Machine: ResMed AirCurve 10 ASV Mask Type: Full face mask Mask Make & Model: ResMed AirTouch F20 full mask Humidifier: ResMed AirCurve 10 ASV CPAP Pressure: EPAP: 5.0-7.0 PS: 2.0-9.0 CPAP Software: OSCAR
ResMed AirCurve 10 ASV settings for first night of treatment: Do you agree?
Thanks to the suggestions of the moderators here, I just obtained a ResMed AirCurve 10 ASV device to replace my AirSense 11 for almost all clear-airway events and periodic breathing pattern.
I believe moderators here suggested that I start ASV with the ASVAuto default settings and later upload OSCAR screenshots to see if I need manual adjustments.
The ResMed titration guide suggests the following ASVAuto settings:
Min EPAP 4 cm H2O
Max EPAP 15 cm H2O
Min PS 3 cm H2O
Max PS 15 cm H2O
Ramp OFF
Do you agree with these default settings when I start my first night with ASV? Thanks!
Machine: ResMed AirCurve 10 ASV Mask Type: Full face mask Mask Make & Model: ResMed AirTouch F20 full mask Humidifier: ResMed AirCurve 10 ASV CPAP Pressure: EPAP: 5.0-7.0 PS: 2.0-9.0 CPAP Software: OSCAR
Machine: ResMed AirCurve 10 ASV Mask Type: Full face mask Mask Make & Model: ResMed AirTouch F20 full mask Humidifier: ResMed AirCurve 10 ASV CPAP Pressure: EPAP: 5.0-7.0 PS: 2.0-9.0 CPAP Software: OSCAR
RE: ResMed AirCurve 10 ASV settings for first night of treatment: Do you agree?
The settings you outlined are exactly what Resmed recommends in the titration protocol. We can look at the resulting data from your first night and suggest optimization if needed.
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Machine: ResMed AirCurve 10 ASV Mask Type: Full face mask Mask Make & Model: ResMed AirTouch F20 full mask Humidifier: ResMed AirCurve 10 ASV CPAP Pressure: EPAP: 5.0-7.0 PS: 2.0-9.0 CPAP Software: OSCAR
Machine: ResMed AirCurve 10 ASV Mask Type: Full face mask Mask Make & Model: ResMed AirTouch F20 full mask Humidifier: ResMed AirCurve 10 ASV CPAP Pressure: EPAP: 5.0-7.0 PS: 2.0-9.0 CPAP Software: OSCAR
Machine: ResMed AirCurve 10 ASV Mask Type: Full face mask Mask Make & Model: ResMed AirTouch F20 full mask Humidifier: ResMed AirCurve 10 ASV CPAP Pressure: EPAP: 5.0-7.0 PS: 2.0-9.0 CPAP Software: OSCAR
04-18-2023, 04:01 PM (This post was last modified: 04-18-2023, 04:15 PM by DevinJones.)
RE: ResMed AirCurve 10 ASV settings for first night of treatment: Do you agree?
Here are the results for my first two nights with my new ResMed AirCurve 10 ASV after I previously experienced many central sleep apnea events and periodic breathing pattern with a ResMed AirSense 11 APAP:
--------
Night 1 on April 15, 2023, with no CSA, no OSA, no periodic breathing, but with 36 flagged hypopneas, five UAs, frequent (97) large leaks, air-filled cheeks, and some aerophagia while I was using a DreamWear nasal mask and mouth tape:
Device Settings
Mode ASVAuto
Min EPAP 5.00 cmH2O
Max EPAP 15.00 cmH2O
PS Min 3.00 cmH2O
PS Max 15.00 cmH2O
Antibacterial Filter Yes
Climate Control Auto
Essentials On
Humidifier Status On
Humidity Level 4
Mask Nasal
Ramp Off
Smart Start Off
Temperature 27 ºC
Temperature Enable On
AHI 5.26
ResMed AirCurve 10 ASV
PAP Mode: ASVAuto
Date 4/15/2023
Start 02:04:35
End 09:52:36
Hours 07:48:01
Night 2 on April 18, 2023, with no CSA, no OSA, no periodic breathing, but with 28 flagged hypopneas, 31 UAs, only one large leak, less of a problem with air-filled cheeks and aerophagia while I was using a DreamWear full-face mask:
Device settings (different from the settings on night 1) and the data for night 2 are in the attached OSCAR screenshots.
Thank you very much, Sleeprider and others, for your analysis and suggestions for adjustments to device settings.