Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Mother's ResMed Astral 150 not keeping oxygen level high enough/what should I ask?
#1
Question 
Mother's ResMed Astral 150 not keeping oxygen level high enough/what should I ask?
Mother's oxygen level stays high enough (93-98) on cannula with concentrator or tank.

When using ResMed Astral 150 (with full face mask), her oxygen level has started to not get high enough, (high 70's or 80's). The ResMed used to keep her oxygen levels high enough.

What ought I ask the RT from Apria and the RN from pulmonologist to make sure the ResMed ventilator gets checked out properly?

Thanks very much.


P. S. When her level is too low, I have her take long, deep breaths, and she adjusts her recliner so she is sitting up more. That only helps a little. If you have any other remedies, I would like to have more helps at hand.


Thank you very much.
====
Am caring for nonagenarian parents:  one uses a ResMed Airsense 11, and the other a ResMed Astral 150, full face masks, oxygen, humidifiers, modems.

Anyone else caring for very elderly loved ones?
Post Reply Post Reply
#2
RE: Mother's ResMed Astral 150 not keeping oxygen level high enough/what should I ask?
I don't know if you're already adding the oxygen into the Astral 150 (only the 150 has this) however that is what can be done with the correct oxygen hose. I'd probably start there, if oxygen isn't attached to the Astral 150 now.

It would likely just clip into the oxygen port. I do think there's settings regarding oxygen with the Astral 150, including adding an FIO2 sensor.

Page 21 of the Astral manual cautions to turn off oxygen, start Astral, then turn oxygen on. Reverse for turning off, ox off then Astral.

See images.

   

   
Mask Primer

Positional Apnea

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: Mother's ResMed Astral 150 not keeping oxygen level high enough/what should I ask?
Sarcastic Dave, thanks for your kind help.

I forgot to write that the current oxygen input is through a valve at the back of the machine.

The Rt from Apria called and is overnighting an elbow fitting that will go on the front of the machine between the filter and the hose.

She is to use the cannula under the face mask until the adaptor arrives.

That is plan A.
====
Am caring for nonagenarian parents:  one uses a ResMed Airsense 11, and the other a ResMed Astral 150, full face masks, oxygen, humidifiers, modems.

Anyone else caring for very elderly loved ones?
Post Reply Post Reply
#4
RE: Mother's ResMed Astral 150 not keeping oxygen level high enough/what should I ask?
OK. The one on the back was what I was referring. It's likely down to see if the more direct method of oxygen brings levels up. Keep us updated and best wishes it works great.
Mask Primer

Positional Apnea

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#5
RE: Mother's ResMed Astral 150 not keeping oxygen level high enough/what should I ask?
I don't think we can see data from the Astral 150, but there are several possibilities that SpO2 is not being maintained. In general, oxygenation is improved with increased PEEP (positive end expiratory pressure) provided by a ventilator, while CO2 exchange and ventilation is enhanced with pressure support or IPAP. When an oxygen bleed is used with a ventilator, we have to account for the dilution caused by excess flow that is vented from the mask vent. With a cannula, oxygen is delivered directly to the nose, but with any positive pressure device, that oxygen is highly diluted, meaning the fraction of inspired oxygen (FiO2) will be lower, even with higher bleed rates. We can roughly calculate the amount of oxygen that must be added to a circuit to achieve the FiO2 requirement, and we can titrate oxygen bleed rates by observing SpO2. You have clearly observed that more oxygen is needed, which means you need to turn up the bleed rate (oxygen delivery in Liters/minute) to a higher rate until SpO2 stabilizes above 90%. So it's possible machine settings are not optimized, and likely that the oxygen generator needs to be adjusted to provide a higher oxygen feed rate. I recommend you read our wiki on oxygen bleed with positive pressure to better understand how this works. https://www.apneaboard.com/wiki/index.ph..._with_CPAP
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  LIQUID OXYGEN LEGISLATION srlevine1 0 95 12-17-2024, 01:35 AM
Last Post: srlevine1
  VCOM & Bi-Level MrIvanDrago 49 2,709 12-14-2024, 01:30 PM
Last Post: ChadBSr
  Medical Evidence of High Res Data...a physician perspective Loggerhead 10 452 12-13-2024, 07:12 PM
Last Post: notausername
  how high was your AHI/RDI and how long did it take for you to sleep with CPAP? mugen4u 6 291 12-03-2024, 06:32 PM
Last Post: Gary1of2
  Can high loop gain be changed? NoddingHacker 10 1,289 12-01-2024, 08:57 PM
Last Post: bh1332
  [Treatment] Central Sleep Apnea at high elevation knuddr 35 4,992 11-30-2024, 04:45 PM
Last Post: knuddr
  Huge SpO2 dips at sleep onset (with and without bi-level) THEVGE 19 864 11-29-2024, 10:32 AM
Last Post: robbob2112


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.