I'm running out of upload quota. How do you guys usually manage that? Upload and link to an outside source?
Thank you again.
Hello Guest,
Welcome to Apnea Board !
[Pressure] Many Settings - Same Result - OSCAR Detectives take a look?
|
09-03-2021, 08:42 AM
RE: Many Settings - Same Result - OSCAR Detectives take a look?
Is this what you had in mind?
I'm running out of upload quota. How do you guys usually manage that? Upload and link to an outside source? Thank you again.
09-05-2021, 07:49 PM
(This post was last modified: 09-05-2021, 07:50 PM by BionicBreather.)
RE: Many Settings - Same Result - OSCAR Detectives take a look?
Attached are 5ish minute views of the flow curve of events from the last two evenings. Also attached are PDF files from the pulse oximetry software.
I have resized these images to reduce storage size. Any input is welccome. Thanks for having a look Another 3 images.
09-05-2021, 08:10 PM
RE: Many Settings - Same Result - OSCAR Detectives take a look?
The oximetry reports appear to be from the Contec-50 series. If so, you can import the ".SpO2" binary data files into OSCAR to be included with your sleep data.
Crimson Nape
Apnea Board Moderator Project Manager for OSCAR - Open Source CPAP Analysis Reporter www.ApneaBoard.com ___________________________________ Useful Links -or- When All Else Fails: The Guide to Understanding OSCAR OSCAR Chart Organization Attaching Images and Files on Apnea Board Apnea Helpful Tips INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
09-07-2021, 08:21 AM
RE: Many Settings - Same Result - OSCAR Detectives take a look?
(09-05-2021, 08:10 PM)Crimson Nape Wrote: The oximetry reports appear to be from the Contec-50 series. If so, you can import the ".SpO2" binary data files into OSCAR to be included with your sleep data. Ah yes! It does. The link on your profile regarding screenshots was super helpful. I have been doing this with PRNTSCRN in windows and then tediously fine tuning the pictures in Paint. For the next few nights I'll be running the pulseox in sync with my cpap and then upload the deep dive results Gideon was requesting. This brings me to an important question regarding the forum: There's a three image limit per post to (obviously) protect the server. But I can reply to my own posts and upload an additional three images, rinse and repeat. Because my sys admin skills are at a cave man level, I don't know if this "work around" is actually a negative for the server's expenses/capacity. The flow limit deep dive is probably going to require more than three images. Is replying to my own posts with additional images against the forum rules?
09-17-2021, 07:01 AM
(This post was last modified: 09-17-2021, 07:04 AM by BionicBreather.)
RE: Many Settings - Same Result - OSCAR Detectives take a look?
(08-31-2021, 09:44 PM)Gideon Wrote: Deep Dive. 1. Imported Pulse Ox and used the align with CPAP data setting as recommended by OSCAR. i. Set SpO2 Y-Axis to 90-100 for more ganular view. Pulse Ox baseline was 96. ii. Can see some de-sats below 90, appear to last only a few seconds. Do not repeatedly correlate with OA or Hypos. Screenshot attached. iii. Total time in desat is 1.72% for this night. 2. View set to 5 minutes. i. 14 arousals. ii. Cannot find a regularity to what proceeds OAs, Hypos. Reading the the flow curve it does not appear rounded at the top as the Wiki states is ideal. I'm not clear if I've set up the Pulse Ox correctly. I get the sense its reporting paramters can be adjusted (baseline, what counts as a desat) and I'm unlcear if its relevant or needs fine tuning to get better data. 3. Increase pressure to "S" ? Would you please point me to what that is. Is "s" a mode of some kind? 4. Attached oodles of screenshots. More documentation. 5 minute views as requested of each event. Another 3. Last 3.
10-03-2023, 09:17 AM
RE: Many Settings - Same Result - OSCAR Detectives take a look?
Your EPAP pressure is not high enough to adequately control your obstructive events - do you have your sleep study results to share where you were titrated at? Looks like you might consider increasing your EPAP / IPAP +1cmH20 per night until the light blue events disappear. For example, tonight you'd want to try EPAP 5, IPAP 10 PS 5 and if you still see prevalent obstructive events then try EPAP 6, IPAP 11 PS 5
10-16-2023, 10:53 AM
RE: Many Settings - Same Result - OSCAR Detectives take a look?
Thank you! Thank you! Thank you!
I have set my epap / ipap at 7 / 13 and feeling a much better. I didn't realize how hard I was working to breath until I tried this setting. |
« Next Oldest | Next Newest »
|
Possibly Related Threads... | |||||
Thread | Author | Replies | Views | Last Post | |
OSCAR Result Interpretation & Clear Airways | Steven Alexandro | 17 | 681 |
Yesterday, 11:23 AM Last Post: Dormeo |
|
Titration Result Analysis | indreams | 0 | 137 |
10-31-2024, 08:11 PM Last Post: indreams |
|
Airsense10 AHI result | yoyoman2024 | 7 | 456 |
10-15-2024, 06:43 PM Last Post: yoyoman2024 |
|
[CPAP] Could use some help with OSCAR analysis/interpretation re: Airsense11 settings | thesquimmgroup | 6 | 465 |
10-14-2024, 01:28 PM Last Post: thesquimmgroup |
|
Question about IPAP settings on OSCAR | Dreamless | 6 | 386 |
08-31-2024, 05:11 PM Last Post: Dreamless |
|
Therapy Result Changes | stevesleep | 11 | 663 |
08-12-2024, 06:10 AM Last Post: stevesleep |
|
[Pressure] Help with pressure settings please (OSCAR data included) | jc12312 | 2 | 264 |
07-26-2024, 09:02 AM Last Post: jc12312 |