Ti min 0.3 sec
Ti max 2.0 sec
Hello Guest,
Welcome to Apnea Board !
Breath pattern question
|
03-16-2022, 12:28 PM
RE: Breath pattern question
I believe the defaults are:
Ti min 0.3 sec Ti max 2.0 sec
03-16-2022, 12:35 PM
RE: Breath pattern question
03-16-2022, 12:42 PM
RE: Breath pattern question
The Ti min of 3.0 could be the culprit in all this. Definitely set that to 0.3.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-16-2022, 12:44 PM
(This post was last modified: 03-16-2022, 12:47 PM by SarcasticDave94.
Edit Reason: Edit
)
RE: Breath pattern question
Sorry to ask David because I've forgotten, but was that VAuto a used unit? If that's the case, best practice is to reset the device to default before any usage.
PS wouldn't Ti Min/Max at 3.0 force all inhales to be 3 seconds, regardless of spontaneous efforts otherwise?
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-16-2022, 12:50 PM
RE: Breath pattern question
You can perform a factory reset in the clinician menu. This will set both back to the base times.
If you have/had a Ti min set to 3 seconds, then that could very well be your problem. The Ti min setting determines how soon, or long, before the machine can drop the IPAP pressure. The Ti max setting controls the maximum time that the IPAP pressure will be maintained. I believe the default settings are, Ti min=.2 and the Ti max=2.8. For some reason, a Ti max setting of 3.1 feels more comfortable for most people. Leave the Ti min at the .2. The Ti min setting rarely needs to be changed. - Red
Crimson Nape
Apnea Board Moderator Project Manager for OSCAR - Open Source CPAP Analysis Reporter www.ApneaBoard.com ___________________________________ Useful Links -or- When All Else Fails: The Guide to Understanding OSCAR OSCAR Chart Organization Attaching Images and Files on Apnea Board Apnea Helpful Tips INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-16-2022, 12:52 PM
RE: Breath pattern question
(03-16-2022, 12:44 PM)SarcasticDave94 Wrote: Sorry to ask David because I've forgotten, but was that VAuto a used unit? If that's the case, best practice is to reset the device to default before any usage. Unless they can reset the odometer on the machine I don't believe it was a used one. It wasn't supposed to be a used one. I "think" it has 200 and something hours on it. The first time I used it was Jan 3rd and I've used it for ~8 hrs every night since.
03-16-2022, 02:10 PM
RE: Breath pattern question
I'm pretty sure this was operator error. Anyway, by changing TiMin back to 0.3 the anomaly will be gone. 100% sure.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-16-2022, 06:05 PM
RE: Breath pattern question
Yes, I hope it's mystery solved. I've got the settings as Ti Min 0.2 and Ti Max 3.1 so we'll see what OSCAR has to say tomorrow morning. Thank you all again.
03-17-2022, 06:58 AM
RE: Breath pattern question
No worries! Clearly, the 3.0 second I-time is readily apparent when plotted against the timeline:
|
« Next Oldest | Next Newest »
|
Possibly Related Threads... | |||||
Thread | Author | Replies | Views | Last Post | |
Hard to breath with CPAP and RSD(CRPS) | Biggz | 92 | 11,129 |
11-16-2024, 02:16 PM Last Post: Biggz |
|
New Weird non-breathing pattern showing up | HarleyGirl63 | 9 | 284 |
11-13-2024, 06:50 PM Last Post: HarleyGirl63 |
|
Odd breathing pattern? | terumi | 1 | 179 |
11-05-2024, 11:06 AM Last Post: PeaceLoveAndPizza |
|
Falling asleep holding breath | onetwothree | 8 | 428 |
10-28-2024, 12:33 AM Last Post: onetwothree |
|
Weird Flow Rate Pattern Started 3 days ago | Dodies | 2 | 228 |
09-30-2024, 11:42 PM Last Post: Dodies |
|
pattern in high resolution mask pressure in resmed device | rberry1 | 1 | 297 |
09-11-2024, 07:51 PM Last Post: Deborah K. |
|
[Equipment] Air Fit F20 Face Mask - Condensation Issue from Breath | xirokx | 21 | 915 |
08-25-2024, 08:50 AM Last Post: xirokx |