Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Low AHI vs. low Flow Limitation
#11
RE: Low AHI vs. low Flow Limitation
Ok.back with nights worth of data based on the recommended settings...The  difference is that I switched pillows from Airfit P10 to dreamwear and took an allegra first two nights..(forgot to take for 3rd night)


                         

Zoomed breathing samples...

                       

Looking forward towards your recommendations...

Thanks again
Post Reply Post Reply
#12
RE: Low AHI vs. low Flow Limitation
Flow limitation continues to be prominent in your results. It would probably take bilevel pressure to resolve. Results are slightly improved or comparable to your original results at fixed pressure 7, EPR 1, except tidal volume is improved. Inspiration ratio remains 1.23 to one, and we would ideally like to see between 0.5 and 1.0 to one. You have low AHI, so your pressure and settings may be optimized for CPAP. Without more information on what the source of the flow limitation might be, I'm at a loss for additional suggestions.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#13
RE: Low AHI vs. low Flow Limitation
Just got my new Resmed AirCurve 10 Vauto in an attempt to reduce Flow Limitations...Screenshots below...Any suggestions???Thanks!!


               
Post Reply Post Reply
#14
RE: Low AHI vs. low Flow Limitation
Change trigger sensitivity to high or very high to reduce the centrals. No other changes needed.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#15
RE: Low AHI vs. low Flow Limitation
Hi again...just to followup on your suggestion to increase trigger to very high...

Changes made: Trigger-very high

                        TiMax-From 2.8-2.6

                         Lowered PS to 4.0

                                    

cannot figure out which change was most signifcant....It appears to me that the pressure settings are more like a fixed bilevel than auto..Is that correct?

If these settings are ok...should I gradually begin to increase PS to reduce Flow limitations gradually .2 at a time?

Thanx again for your valuable insights and suggestions Thanks
Post Reply Post Reply
#16
RE: Low AHI vs. low Flow Limitation
Yes, if you want pressure to vary you will need to increase the max IPAP pressure. EPAP 8 plus 4.0 PS equals 12, so pressure is currently fixed. I think it would be fine to increase IPAP max to 14.0 and that might help with the remaining hypopnea. This looks a lot better.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Flow Limitations? gr3gg0 9 358 4 hours ago
Last Post: Deborah K.
Exclaimation [CPAP] Increased EPR reduces flow limits even when EPAP is simultaneously decreased G. Szabo 5 395 11-16-2024, 09:16 PM
Last Post: SeePak
  EPR 1, hight flow limit, but lots of CA throwaway1242 13 750 11-15-2024, 07:30 PM
Last Post: OpalRose
  Low AHI, high flow limits theswerve 1 166 11-15-2024, 03:30 PM
Last Post: PeaceLoveAndPizza
  Flow rate analysis jake79 4 257 11-12-2024, 12:27 PM
Last Post: jake79
  Stuffy Nose - Flow Limitation? mark.apnea 2 189 11-12-2024, 11:20 AM
Last Post: G. Szabo
  Flow Rate and Flow Limit Graphs ovmeadows 22 4,376 11-11-2024, 07:34 PM
Last Post: ovmeadows


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.