#31
RE: Dreamstation 2 vs. Aircurve 10 (Dry, Puffy Eyes)
I don't necessarily trust the numbers and prefer to look at representative samples of the flow rate chart at a zoom level of 3-4 minutes. I have found that the flow limit index is a relative metric and shows how much flow limitation increases for an individual, and is not comparable to other individuals. We often shoot for a 95% flow limit below 0.10, but I have moved away from that without some analysis of the flow chart. If someone has persistently flat, flow limited inspiration they may have a very low flow limitation index, but someone with a normal, rounded inspiration curve with occasional flow limited periods will have a higher flow limit index. That said, in the mile-high view of flow, you seem to have a balanced inspiration / expiration and the 95% of 0.1 is very good as is the I:E ratio.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#32
RE: Dreamstation 2 vs. Aircurve 10 (Dry, Puffy Eyes)
Thank you for all of your input @SleepRider. You have really helped me out with things this week. Appreciate it!
Post Reply Post Reply
#33
RE: Dreamstation 2 vs. Aircurve 10 (Dry, Puffy Eyes)
I wanted to update this thread on my recent changes and findings. After battling a slight cough and cold last night, I got a solid night of therapy in with the settings that @SleepRider has suggested. I switched to the F30i Mask from the F30, and it seems to have helped my eyes along with the eye drops that @SleepRider recommended. I also changed the humidity from 'Auto' to 'Manual' and set it to '5' and upped the tube temp by one degree as well, and no rainout and helped a little with dry mouth.

I think things are trending in a good direction after the tweaks in the past few weeks and about nine months into PAP therapy. Attached is an OSCAR screenshot of last night, and if there are any additional tweaks, I am always open to suggestions. Thank you all!


Aircurve 10 (16.0/9.0 PS4.0)

[attachment=55898]
Post Reply Post Reply
#34
RE: Dreamstation 2 vs. Aircurve 10 (Dry, Puffy Eyes)
MrIvanDrago - You keep duplicating your post from yesterday. Please do not do that.

- Red
Crimson Nape
Apnea Board Moderator
Project Manager for OSCAR - Open Source CPAP Analysis Reporter
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#35
VAUTO Settings vs. CPAP Settings
Quick question. 

On the ResMed Aircurve, is VAUTO mode 15.0/5.0 PS 3.0 the same as on a Resmed APAP 5.0-15.0 EPR 3.0?
Post Reply Post Reply
#36
Rainbow 
RE: VAUTO Settings vs. CPAP Settings
No. EPR is subtracted from the IPAP pressure for the EPAP pressure. PS is added to the EPAP pressure for the IPAP pressure.

Ergo, Vauto of EPAP 5 IPAP max 15 PS 3 is equal to Autoset range 8-15 EPR 3. For Vauto add PS of 3 to the EPAP (5+3=8 IPAP) up to IPAP max (15). For Autoset subtract 3 from IPAP pressure (8-3=5 for EPAP) up to IPAP pressure max of 15.

Confusing enough?
Post Reply Post Reply
#37
RE: VAUTO Settings vs. CPAP Settings
(11-17-2023, 05:56 PM)PeaceLoveAndPizza Wrote: No. EPR is subtracted from the IPAP pressure for the EPAP pressure. PS is added to the EPAP pressure for the IPAP pressure.

Ergo, Vauto of EPAP 5 IPAP max 15 PS 3 is equal to Autoset range 8-15 EPR 3. For Vauto add PS of 3 to the EPAP (5+3=8 IPAP) up to IPAP max (15). For Autoset subtract 3 from IPAP pressure (8-3=5 for EPAP) up to IPAP pressure max of 15.

Confusing enough?
I think I get it, so then how would you set APAP 5.0-15.0 EPR 3.0 on the Aircurve to mimic those settings?
Post Reply Post Reply
#38
RE: VAUTO Settings vs. CPAP Settings
(11-17-2023, 05:31 PM)MrIvanDrago Wrote: Quick question. 

On the ResMed Aircurve, is VAUTO mode 15.0/5.0 PS 3.0 the same as on a Resmed APAP 5.0-15.0 EPR 3.0?

No. In bilevel terms, the Vauto is working at 8.0/5.0 to 15.0/12.0;

the Resmed APAP settings of 5-15 with EPR 3 are 5.0/4.0 to 15.0/12.0. So until pressure is over 8.0 you won't see the minimum Vauto pressure of 8.0/5.0.  In other words, the low pressures on the Autoset are essentially unresponsive below 7.0 cm in terms of raising EPAP. Also the Vauto offers fine-tuning of TiMin, TiMax, and trigger/cycle sensitivity which is not available on the Autoset.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#39
RE: VAUTO Settings vs. CPAP Settings
(I noticed SleepRider explained it before I posted, but I’ll do so anyway as it may fill in some of the finer details)

As CPAP pressure does not go below 4 cmH2O, if you have you Autoset (APAP) set to 5-15 EPR 3, the pressure would work as follows:

EPAP 4 IPAP 5

Then as pressure goes up for whatever reasons

EPAP 4 IPAP 6
EPAP 4 IPAP 7
EPAP 5 IPAP 8
EPAP 6 IPAP 9

And so on until IPAP max 15 is reached.

On the AirCurve the closest you can come is to set EPAP to 4 and IPAP max to 15 with a PS of 3. This results in:

EPAP 4 IPAP 7
EPAP 5 IPAP 8
EPAP 6 IPAP 9

Until IPAP max of 15 is reached. As you can see not quite the same thing.
Post Reply Post Reply
#40
RE: VAUTO Settings vs. CPAP Settings
Ok this all makes sense now. Thank you for the input. My point to this question if you were wondering is I am trying to adjust my Aircurve to the lowest possible pressures to get benefit. The reason behind this is complicated, but the theory is the higher pressures are an issue for me personally causing other problems. I was not having issues when I was on my cpap, but my AHI I thought could be improved, it was always below 5, but in the 1-3 range with many hypopneas.

Last night I adjusted my Aircurve to 15.0/4.0 PS4.0. If someone has an idea to get it closer to my old setting on the Dreamstation (15.0-5.0, CFLEX+1) OR to even further simplify my settings, please feel free to suggest, many of you have been of great help!

Last night OSCAR:

[attachment=56231]
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Colorado Apnea Journey - Complex and Mostly CSA Questions CLCCW 19 670 12-21-2024, 11:32 AM
Last Post: CLCCW
  Been on CPAP therapy for 3 years feel worse than ever faixapreta 26 1,010 12-16-2024, 05:09 PM
Last Post: G. Szabo
  Therapy optimization and CPAP tuning Duck020 2 319 11-16-2024, 12:12 AM
Last Post: Duck020
  My Personal CPAP Journey SilverGuy 5 326 11-15-2024, 10:52 AM
Last Post: SilverGuy
  Back sleeping negatively effect CPAP therapy? sleepingbetterinFl 4 415 11-11-2024, 12:22 PM
Last Post: Big Guy
  Recently started CPAP therapy (AHI 56.9) naani 15 1,031 11-03-2024, 11:08 AM
Last Post: G. Szabo
  My journey to eliminate leak induced arousals PeterJohn 12 754 10-31-2024, 10:12 AM