As you say, both your charts report as the VAuto machine. If you are using both machines in the same night, perhaps try creating different profiles for each machine to make sure there is no confusion between them. Import each SD card into each of the different profiles.
One thing I notice from your charts is that you have no obstructive apneas, few hypopneas and mostly or all central apneas. So the first question I must ask is whether the ST-A machine with iVAPS mode is, in fact, the correct machine and mode for you, or whether an ASV is more appropriate.
The next thing I need to ask is what your underlying objective is? I note that you said in your original post that you have "shallow breathing", which leads me to anticipate that you would like to increase your tidal volume. Tidal volume is generally increased by increasing pressure support, but you can do that with either the VAuto or the ST-A machines in VAuto, S, ST or iVAPS modes. So let's first explore the modes a bit.
- With VAuto, the min EPAP (Expository Positive Air Pressure) is set to overcome obstructive apneas (OAs), but it is increased by the auto algorithm if OAs, flow limits or snore occur. It has a constant PS (Pressure Support) added to determine the IPAP (Inpository Positive Air Pressure, IPAP = EPAP + PS). The change between EPAP and IPAP is determined by detection of your spontaneous breathing.
- With S mode, both the EPAP and IPAP are fixed. The PS is simply the difference between them (i.e. PS = IPAP - EPAP). The pressure change between EPAP and IPAP is determined by detection of your spontaneous breathing.
- ST mode is the same as S Mode except that the change between EPAP and IPAP is determined by detection of your spontaneous breathing only if you are breathing at a rate faster than the backup rate, which can be either fixed or variable if you are using "intelligent" backup rate (iBR). If not, the change from EPAP to IPAP is triggered automatically at the backup rate. The change from IPAP to EPAP is prevented for the duration of Ti min and occurs automatically if it has not already happened spontaneously before Ti max.
- iVAPS is similar to ST mode. The difference is that PS is variable, not fixed. The purpose of varying PS is to achieve an "assured" (i.e. specified minimum) volume.
- iVAPS can also be similar to VAuto, with the addition of backup rate and variable PS, if Auto EPAP is enabled. However, there is a possible problem with the AutoEPAP option because MaxEPAP cannot be less than 8 (experience from my machine, not something I have found to be documented. Perhaps this is not true for your machine).
Now, you might think that iVAPS is the perfect mode if your objective is to obtain a particular tidal volume. But is is not used often because (1) it is complex, (2) it leads to rapid pressure changes and (3) it is not often necessary because perfectly good results can be obtained with S or ST modes. For example, the reason iVAPS is correct for me is that my minute vent changes during the night (it drops by 40 to 50% during REM) so the min PS (that I could apply in S or ST mode) works fine during non-REM but is not sufficient during REM. My SpO2 follows my Mv - and keeping that above 90% that is ultimately what I want to achieve.
Looking at the charts that you provided:
- You had no obstructive apneas and your pressure barely changed throughout the night, so I would suggest that you probably do not need an algorithm that auto adjusts EPAP.
- Your respiratory rate, tidal volume and minute vent trends were all reasonably steady for the whole night, so you might not need auto adjustment of PS.
So I am going to suggest that you try the following settings on your ST-A machine:
Mode: ST
IPAP: 10 (increase this if your Mv or Vt is lower than you need. Refer to your median values in the left panel if your chart trends are approximately constant)
EPAP: 6 (so PS = 4, while IPAP = 10)
iBR: On
TPR: 12 (this sets your min RR to 8)
Ti Max - Check the setting on your VAuto, it should be the same as that
Ti Min - Check the setting on your VAuto, it should be the same as that
Rise Time - You do not have a Rise Time setting on the VAuto so the trick here is to zoom right in on your mask pressure for the VAuto and the ST-A and adjust Rise Time to make the shape of increase in the pressure wave look the same. This is possibly the maximum value that can be set relative to Ti Min (bacause it cannot be as long as or longer than Ti Min)
Trigger - Check the setting on your VAuto, it should be the same as that. Otherwise Medium.
Cycle - Check the setting on your VAuto, it should be the same as that. Otherwise Medium.
Oh, I must point out that I am working off the clinical guide for the Lumis 150 ST machine. It is a little different to yours as I do not have ST-A mode, but ResMed did confirm for me that the "-A" simply refers to alarms, not therapeutic function, so I think I have what I need. If I do say something that you do not have in your machine, please let me know.
Getting ST mode working first is one step in the strategy that I used to get iVAPS working. I used ST mode to determine the effects of changing:
- IPAP
- Fixed backup rates
- Target patient rates when using intelligent backup
- Rise Time and Ti Min
- Trigger and Cycle
- Different masks
To monitor the effects of each change (only one at a time, for a couple of nights then I reversed the change), I found I needed to track the rates of spontaneous trigger and cycle. Unfortunately, these are not reported in OSCAR so I do the following:
- View the Sleep Report on the machine.
- Make sure the reporting period is set to "Daily"
- Scroll right down to the bottom of the list to see the values
- Record them in a spreadsheet - along with all of the settings or other changes.
Once I had my best ST mode settings configured, I switched to iVAPS with target volumes less than my typical volumes but set to trigger the increase in PS when needed (I used Mv to determine this). I initially set my PS max to be 1 more than PS min and increased from there as necessary. If you do need (or want to try) iVAPS mode, perhaps this will work for you too.
I hope this information and these suggestions help. I am mindful that I am working from my personal experiences, which will be different to yours because our underlying conditions are not the same. But certainly, a PS range from 4 to 12 would be the stuff of nightmares to me, so I am not surprised that you are finding iVAPS difficult as you have it set.