REMstar error codes
From Apnea Board Wiki
Here are the Error Codes for REMstar Machines:
E-0 - E-20
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-0 | NO_ERROR | Reserved | N/A |
E-1 | ERR_SOFTWARE | Reserved | Replace Main PCA |
E-2 | ERR_PROGRAM_CRC_FAILURE | Defective FLASH chip | Replace Main PCA |
E-3 | ERR_EXTERNAL_RAM_FAILURE | Defective RAM chip | Replace Main PCA |
E-4 | ERR_FIQ_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-5 | ERR_FIQ_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-6 | ERR_NEST_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-7 | ERR_NEST_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-8 | ERR_IRQ_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-9 | ERR_IRQ_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-10 | ERR_TMR_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-11 | ERR_TMR_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-12 | ERR_SVC_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-13 | ERR_SVC_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-14 | ERR_THREAD_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-15 | ERR_ARM_UNDEFINED_INSTRUCTION_EXCEPTION | Defective FLASH chip | Replace Main PCA |
E-16 | ERR_ARM_SWI_EXCEPTION | Defective FLASH chip | Replace Main PCA |
E-17 | ERR_ARM_PREFETCH_EXCEPTION | Defective FLASH chip | Replace Main PCA |
E-18 | ERR_ARM_ABORT_EXCEPTION | Defective RAM chip | Replace Main PCA |
E-19 | ERR_ARM_RESERVED_EXCEPTION | Defective FLASH chip Defective microprocessor |
Replace Main PCA |
E-20 | ERR_AIC_DEFAULT_HANDLER | Defective FLASH chip | Replace Main PCA |
E-21 - E-40
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-21 | ERR_AIC_SPURIOUS_HANDLER | Defective FLASH or RAM chip | Replace Main PCA |
E-22 | ERR_CAL_DATA_CRC_FAILURE | Improper calibration Defective FLASH chip |
Recalibrate Replace Main PCA |
E-23 | ERR_INVALID_DRIFT_INDEX_AT_CAL | Improper calibration Occluded or restricted pressure tube Defective flow sensor |
Recalibrate Replace pressure tube Replace Main PCA |
E-24 | ERR_UNRECOGNIZED_CAL_VERSION_BY_SW | Improper calibration Defective FLASH chip |
Recalibrate Replace Main PCA |
E-25 | ERR_DRIFT_FLOW_TOO_HIGH | Occluded or restricted pressure tube Defective flow sensor |
Replace Flow Tube Replace Main PCA |
E-26 | ERR_UI_DISPLAY_FAILURE | Defective LCD driver chip | Replace Main PCA |
E-27 | ERR_DRIFT_PRESSURE_TOO_HIGH | Occluded or restricted pressure tube Defective pressure sensors |
Replace pressure tubing Replace Main PCA |
E-28 | ERR_CAL_DATA_BLANK | Improper calibration Defective FLASH chip |
Recalibrate Replace Main PCA |
E-29 | ERR_DRIFT_SECONDARY_PRESSURE_TOO_HIGH | Occluded or restricted pressure tube Defective pressure sensors |
Replace pressure tubing Replace Main PCA |
E-30 | ERR_INVALID_NVDATA_SCHEMA | Normal for first time power up Defective EEPROM |
Remove & restore power Replace Main PCA |
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-31 | ERR_INVALID_NVDATA_LOCK | Defective EEPROM | Replace Main PCA |
E-32 | ERR_INVALID_NVDATA_CHECKSUM | Normal for first time power up Defective EEPROM |
Remove & restore power Replace Main PCA |
E-33 | ERR_NVDATA_STORAGE_ERROR | Defective EEPROM | Replace Main PCA |
E-34 | ERR_NVDATA_QUEUE_FULL | Defective EEPROM | Replace Main PCA |
E-35 | ERR_NVDATA_READBACK_FAILURE | Defective EEPROM | Replace Main PCA |
E-36 | ERR_NVDATA_RANGE_FAILURE | Defective EEPROM | Replace Main PCA |
E-37 | ERR_CORRUPT_RTC_VALUE | Defective microprocessor | Replace Main PCA |
E-38 | ERR_RTC_NOT_TICKING | Defective microprocessor Defective crystal |
Replace Main PCA Replace Main PCA |
E-39 | ERR_UI_QUEUE_FULL | Defective PCA | Replace Main PCA |
E-40 | ERR_INVALID_BIST_TEST_CALL | Invalid RASP command | Reissue RASP command |
E-41 - E-60
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-41 | ERR_BIST_THREAD_LOCKED | Defective PCA | Replace Main PCA |
E-42 | ERR_ALARM_VOLTAGE_FAILURE | Defective power supply | Replace Main PCA |
E-43 | ERR_12VREF_FAILURE | Defective power supply | Replace Main PCA |
E-44 | ERR_5VREF_FAILURE | Defective power supply | Replace Main PCA |
E-45 | ERR_VBULK_FAILURE | Defective power supply | Replace Main PCA |
E-46 | ERR_NEG15VREF_FAILURE | Defective power supply | Replace Main PCA |
E-47 | ERR_FL12_SENSOR_RAILED | Occluded or restricted pressure tube Defective flow sensor |
Replace flow tube Replace Main PCA |
E-48 | ERR_PV2_SENSOR_RAILED | Occluded or restricted pressure tube Defective pressure sensor |
Replace pressure tube Replace Main PCA |
E-49 | ERR_FLASH_FAILURE | Defective FLASH chip | Replace Main PCA |
E-50 | ERR_UNRECOGNIZED_BOARD_REV | Needs new software Defective PCA |
Replace Main PCA Replace Main PCA |
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-51 | ERR_MOTOR_FAILURE | Defective motor drive chip | Replace Main PCA |
E-52 | ERR_BLOWER_RPM_FAILED_TOLERANCE | Stalled or locked blower | Replace Blower Replace Main PCA |
E-53 | ERR_BLOWER_OFF_CURRENT_HIGH | Stalled or locked motor Defective motor drive chip |
Replace Blower Replace Main PCA |
E-54 | ERR_BLOWER_ON_CURRENT_HIGH | Stalled or locked motor Defective motor drive chip |
Replace Blower Replace Main PCA |
E-55 | ERR_PRINTF_BUFFER_OVERFLOW | Defective PCA | Replace Main PCA |
E-56 | ERR_THERAPY_EVENT_QUEUE_FULL | Defective PCA | Replace Main PCA |
E-57 | ERR_KEYPAD_STUCK_KEY | Verify all keys protrude through top Defective Keypad Defective pad on PCA |
Align keys with top Replace Keypad Replace Main PCA |
E-58 | ERR_UNUSED_58 | Reserved | Reserved |
E-59 | ERR_UNUSED_59 | Reserved | Reserved |
E-60 | ERR_UNUSED_59 | Reserved | Reserved |
E-61 - E-80
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-61 | ERR_UNUSED_61 | Reserved | Reserved |
E-62 | ERR_UNUSED_62 | Reserved | Reserved |
E-63 | ERR_WATCHDOG_FAILURE | Defective microprocessor | Replace Main PCA |
E-64 | ERR_UNEXPECTED_WATCHDOG | Defective FLASH chip Defective RAM chip |
Replace Main PCA Replace Main PCA |
E-65 | ERR_UNUSED_65 | Reserved | Reserved |
E-66 | ERR_RTOS_INIT_FAILURE | Defective PCA | Replace Main PCA |
E-67 | ERR_SERIAL_QUEUE_FULL | Defective PCA | Replace Main PCA |
E-68 | ERR_SNORE_SAMPLING_THREAD_LOCKED | Defective PCA | Replace Main PCA |
E-69 | ERR_MAIN_EXECUTION_THREAD_LOCKED | Defective PCA | Replace Main PCA |
E-70 | ERR_INTERNAL_RAM_FAILURE | Defective microprocessor | Replace Main PCA |
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-71 | ERR_EVENT_LOG_QUEUE_FULL | Defective PCA | Replace Main PCA |
E-72 | ERR_TREND_DIV_BY_ZERO | Defective RAM chip | Replace Main PCA |
E-73 | ERR_ROLLOVER_CASE | Defective RAM chip | Replace Main PCA |
E-74 | ERR_WRONG_DISPLAY_TYPE | Defective PCA | Replace Main PCA |
E-75 | ERR_ALARM_CAPACITOR_FAILURE | Defective PCA | Replace Main PCA |
E-76 | ERR_OVER_PRESSURE | Occluded or restricted tube Defective PCA |
Replace Pressure tubing Replace Main PCA |
E-77 | ERR_UNDER_PRESSURE | Occluded or restricted tube Defective PCA |
Replace Pressure tubing Replace Main PCA |
E-78 | ERR_OVER_PRESSURE_REG | Occluded or restricted tube Defective PCA |
Replace Pressure tubing Replace Main PCA |
E-79 | ERR_UNDER_PRESSURE_REG | Occluded or restricted tube Defective PCA |
Replace Pressure tubing Replace Main PCA |
E-80 | ERR_ROTOR_LOCKED | Locked Motor Defective PCA |
Replace Motor Replace PCA |
E-80 - E-86
See Also
Smart card error codes
Donate to Apnea Board