370
edits
Changes
→What is the oxygen concentration in the diluted CPAP circuit?
=What is the oxygen concentration in the diluted CPAP circuit?=
(oxygen flow rate (litres) at 100% O2 + air volume (litres) at 21% O2) / Total air volume = oxygen concentration %
therefore, at 1 litres/minute oxygen addition and assuming a total CPAP flow of 60 30 L/min:(1-L/min x 100) + (6030-L/min x 21) = 1360 730/6031-L/min = 2223.55% Oxygen.
at 1 litre/minute oxygen
(3-L/min x 100) + (6030-L/min x 21) = 1560 930/ 60 33 L/min = 2628.18% oxygen.
Any variable can be changed based on available data. If the flow rate is 40 L/min from the CPAP, the oxygen concentrations will be higher. If leaks are present or pressure results in a higher average vent-rate, the oxygen concentrations will be lower.
[[file:Resmed Flow vs Pressure.png]]
=xPAP Flow Rates vs Pressure=
Donate to Apnea Board