 8 1,139 05-21-2024, 03:06 PM
Last Post: HenryECole81


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.