[Equipment] Please need some input regarding new PR auto bipap machine.
#21
RE: Please need some input regarding new PR auto bipap machine.
We see a drop in VS, H and RERA, but a small increase in FL that I can't explain because we should have targeted that. Periodic breathing and large leak are also reduced and a minor increase in CA. By the numbers, it is fine. How did this feel in comparison to lower PS?
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#22
RE: Please need some input regarding new PR auto bipap machine.
(08-23-2020, 10:28 AM)Sleeprider Wrote: We see a drop in VS, H and RERA, but a small increase in FL that I can't explain because we should have targeted that.  Periodic breathing and large leak are also reduced and a minor increase in CA.  By the numbers, it is fine. How did this feel in comparison to lower PS?

I seem to sleep better with the lower PS (the pressure at this range of 4.5 I'm not used to yet).
Post Reply Post Reply
#23
RE: Please need some input regarding new PR auto bipap machine.
Hello again. I'm not sleeping as soundly with the bipap, mainly because the pressure on inhale is too blunt/strong. What pressure setting do I need to lower to make the inhale pressure less? It doesn't seem to taper, it's either air/no air.

Thank You
Post Reply Post Reply
#24
RE: Please need some input regarding new PR auto bipap machine.
We can drop back pressure support, also, it is possible to taper the rise time from EPAP to IPAP by changing BiFlex settings or turning off BiFlex and using a rise time setting that is longer than 0.3 seconds. Once again, this is an advantage of Resmed where the spontaneous breathing effort causes the rise in pressure support, which is then proportional to the respiratory flow. Let's start by understanding what BiFlex setting you are currently using. If you want to calibrate rise time, we can turn off BiFlex and use a setting of 0.5 or 0.6. Higher settings extend the time it takes for pressure support to be fully on. This is a trial and error setting, and you just need to try it while relaxing and breathing with the machine. Find the setting that feels most natural from 0.3 up to 1.0.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#25
RE: Please need some input regarding new PR auto bipap machine.
(08-25-2020, 11:40 AM)Sleeprider Wrote: We can drop back pressure support, also, it is possible to taper the rise time from EPAP to IPAP by changing BiFlex settings or turning off BiFlex and using a rise time setting that is longer than 0.3 seconds.   Once again, this is an advantage of Resmed where the spontaneous breathing effort causes the rise in pressure support, which is then proportional to the respiratory flow.  Let's start by understanding what BiFlex setting you are currently using.  If you want to calibrate rise time, we can turn off BiFlex and use a setting of 0.5 or 0.6. Higher settings extend the time it takes for pressure support to be fully on.  This is a trial and error setting, and you just need to try it while relaxing and breathing with the machine. Find the setting that feels most natural from 0.3 up to 1.0.

That's exactly what I needed to know, I will experiment with rise time settings and let you know how it goes.  Thank you very much!
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  need help with a new BIPAP Hanma 2 174 12-09-2024, 07:29 AM
Last Post: Hanma
  Airsense 11 Auto - What is auto-off supposed to do? Schnoozer 3 315 10-31-2024, 07:33 PM
Last Post: Warnick13
Exclaimation CPAP Machine Choices - read this before you accept a new machine SuperSleeper 305 285,010 10-28-2024, 07:30 PM
Last Post: ZacharySimon
Information Recommendation for new Bipap machine HoodedRobin 4 273 10-16-2024, 07:13 PM
Last Post: Deborah K.
  Just getting startedand need input gr3gg0 1 255 10-15-2024, 04:42 PM
Last Post: Old Steve
  [Treatment] Issues with auto bipap and aero haha Xboxrrod 22 1,292 10-12-2024, 06:10 PM