Hi, all! I am preparing for my doc visit tomorrow and am looking at my data in Rescan, which I usually ignore in favor of sleepyhead. Can anyone tell me what I am seeing here? No events around this time at all.
Thanks. Charlie Sue[attachment=2297]
03-21-2016, 07:55 PM
RE: Rescan flow rate graph interpretation
That's not flow. It's flow limitation. Flow limitation is a scoring on a 0 to 1 scale of the shape of the inhalation waveform.
Note the little waveform pictures on the left axis. When the inhalation waveform "flat-tops" it indicates flow limitation. The Autoset should raise pressure on flow limitation. Admin Note:
JustMongo passed away in August 2017
Click HERE to read his Memorial Thread ~ Rest in Peace ~
03-21-2016, 07:58 PM
RE: Rescan flow rate graph interpretation
Hmmm.. kind of looks like it could be your pressure graph, esp: since your pressure is fairly low and close..(7-10). It would probably jump up and down that much since your pressure is that low and the setting is only 3 apart. . That's fairly low for apnea settings, but as a newbee to CPAP ya gotta start somewhere. Good Luck tomorrow..
Yesterday is history; Tomorrow is a mystery; Today is a gift; Thats why its called "The Present".
03-21-2016, 09:42 PM
RE: Rescan flow rate graph interpretation
I think Mongo's on the right track. That's the flow limitation chart. From Resmed's site: The S9 AutoSet responds to flow limitation to pre-emptively treat apnea events and prevent more serious, subsequent events from occurring.
03-21-2016, 10:05 PM
RE: Rescan flow rate graph interpretation
Hi Charlie Sue,
Good luck to you at your Dr. appointment, tomorrow.
trish6hundred
03-22-2016, 07:10 PM
(This post was last modified: 03-22-2016, 07:11 PM by 0rangebear.)
RE: Rescan flow rate graph interpretation
(03-21-2016, 07:55 PM)justMongo Wrote: That's not flow. It's flow limitation. Flow limitation is a scoring on a 0 to 1 scale of the shape of the inhalation waveform. justMongo is right ! Go to this link to learn more http://www.apneaboard.com/wiki/index.php...imitations
2004-Bon Jovi
it'll take more than a doctor to prescribe a remedy Observations and recommendations communicated here are the perceptions of the writer and should not be misconstrued as medical advice. |
« Next Oldest | Next Newest »
|
Possibly Related Threads... | |||||
Thread | Author | Replies | Views | Last Post | |
Flow Limitations? | gr3gg0 | 3 | 91 |
1 hour ago Last Post: G. Szabo |
|
EPR 1, hight flow limit, but lots of CA | throwaway1242 | 13 | 614 |
5 hours ago Last Post: OpalRose |
|
Low AHI, high flow limits | theswerve | 1 | 58 |
9 hours ago Last Post: PeaceLoveAndPizza |
|
[CPAP] Increased EPR reduces flow limits even when EPAP is simultaneously decreased | G. Szabo | 4 | 288 |
11-14-2024, 12:15 PM Last Post: SeePak |
|
OSCAR Leak Rate Question | AndyB | 5 | 276 |
11-13-2024, 10:45 AM Last Post: Epark617 |
|
[CPAP] OSCAR Data Interpretation Help | D Who117 | 0 | 100 |
11-12-2024, 08:43 PM Last Post: D Who117 |
|
[CPAP] Pimp My CPAP! (Please rate my data) | walshy1009 | 6 | 286 |
11-12-2024, 06:58 PM Last Post: walshy1009 |