Viatom/Wellue versus CMS50F Oximeters - A Comparison
I have just got a new CMS50F because it was going super cheap, and I've run it alongside the Viatom Visual Oxy (which is pretty much the same as the Viatom/Wellvue CheckMe O2 or SleepU models, the wrist-worn ones with the soft ring sensor). I hope this isn't too in-depth or unwanted, feel free to ignore it if so.
 
TLDR: My preference overall is for the Viatom/Wellview. but there are several reasons that might make you prefer the CMS50F.  

The long story... 

Cost: CMS50F is way cheaper. So if cost is an issue, and you definitely want a wrist-based product, this is probably the one.

Sensor: The Viatom ring sensor is by FAR superior IMO. It stretches significantly, it has remained intact in nearly 2 years of use off and on, and it doesn't tend to lose contact with the finger. By contrast, although the traditional duck-bill sensor with the CMS50F is a bit more reliable than the fingertip ones, I STILL had to tape it in place to stop it from falling off, and even then, it would shift around enough to lose up to 20 seconds of data at a time a few times per night.

They also have different behaviours if they do slip off your finger. The CMS50F just keeps recording until you stop it (or turn off the device) whereas the Viatom prompts you to reinsert your finger for 10 seconds (allowing you to swap fingers, for instance) before stopping the recording and powering off. You can also press the power button to skip the countdown and just save it or hold down the power to turn it off, which will also autosave it. 

Recording time: Depends. Viatom has more storage space overall (40 hours in 10 hr slots) but Contec products can record up to 24 hours in one session. My biggest (and only real) pet peeve is that the Visual Oxy that I use will stop recording after 10 hours, rather than continue recording in the next slot. I believe that Crimson said to me that one of the models will continue recording into a second session if you sleep longer than that, but I'm not exactly sure which model(s) have that function.

I'm not sure how many individual sessions that the CMS50F can save (the default is 2, I think), but Viatom is fixed to four, no matter how short. So even if you have only recorded 5 minutes four times, rather than being able to use the rest of the time in a fifth slot, it will record over the oldest slot instead. I haven't done enough testing of the CMS50F to know for sure how it works. With the Viatom, though, you can of course stop the recording and restart again if you wake after a few hours and feel like you're going to sleep longer.

Comfort: Viatom. The device is smaller, lighter and more comfortable to wear. The Viatom is around 0.9-1.1oz (25-31g) whereas the CMS50F is around 1.8oz (50g). The strap is softer (and the device can be removed from it for easy cleaning) compared to the chunkier screen and traditional watch strap of the CMS50F. As above, the sensors are more comfortable, for me, for the Viatom, too.

The only minor issue is that where the Visual Oxy (not necessarily other models) slots into the strap, there's a small metal button that helps hold it in place. I'm allergic to nickel and this irritated my skin, especially being under plastic, but that was easily fixed with some clear nail polish and/or a piece of adhesive medical tape over the back of the strap where the metal sits. And by contrast the CMS50F has a metal buckle and the back of the main unit, so even more of an area for potential sensitivity.

If you also have a strong preference for which wrist to use, be aware that if you want the screen to be the right way up and the sensor port closest to your fingers, you would need to wear the Viatom on your left wrist and the CMS50F on the right. Both sensors should be long enough to be worn both ways, but if you have long hands, you may not find that possible.

Data captured: Viatom because it also includes movement sensors as well as pulse and SpO2. CMS50E has only pulse and SpO2. They both cover similar ranges of SpO2 and pulse, with similar levels of accuracy depending on environmental factors, etc. 

Battery Life: CMS50F gets 20 hours, apparently, and the Viatom wrist products get 16 hours (apart from the CheckMe O2 Max, which has a 72-hour battery). In practice though, the Viatom has always allowed me to get two days worth of recordings on one charge, and it only takes an hour or so to top up.

Software: Viatom. Although the Contec software comes with some better reports, the graphics are kind of ugly, and it can be a nightmare getting the right drivers and software version on certain systems. I think Contec makes software only for PC, but as far as I can tell, it has no apps for your smartphones. Viatom makes software for the PC and Macs, plus it does have apps for both Android and iOS, and the data can be imported into OSCAR from either the computer or the smartphone. The smartphone also allows you to monitor and record your data directly to the phone via Bluetooth,  although I am not sure if that data is also possible to import into OSCAR, or just the overnight sessions. However, you can't monitor your sats with the CMS50F live because of the lack of Bluetooth, and the finger probe using the USB port on the device. If you wanted to monitor or record the data live into OSCAR, you would need to use either a Bluetooth enabled Contec product or a finger clip device like the CMS50E while it's plugged in via the USB cable. 

Also - one other issue currently with OSCAR is that the newer (v2.2 firmware, I think) CMS50F devices can't be accessed by OSCAR to extract the data directly. You would need to save the data using the Contec software and then upload that, instead, which is the same kind of process you have to use with the Viatom data files. Either way, it still works, it's just an extra step for now, at least.

The main problem with the Viatom software is that you can't tweak the graphs or change the zoom - it shows you a zoomed-out and a zoomed-in view of the graph, but it's fixed. However, OSCAR gives you that functionality, so it's not a huge issue. There is a bug with one measure (if you have over 250 events, it miscalculates the ODI) but that will be fixed in the next version.

Date/Time Accuracy: CMS50F but only because of the current OSCAR interface. Both products store the date and time, and can be synchronised through the software (or simply by connecting via Bluetooth to your phone for the Viatom). When it comes to synchronising your pulse ox data to your CPAP data, it's easier for the CMS50F because the import wizard in OSCAR allows you to manually set the start time to match. With Viatom, you have to currently edit the file as the start time is hardcoded in the file, and you need to be comfortable using a hex editor to do that. It's not technically hard when you know how, but it's probably a challenge for many people.

Hopefully, in future, the Viatom import tool will allow you to manually correct it. As long as your devices are all set to the same time, though, they should line up close enough, and you can also tweak the time shift for the CPAP data as well, if necessary.

Accuracy: Either. I recorded myself one night using both products, just to find this out. Overall, they were pretty much the same, as far as I could tell. There were a few glaring differences - when my finger slipped out of the CMS50F, it produced all kinds of spurious data at the extreme ends of the spectrum that was clearly inaccurate, while the Viatom seemed to be a little better at that. I did see that there were some differences where my oxygen levels were spiking and plummeting precipitously, probably to do with the frequency of data polling, but I don't know the exact figures for either product. That said, the overall accuracy is about the same. Just bear in mind that the lower your SpO2 gets, the less reliable the measurements become.

Ease of use: Viatom. Out of the box, both of them record automatically as soon as you switch it on and put your finger in. A quick press of the power button on both swaps between the time/date and showing your current pulse and oxygen levels. The CMS50F screen is bigger, has a variety of brightness options, and may be easier to read for some. The Viatom is smaller and simpler, without the colours and extra data and may be preferred for the simpler, cleaner interface. 

However, the real ease of use comes in getting it to work with your computer, extracting the data, etc. Viatom has one software package for their pulse oximeters and needs no special drivers. the Contec products need specific driver versions, depending on the firmware and product, and different software and versions, same thing. On top of that, while the Viatom is simply plugged into the USB port on your PC or automatically connects through Bluetooth to your phone/tablet, there is often an issue with the correct order of plugging in the cable, the device, starting the software, rebooting your computer etc that can be endlessly frustrating for the Contec products. And as I have ALSO discovered, if you have multiple devices that use different drivers, you are a little stuffed. My old CMS50E won't communicate with the computer with the new drivers needed for the CMS50F, so bear that in mind if you have multiple devices. They also use different versions of the software, (so I currently have three separate programs from two manufacturers to get data off three different pulse oximeters. LOL) I also had to play hunt-the-data-files with the Contec products because of an issue with where the Contec software tries to save the files in Windows 10.

Alarms/Alerts: It depends. Audible alerts? CMS50F for sure. You can set the max and min thresholds for both pulse (between 0 and 254bpm) and oxygen saturation (0-100%), plus there is an audible finger-out alarm. However, it's an all or nothing. Technically you could set the boundaries to very high/low figures that you're unlikely to hit if you only want the finger-out alarm, though. It also comes with a pulse beep option, which can be set separately from the alarms. There are 4 volume settings. All settings (aside from a couple) can be set manually via the device menu. 

All the Viatom watch-type devices come with vibration alarms triggered by low oxygen levels (80-95%) with the intention of waking you more gently. The intensity can be varied between 1 and 5. and turned off completely. These values must be changed through the software. The CheckMe O2 Max comes with audible alarms too, I believe. 

Accessories/replacements: Both. You can get replacement finger probes and USB cables for both products, though maybe a little easier for Contec. The prices seem similar in my area but YMMV. You MUST use the USB that comes with the Viatom to access the data on your computer and for charging and though I've seen some people say you don't with the Contec products, I've never got it to work unless I use the one that came with it, either. 

Reliability: Viatom. As mentioned above, the CMS50E that I had, the display became unreadable after a relatively short period of time. I now have to use my smartphone in a dark room in order to read the screen, even with it set to the maximum brightness on the device. It still works as well as it previously did (ignoring the software and driver issues) but that is a huge problem for me.

The Viatom has been much better. I was worried the ring sensor would snap really easily but touch wood, so far it hasn't. I did manage to break part of the plastic strap, but this did glue back and Viatom replaced the whole unit for me, as they don't sell replacement straps alone for my specific product, the Visual Oxy. That's really impressive considering the unit was out of warranty, but it also costs £180 or so brand new now and it was significantly cheaper when I purchased it (about half the price 2 years ago). 

I did find that, with the Viatom cable, the two plastic parts that house the custom circuit board on the wider end of the USB cable tend to separate easily, and as the PCB is attached to the cable with three tiny solder joints, these can easily break if you aren't careful or are rough with it. For protection, I've wrapped a little tape around the plastic just to keep the two sides from working loose, but just be sure to never pull on the cable itself to unplug it from the computer.

Customer Service: Viatom.  While there are definitely some minor issues with language with both companies, Viatom's staff have proven to be far more helpful than I ever expected. They provided me with a tool to convert the data from an older software product to allow it to be viewed in their new one, there was the above-mentioned help with the strap (and they actually apologised that they couldn't give me either just a strap separately, or a brand new product, only a refurbished one - while letting me keep the existing one too). They clarified some of the details about how their product works and calculates the statistics and they've always been endlessly polite. Sometimes the response time has been a little slow, but they've always resolved my issues for me.

I found that Contec support was a lot less helpful. They don't seem to be much use at figuring out the program version/driver version/etc issues, at least in my experience. I've had two CMS50Es, both of which has issues with connecting to the computer to download the data (which defeats the whole purpose). The first one ended up dying after a year - the screen was just black, it would stay on permanently until the battery ran out but didn't seem to be recording or measuring any data. The second one wouldn't connect at all to my laptop when it was updated to windows 8 and I simply used it for live on-the-spot measurements rather than recording overnight. Although it was barely a year old, Contec refused to exchange it, even after such a fundamental failure in the first one, and never managed to get the software working on my laptop after I upgraded, so I moved onto Viatom instead. 

Overall conclusion: Viatom, is (IMO) the better device overall, especially if you want a slimmer, lighter product, up to 40 hours of data in four sessions, a more comfortable and reliable finger sensor, Bluetooth, smartphone apps and/or Mac software, no fussing around with drivers or software versions and more reliability long term. 

However, if you want to record up to a maximum of 24 hours total in one or more sessions, budget is important, you want/need audible warning alarms or a bigger screen, you're only using a PC and you're not worried about capturing/monitoring data in real-time, the CMS50F may well do everything that you need.
Post Reply Post Reply
#2
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
After a little more testing, I have a couple of other points to add. 

Low Battery and Charging:  It depends on what you require. Both have an icon on the screen that shows the approximate charge of the battery. When the battery is low on the CMS50F, it will sound an alarm continuously, as long as you have the alarms turned on. If they're turned off, it will remain silent until it switches off. The Viatom doesn't have an alarm, so it will simply turn off (saving the data in the process). I have left both Viatom and Contec devices unused for long stretches of time (months at least, if not a year) and both have retained most of the charge during that time. It is recommended to charge them again if they have been left for a while without use (or to periodically charge them between uses).

When it comes to fully charging from completely empty, the Viatom takes around two hours, whereas the CMS50F takes longer, somewhere around three-plus hours, in my case. If you are leaving either one plugged into the USB for a short time every day or longer every other day, though, you should be fine for regular overnight use. If you store the devices, leads etc in the box, it's very easy for the CMS50F to push up against the top which can easily drain the battery - even the first time I opened the box from new it was already turned on. While charging, the CMS50F has a very bright LED next to the USB port - this will shine red when charging and green when fully charged. The Viatom shows an animated icon of the battery charging and a fully charged battery when finished.
Post Reply Post Reply
#3
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
Excellent comparison.
I have a Viatom ring sensor but it is the one with the yellow band around it (Model Wear O2 4843).
The problem with this thing is that It does not export data and the only way you can see your data is through their own mobile app.
I wish I had spent the few extra dollars and gone with the Ring sensor that you tested.
Post Reply Post Reply
#4
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
S. Manz - After reading your post above, I tried searching for your oximeter model without success. I suggest that you create a review of your oximeter and its software. This may help others if they happen to come across your model while searching for an oximeter.
Crimson Nape
Apnea Board Moderator
Project Manager for OSCAR - Open Source CPAP Analysis Reporter
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#5
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
Thanks. Yes, its sold by GetWellue.com and is sometimes marketed as the "FL320" ring oximeter. It is however made by Viatom but the stamped letters on my unit are wornout so I can't make out details. I wouldn't recommend it as it has no export capability and it would be a waste of money if it is to be used with Oscar. That being said, it is very convenient to wear overnight and can record up to 16 hours of data but only at 1min intervals.
I have attached an image below
.
[attachment=36004]
Post Reply Post Reply
#6
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
(05-06-2021, 02:32 PM)Ratchick Wrote: Date/Time Accuracy: CMS50F but only because of the current OSCAR interface. Both products store the date and time, and can be synchronised through the software (or simply by connecting via Bluetooth to your phone for the Viatom). When it comes to synchronising your pulse ox data to your CPAP data, it's easier for the CMS50F because the import wizard in OSCAR allows you to manually set the start time to match. With Viatom, you have to currently edit the file as the start time is hardcoded in the file, and you need to be comfortable using a hex editor to do that. It's not technically hard when you know how, but it's probably a challenge for many people.

Hopefully, in future, the Viatom import tool will allow you to manually correct it. As long as your devices are all set to the same time, though, they should line up close enough, and you can also tweak the time shift for the CPAP data as well, if necessary.

I recently bought the Wellue/Viatom 02 Ring myself, and have been using my phone as a standard to sync the clocks on the O2ring and the AirSense 10 Autoset I've been using each night before bed.  As you mentioned I use the ViHealth app to sync the ring to the phone by connecting to it via bluetooth, and I use a clock app with a second hand to sync the AS10.  (which seems to loose about 7 seconds per day)  The AS10 time setting (which is on the clinical menu for some odd reason) lets you set the minute first, and then the hour, then starts the clock from when you finish setting the hour.  So I start by setting the minute ahead by 1, press the button, and let it sit on the hour set mode until the second hand reaches 12 at which point I press the button on the AS10 again to start the clock.  I then wait a minute to confirm that the clocks on the Phone and the AS10 tick over at approximately the same time. This seems to work pretty well.  

The other day my clinician called in the morning to change our appointment time, and actually woke me up, at which point I removed my mask, (having my sleep clinician inadvertently wake me up struck me as kind of ironic but that's neither here nor there Too-funny ) but I kept the O2 Ring on until shortly after our short conversation.  This was reflected in my data by an end to the CPAP waveforms and a corresponding rise in my pulse rate until the point when I took off the ring shortly afterwards. That correspondence has given me some confidence in the level of synchronisation achieved.

- Neelix
Post Reply Post Reply
#7
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
I'm looking at a product with the description "Wellue Oxylink Wireless Bluetooth Pulse Oximeter, Oxygen Saturation Monitor with Audio Reminder on Device & Smart phone Free APP, Finger Wearable Pulse Rate Tracker" on newegg. Is this the one that can't export data to OSCAR or not?

Also in the product description it says
"The report records the data every 1 min, cannot show oxygen changes within a few seconds, therefore it can't be used for sleep breath problems."
should I be looking elsewhere?
Post Reply Post Reply
#8
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
Keep looking.   As far as I know, only the O2Ring, SleepU, and the Checkme are OSCAR compatible.

Since this is the Product Review area, please let me know if you need to post additional information and I will split these posts out into their own thread.
Crimson Nape
Apnea Board Moderator
Project Manager for OSCAR - Open Source CPAP Analysis Reporter
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#9
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
If you can get the data using the ViHealth app on your phone, via bluetooth, then you can (theoretically) import it to OSCAR by copying the file to your computer, or hopefully should be able to. It's easier from the PC/Mac software, but you definitely can get it just from your phone. It's worth noting that my Visual Oxy is recognised by Viatom's PC software as being a CheckMe O2 (and having used a CheckMe O2 and found that only, really, the band was different from the Visual Oxy model), I suspect that many similar models are being created that all have the same underlying data storage.

I have now got myself a CheckMe O2 Max, which seems to be their top of the line pulse oximeter. And whoo boy, it's nice.

1) the battery life is 72 hours (roughly).
2) it has both vibration and/or audio alerts for both high and low pulse, and low SpO2.
3) It's a little bigger and heavier than the VisualOxy Max etc, but it's still lighter than the CMS50-F.
4) It can still only be worn on the left wrist if you want to read the screen the right way up, but works just as well on the right one if you just want to record the data and you can read upside down.
5) The screen is bigger. It also displays the 4-digit code you need to link to it within the cellphone app on the screen so it's easy to see which one is which.
6) I find that the icons on the screen are easier to recognise and understand just because of the increased size.
7) When you have it plugged into the USB port and open up O2 Insight Pro, this model is better at actually displaying when it has connected and also properly showing that it has simply gone back to charging after the program is closed. The VisualOxy doesn't do that, so to be sure it's going back to charging, I usually remove and replace the connector.
8) Unlike the CheckMe O2 (the non-Max version) and the Visual Oxy models, this one DOES continue recording immediately at the end of a ten-hour session, starting another within a couple of seconds. MAJOR bonus points for this. No more having to wake up to check it's still running!
9) Although the Visual Oxy has two modes (the nighttime recording mode by default, and the monitoring mode where the files can only be seen on the phone app, and which record data like steps taken, average O2 etc but not as many statistics given on the report as with the night recording) the CheckMe O2 Max only has the one - but honestly that doesn't really matter, the only thing it doesn't do is count steps, and honestly I don't use it for that anyway.
10) It will start recording without being prompted, if it is still on and you insert your finger into the probe, even if you had finished the previous recording.
11) You can also set alerts in the phone app for finger out as well as both pulse and SpO2 triggers, separate from the alert from the device itself.
12) the strap for both the CheckMe O2 and CheckMe O2 Max is much sturdier than the one for the Visual Oxy. Some people have said it's a little awkward with arthritis but I don't find it too hard now I have the knack. it's easier if you can get your thumb under the strap and your fingers on top to press the studs through the band.
13) Some of the other Viatom/Wellue products "fake" 2 second data sampling by simply doubling the 4 second data points in come versions of the app or software (I can't remember which one off the top of my head, I suspect Mac or iPhone versions though). This, however, ACTUALLY records the data every two seconds, so this has twice the granularity of data than the other Viatom/Wellue products, that I know about.
14) Both the VisualOxy and the CheckMe O2 Max can connect to the app via Bluetooth for real-time monitoring and alerts. CheckMe O2 Maz shows the PI on the phone app, but not within the data file that I can find, whereas the Visual Oxy doesn't show the PI at all.
15) When you turn on the CheckMe O2 Max, it also warns you with a flashing crossed-out bell icon for a few seconds when all warnings are turned off before it starts recording. However, I've discovered that you can actually change the settings via the phone app while you're recording (such as if it vibrates or beeps, or how loud/how strong it is etc, or even changing the threshold levels). Probably not something that most people would do on the regular, but useful if you started the session and it keeps waking you up too many times or whatnot.

I will probably write up a full review of this one - it's not cheap but I'm loving it, so I hope this helps on top of the previous comparison.
Post Reply Post Reply
#10
RE: Viatom/Wellue versus CMS50F Oximeters - A Comparison
(01-01-2022, 05:41 AM)Ratchick Wrote: 13) Some of the other Viatom/Wellue products "fake" 2 second data sampling by simply doubling the 4 second data points in come versions of the app or software (I can't remember which one off the top of my head, I suspect Mac or iPhone versions though). This, however, ACTUALLY records the data every two seconds, so this has twice the granularity of data than the other Viatom/Wellue products, that I know about.

It's kinda funny that their web site says:

Quote:Tracks your blood oxygen saturation level, heart rate, and body movements every second automatically and accurately.


...but then doesn't highlight the actual resolution later:

Quote:Print & share two-seconds interval, high-resolution PDF/CVS format oximetry reports with your doctor. 

One other question: can you finally adjust the brightness of the display? I had to stick black electrical tape over my SleepU!
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
Angry [Product Review] Wellue 02 Ring Customer Support Review r0nwilliams 3 986 11-11-2024, 11:37 AM
Last Post: notausername
  [Product Review] Wellue SleepU Wrist Pulse Oximeter Crimson Nape 23 9,594 08-19-2024, 11:09 AM
Last Post: somewherein413
  [Product Review] SleepHQ versus Oscar CPAP software Sperafucile 4 6,740 09-01-2023, 09:56 PM
Last Post: Lucky7
  [Product Review] Wellue Recording ECG with AI Analysis Crimson Nape 3 3,615 01-15-2022, 10:54 AM
Last Post: Crimson Nape
Thumbsup [Product Review] CMS50F Pulse Oximeter (USB, no bluetooth) Peter_C 21 18,719 09-05-2021, 10:13 PM
Last Post: CorruptAlligator
Smile [Product Review] Wellue O2Ring initial review CZOscar 14 14,964 08-27-2021, 02:40 PM