I was looking at the settings and they were the same in terms of max and min etc and was scratching my head.
Then I dug into the 2 clinicians manuals
It seems Resmed counts OA as a reduction in flow of more than 90% for >10 seconds and Hypopneas as a greater than 50% reduction for >10 seconds.
Intellipap 2 does them at 75% and 40% respectively hence it is set to be more sensitive. The Intelliaps 2 is also adjustable so I can change the settings to match.
Mystery solved, I think?
Experts what do you think about my logic??