[Equipment] transferring settings from PR unit to Resmed unit
#1
transferring settings from PR unit to Resmed unit
The Respironics System One cpap unit that I've been using for a long time (11,700 hours on the blower) is on the recall list. I am switching over to a used Resmed Airsense 10 Autoset. With the old setup I've had good results, using a nasal mask and taping my mouth, with average AHI under 1.0

In clinical settings on the new unit I entered the same auto pressure settings (ipap 16, epap 8) that I had on the old unit. But on my first night using the Airsense, my checks were puffing out and sleep was disrupted. I am wondering why the same pressure settings are not translating well between the two units. Should I be lowering the ipap and/or epap pressure settings on the new unit?
Post Reply Post Reply
#2
RE: transferring settings from PR unit to Resmed unit
You probably had the flex on the old machine. ResMed does not call it the same thing. It is the EPR. Set your EPR to 3 and you should be about the same as your old one.
Apnea (80-100%) 10 seconds, Hypopnea (50-80%) 10 seconds, Flow Limits (0-50%) not timed  Cervical Collar - Dealing w DME - Chart Organizing
Post Reply Post Reply
#3
RE: transferring settings from PR unit to Resmed unit
Please post a screenshot of your night from OSCAR. I want to be sure which Model you have.
You said AutoSet but the pressure settings are those of a BiLevel with PS=8 which is likely too high.
Post Reply Post Reply
#4
RE: transferring settings from PR unit to Resmed unit
Thanks for that clarification. Pressures of 8-16 are the autoset lower and upper limits, not bipap pressure settings. Screenshot attached. Any insights most appreciated.


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#5
RE: transferring settings from PR unit to Resmed unit
Flow Limits are driving your pressure up, we will limit your high but for now
set Min Pressure = 7
Set EPR = 3, this is a therapeutic value and flex, in this role, has no comparison.
These settings should help the flow limits
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  [Treatment] BIPAP Settings - ResMed AirCurve 11 Vauto dcbauman 5 170 Yesterday, 11:42 AM
Last Post: dcbauman
  New ASV unit. No Doctor help! Barry5k 12 477 12-06-2024, 06:59 PM
Last Post: RayBee
  Help with ASVAuto ResMed AirCurve 10 settings (and esophageal pain) RowG 2 230 11-17-2024, 08:34 AM
Last Post: RowG
  Help with Resmed 11 Settings underoathtdwp 2 336 10-23-2024, 11:58 PM
Last Post: underoathtdwp
  Mold issue / cleaning CPAP (the unit, not tubing), Resmed S9 NightOwl42 11 2,236 09-13-2024, 01:44 PM
Last Post: NightOwl42
  [Equipment] Transition from old ResMed11 to new unit alfarabee 2 247 08-09-2024, 02:09 AM
Last Post: Rickyricardo
  ResMed mask settings - what changes in each type? skypony 14 27,185 08-01-2024, 10:26 AM
Last Post: DaveL


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.