[CPAP] Service Required Error Codes - Inner Workings of a CPAP machine
#1
Service Required Error Codes - Inner Workings of a CPAP machine
[note: some parts of this thread were copied from our old forum]



dunhamg Wrote:Is there something that I can do to reset something, I am leaving for vacation first thing in the morning to Mexico for the week. Last night my Remstar Pro M with AFlex gave me this message. I checked the clinician menu and didn't see anything there.

Thanks for any help.
Grady
Post Reply Post Reply
#2
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
Hi Grady, welcome to Apnea Board!

To help diagnose the issue, it should give you an error code of some type...

Is there a specific error code or does it simply say "Service Required"?

Also, does your machine have a smart card? If so, try removing the card and seeing if that solves the problem.

Here are the Error Codes for REMstar Machines:


Error Code - Description - Probable Cause - Corrective Action

E-0 NO_ERROR •Reserved •N/A
E-1 ERR_SOFTWARE •Reserved •Replace Main PCA
E-2 ERR_PROGRAM_CRC_FAILURE •Defective FLASH chip •Replace Main PCA
E-3 ERR_EXTERNAL_RAM_FAILURE •Defective RAM chip •Replace Main PCA
E-4 ERR_FIQ_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-5 ERR_FIQ_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-6 ERR_NEST_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-7 ERR_NEST_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-8 ERR_IRQ_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-9 ERR_IRQ_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-10 ERR_TMR_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-11 ERR_TMR_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-12 ERR_SVC_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-13 ERR_SVC_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-14 ERR_THREAD_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-15 ERR_ARM_UNDEFINED_INSTRUCTION_EXCEPTION •Defective FLASH chip •Replace Main PCA
E-16 ERR_ARM_SWI_EXCEPTION •Defective FLASH chip •Replace Main PCA
E-17 ERR_ARM_PREFETCH_EXCEPTION •Defective FLASH chip •Replace Main PCA
E-18 ERR_ARM_ABORT_EXCEPTION •Defective RAM chip •Replace Main PCA
E-19 ERR_ARM_RESERVED_EXCEPTION •Defective FLASH chip •Defective microprocessor •Replace Main PCA •Replace Main PCA
E-20 ERR_AIC_DEFAULT_HANDLER •Defective FLASH chip •Replace Main PCA
E-21 ERR_AIC_SPURIOUS_HANDLER •Defective FLASH or RAM chip •Replace Main PCA
E-22 ERR_CAL_DATA_CRC_FAILURE •Improper calibration •Defective FLASH chip •Recalibrate •Replace Main PCA
E-23 ERR_INVALID_DRIFT_INDEX_AT_CAL •Improper calibration •Occluded or restricted pressure tube •Defective flow sensor •Recalibrate •Replace pressure tube •Replace Main PCA
E-24 ERR_UNRECOGNIZED_CAL_VERSION_BY_SW •Improper calibration •Defective FLASH chip •Recalibrate •Replace Main PCA
E-25 ERR_DRIFT_FLOW_TOO_HIGH •Occluded or restricted pressure tube •Defective flow sensor •Replace Flow Tube •Replace Main PCA
E-26 ERR_UI_DISPLAY_FAILURE •Defective LCD driver chip •Replace Main PCA
E-27 ERR_DRIFT_PRESSURE_TOO_HIGH •Occluded or restricted pressure tube •Defective pressure sensors •Replace pressure tubing •Replace Main PCA
E-28 ERR_CAL_DATA_BLANK •Improper calibration •Defective FLASH chip •Recalibrate •Replace Main PCA
E-29 ERR_DRIFT_SECONDARY_PRESSURE_TOO_HIGH •Occluded or restricted pressure tube •Defective pressure sensors •Replace pressure tubing •Replace Main PCA
E-30 ERR_INVALID_NVDATA_SCHEMA •Normal for first time power up •Defective EEPROM •Remove & restore power •Replace Main PCA
E-31 ERR_INVALID_NVDATA_LOCK •Defective EEPROM •Replace Main PCA
E-32 ERR_INVALID_NVDATA_CHECKSUM •Normal for first time power up •Defective EEPROM •Remove & restore power •Replace Main PCA
E-33 ERR_NVDATA_STORAGE_ERROR •Defective EEPROM •Replace Main PCA
E-34 ERR_NVDATA_QUEUE_FULL •Defective EEPROM •Replace Main PCA
E-35 ERR_NVDATA_READBACK_FAILURE •Defective EEPROM •Replace Main PCA
E-36 ERR_NVDATA_RANGE_FAILURE •Defective EEPROM •Replace Main PCA
E-37 ERR_CORRUPT_RTC_VALUE •Defective microprocessor •Replace Main PCA
E-38 ERR_RTC_NOT_TICKING •Defective microprocessor •Defective crystal •Replace Main PCA •Replace Main PCA
E-39 ERR_UI_QUEUE_FULL •Defective PCA •Replace Main PCA
E-40 ERR_INVALID_BIST_TEST_CALL •Invalid RASP command •Reissue RASP command
E-41 ERR_BIST_THREAD_LOCKED •Defective PCA •Replace Main PCA
E-42 ERR_ALARM_VOLTAGE_FAILURE •Defective power supply •Replace Main PCA
E-43 ERR_12VREF_FAILURE •Defective power supply •Replace Main PCA
E-44 ERR_5VREF_FAILURE •Defective power supply •Replace Main PCA
E-45 ERR_VBULK_FAILURE •Defective power supply •Replace Main PCA
E-46 ERR_NEG15VREF_FAILURE •Defective power supply •Replace Main PCA
E-47 ERR_FL12_SENSOR_RAILED •Occluded or restricted pressure tube •Defective flow sensor •Replace flow tube •Replace Main PCA
E-48 ERR_PV2_SENSOR_RAILED •Occluded or restricted pressure tube •Defective pressure sensor •Replace pressure tube •Replace Main PCA
E-49 ERR_FLASH_FAILURE •Defective FLASH chip •Replace Main PCA
E-50 ERR_UNRECOGNIZED_BOARD_REV •Needs new software •Defective PCA •Replace Main PCA •Replace Main PCA
E-51 ERR_MOTOR_FAILURE •Defective motor drive chip •Replace Main PCA
E-52 ERR_BLOWER_RPM_FAILED_TOLERANCE •Stalled or locked blower •Replace Blower •Replace Main PCA
E-53 ERR_BLOWER_OFF_CURRENT_HIGH •Stalled or locked motor •Defective motor drive chip •Replace Blower •Replace Main PCA
E-54 ERR_BLOWER_ON_CURRENT_HIGH •Stalled or locked motor •Defective motor drive chip •Replace Blower •Replace Main PCA
E-55 ERR_PRINTF_BUFFER_OVERFLOW •Defective PCA •Replace Main PCA
E-56 ERR_THERAPY_EVENT_QUEUE_FULL •Defective PCA •Replace Main PCA
E-57 ERR_KEYPAD_STUCK_KEY •Verify all keys protrude through top •Defective Keypad •Defective pad on PCA •Align keys with top •Replace Keypad •Replace Main PCA
E-58 ERR_UNUSED_58 •Reserved
E-59 ERR_UNUSED_59 •Reserved
E-60 ERR_UNUSED_60 •Reserved
E-61 ERR_UNUSED_61 •Reserved
E-62 ERR_UNUSED_62 •Reserved
E-63 ERR_WATCHDOG_FAILURE •Defective microprocessor •Replace Main PCA
E-64 ERR_UNEXPECTED_WATCHDOG •Defective FLASH chip •Defective RAM chip •Replace Main PCA •Replace Main PCA
E-65 ERR_UNUSED_65 •Reserved
E-66 ERR_RTOS_INIT_FAILURE •Defective PCA •Replace Main PCA
E-67 ERR_SERIAL_QUEUE_FULL •Defective PCA •Replace Main PCA
E-68 ERR_SNORE_SAMPLING_THREAD_LOCKED •Defective PCA •Replace Main PCA
E-69 ERR_MAIN_EXECUTION_THREAD_LOCKED •Defective PCA •Replace Main PCA
E-70 ERR_INTERNAL_RAM_FAILURE •Defective microprocessor •Replace Main PCA
E-71 ERR_EVENT_LOG_QUEUE_FULL •Defective PCA •Replace Main PCA
E-72 ERR_TREND_DIV_BY_ZERO •Defective RAM chip •Replace Main PCA
E-73 ERR_ROLLOVER_CASE •Defective RAM chip •Replace Main PCA
E-74 ERR_WRONG_DISPLAY_TYPE •Defective PCA •Replace Main PCA
E-75 ERR_ALARM_CAPACITOR_FAILURE •Defective PCA •Replace Main PCA
E-76 ERR_OVER_PRESSURE •Occluded or restricted tube •Defective PCA •Replace Pressure tubing •Replace Main PCA
E-77 ERR_UNDER_PRESSURE •Occluded or restricted tube •Defective PCA •Replace Pressure tubing •Replace Main PCA
E-78 ERR_OVER_PRESSURE_REG •Occluded or restricted tube •Defective PCA •Replace Pressure tubing •Replace Main PCA
E-79 ERR_UNDER_PRESSURE_REG •Occluded or restricted tube •Defective PCA •Replace Pressure tubing •Replace Main PCA
E-80 ERR_ROTOR_LOCKED •Locked Motor •Defective PCA •Replace Motor •Replace PCA
E-81 ERR_MOTOR_EDGE_MISSED •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-82 ERR_MOTOR_EDGE_EXTRA •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-83 ERR_MOTOR_OVERSPEED •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-84 ERR_MOTOR_BRAKE •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-85 ERR_MOTOR_CURRENT •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-86 ERR_INVALID_ERROR_CODE •Reserved

6.5 SmartCard Error Codes:

Code Label - User Action

C-1 SMARTCARD_12C_WRITE_FAIL Card inserted upside-down. Remove card and reinsert properly. Otherwise, card is corrupt and should be reprogrammed.
C-2 SMARTCARD_BAD_CARD_TYPE Card is not supported by this device. Otherwise, card is corrupt and should be reprogrammed.
C-3 SMARTCARD_BAD_CIB_VERSION Card is corrupt and should be reprogrammed.
C-4 SMARTCARD_SCRIPT_DURING_PROVIDER_MODE Remove card. Exit Provider Mode screens. Then reinsert card.
C-5 SMARTCARD_PRESENT_IN_CAL_MODE Remove card. Exit Calibration Mode. Then reinsert card.
C-6 SMARTCARD_BAD_CIB_CHECKSUM Card is corrupt and should be reprogrammed.
C-7 SMARTCARD_NOT_PRESENT
C-8 SMARTCARD_12C_READ_FAIL Card inserted upside down. Remove card and reinsert properly. Otherwise, card is corrupt and should be reprogrammed.
C-9 SMARTCARD_WRITE_CONFIRM_FAIL Card is corrupt and should be reprogrammed. Otherwise, replace card.
C-10 SMARTCARD_BAD_RI_BRAND Card is corrupt and should be reprogrammed.
C-100 SMARTCARD_BAD_CPB_CHECKSUM Card is corrupt and should be reprogrammed.
C-101 SMARTCARD_BAD_CPB_VERSION Card is corrupt and should be reprogrammed.
C-102 SMARTCARD_INVALID_UNIT
C-103 SMARTCARD_UNKNOWN_CPB_FORMAT Card contains a prescription format that is not supported. Perhaps card is intended for another device.
C-104 SMARTCARD_UNKNOWN_CPB_PRESCRIPTION Card contains a prescription type that is not supported. Perhaps card is intended for another device.
C-105 SMARTCARD_SCRIPT_PARM_RANGE Card contains prescription settings that are outside of this device’s range. Perhaps card is intended for another device.
C-106 SMARTCARD_BAD_CPB_SCRIPT_READY Card is corrupt and should be reprogrammed.
C-107 SMARTCARD_BAD_CPB_START_ADDR Card is corrupt and should be reprogrammed.
C-108 SMARTCARD_BAD_CPB_END_ADDR Card is corrupt and should be reprogrammed.
C-109 SMARTCARD_BAD_CPB_ERASE_PERMISSION Card is corrupt and should be reprogrammed.
C-200 SMARTCARD_BAD_CDCB_VERSION Card is corrupt and should be reprogrammed.
C-201 SMARTCARD_BAD_CDB_VERSION Card contains a storage format that is not supported. Perhaps card is intended for another device.
C-202 SMARTCARD_BAD_CDCB_CHECKSUM Card is corrupt and should be reprogrammed.
C-203 SMARTCARD_BAD_CDCB_CONTROL_BLOCKS Card is corrupt and should be reprogrammed.
C-204 SMARTCARD_BAD_CDCB_START_ADDR Card is corrupt and should be reprogrammed.
C-205 SMARTCARD_BAD_CDCB_END_ADDR Card is corrupt and should be reprogrammed.
C-206 SMARTCARD_BAD_CDB_START_ADDR Card is corrupt and should be reprogrammed.
C-207 SMARTCARD_BAD_CDB_END_ADDR Card is corrupt and should be reprogrammed.
C-300 SMARTCARD_BAD_UIDB_VERSION Card is corrupt and should be reprogrammed.
C-301 SMARTCARD_BAD_UIDB_CHECKSUM Card is corrupt and should be reprogrammed.
C-302 SMARTCARD_BAD_UIDB_START_ADDR Card is corrupt and should be reprogrammed.
C-303 SMARTCARD_BAD_UIDB_END_ADDR Card is corrupt and should be reprogrammed.
C-400 SMARTCARD_BAD_DDCB_VERSION Card is corrupt and should be reprogrammed.
C-401 SMARTCARD_BAD_DDCB_CHECKSUM Card is corrupt and should be reprogrammed.
C-402 SMARTCARD_BAD_DDCB_PROFILE_CONTROL_BLOCKS Card is corrupt and should be reprogrammed.
C-403 SMARTCARD_BAD_DDCB_HISTOGRAM_CONTROL_BLOCKS Card is corrupt and should be reprogrammed.
C-404 SMARTCARD_BAD_DDCB_FOSQ_COMPL_CONTROL_BLOCKS Card is corrupt and should be reprogrammed.
C-500 SMARTCARD_FILE_OPEN_FAIL Card is corrupt and should be reprogrammed.
C-501 SMARTCARD_FILE_TELL_FAIL Card is corrupt and should be reprogrammed.
C-502 SMARTCARD_FILE_SEEK_FAIL Card is corrupt and should be reprogrammed.
C-503 SMARTCARD_FILE_READ_FAIL Card is corrupt and should be reprogrammed.
C-504 SMARTCARD_FILE_WRITE_FAIL Card is corrupt and should be reprogrammed.
C-505 SMARTCARD_FILE_GET_PTR_FAIL Card is corrupt and should be reprogrammed.
C-506 SMARTCARD_FILE_SET_PTR_FAIL Card is corrupt and should be reprogrammed.

SuperSleeper
Apnea Board Administrator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.


Post Reply Post Reply
#3
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
minime Wrote:My husband didn't empty the humidifier before packing his c-pap up, I fear the machine is toast. There is no error code, only that it needs service. No air comes out and the buttons flash and beep. Is there anything we can do to get this working again?

TIA, minime
Post Reply Post Reply
#4
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
minime,(time=1263670937) Wrote:My husband didn't empty the humidifier before packing his c-pap up, I fear the machine is toast. There is no error code, only that it needs service. No air comes out and the buttons flash and beep. Is there anything we can do to get this working again?

TIA, minime
Hi minime, welcome to the forum!

So I take it that the machine got soaked with water, correct? If that's the case, you need to allow it to completely dry out before you plug it back in, perhaps for a few days in a dry area with good air circulation (even in front of a fan if you want). If you plug it back in without all the water evaporated off the circuit boards and electrical connections, it could short out.

If that's what happened (it shorted out), you're right, it may indeed be "toast".

If this machine was given to you through your home health equipment provider (DME) and is a rental, contact them and tell them that it's not working. If not, and it's an expensive machine, you may want to send it to the manufacturer for repairs. Or, you could purchase another one. You can get a very basic CPAP over the Internet for about $150 these days, as long as you have a copy of your prescription from your doctor.
SuperSleeper
Apnea Board Administrator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.


Post Reply Post Reply
#5
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
A member emailed me this on the inner workings of a standard CPAP machine... it may be of help...


A View Inside a Respironics REMstar Pro CPAP Machine


Broken REMstar Pro CPAP Machine A few weeks ago I was looking around our stock room when I came across a non-functioning CPAP machine sitting on a shelf. It was an old REMstar Pro with an E-24 error, usually associated with some sort of water damage (figure 1). With the same level of curiosity I had with machines (radios, tvs, video games, etc.) when I was a kid, I found a screw driver and opened up the machine. We don't repair CPAP machines here, so this was the first time I'd ever seen the inside of a CPAP machine. Not a big surprise, really, except for the fact that the CPAP machine case was filled with mostly sound-absorbing foam. Besides that, of course, there was a circuit board with buttons and an LCD display screen, a blower motor and some tubing connecting the blower to the air outlet port on the CPAP machine.

FIGURE 1
[Image: attachment.php?aid=25]

I've taken pictures of the components of this CPAP machine and I'm going to share them here so that you know a little bit more about your unit, even if it's not the exact same CPAP machine. After all, every CPAP machine works in pretty much the same way. These photos should also prevent you from feeling any temptation to open your CPAP machine out of curiosity. If your CPAP machine is within it's warranty period and you open it, it's likely that your warranty will be voided. So, sit back and enjoy my broken CPAP machine.

Top Enclosure Removed from the CPAP Machine Opening the Respironics REMstar Pro CPAP machine only requires the removal of three philips head screws. Once those screws are removed the top enclosure of the CPAP machine pops right off (figure 2). Some machines are a bit trickier to open, such as the Puritan Bennett 420G. The Puritan Bennett 420G CPAP machine only has four screws, but they're weird ones that appear to require some sort of star-shaped allen-wrench-type of tool for removal. The Puritan Bennett 420G is made in France, which likely explains the unusual type of fastener.

FIGURE 2
[Image: attachment.php?aid=26]

Circuit Board - or Main PCA - of the REMstar Pro CPAP MachineOnce the top enclosure is removed from the CPAP machine the fun can begin. The circuit board - also known as the main printed circuit assembly or Main PCA (figure 3) is attached to the top enclosure and can easily be removed from the top enclosure. Two wires connect the blower assembly to the circuit board. Removing these two wires allows you to completely remove the top enclosure and circuit board from the main body of the CPAP machine. Figure 4 shows the CPAP machine completely separated into two parts.

FIGURE 3
[Image: attachment.php?aid=27]

FIGURE 4
[Image: attachment.php?aid=28]

CPAP Machine in Two Parts


The Main PCA is the control center of the REMstar Pro CPAP System. The Main PCA utilizes a microcontroller chip which reads from and writes to various I/O devices (control pad, LCD, memory, pressure sensor, motor control circuitry, RS-232 interface, etc.). With all of this information, the microcontroller makes appropriate decisions so the CPAP machine blower is properly controlled to deliver the desired CPAP therapy in a safe and reliable manner. The serial number of the CPAP machine is programmed into the memory of the Main PCA during the manufacturing process.

Part of the Main PCA consists of the slot for the Encore SmartCard. More sophisticated CPAP machines like the REMstar Pro offer SmartCard functionality. On the Respironics REMstar series of CPAP machines, the SmartCard is a plastic card similar in size and shape to a credit card. Instead of holding information on a magnetic strip, it holds it in a small silicon chip embedded into the card.

When installed in the REMstar Pro CPAP machine, the SmartCard records the date, time at pressure, pressure levels, and duration of each use. Using the Encore software, the usage data can be downloaded and viewed. You can see the Encore SmartCard in figure 4.

REMstar Pro CPAP Machine Sound Absorbing FoamNotice that the bulk of the interior of the CPAP machine consists of sound- and vibration-absorbing foam (figure 5). My estimate is that about 65% of the interior of the CPAP machine is filled with foam. Surrounded by the foam is the blower assembly, which is obviously a key component of the CPAP machine - the part of the CPAP machine that actually generates air pressure (figure 6). The blower is a three phase brushless DC motor with a permanent magnet rotor and stationary coils that are conenected in a "Y" configuration and serve as its stator coils. The microcontroller sets the motor speed - and this in turn determines the output pressure - by varying the voltage which is delivered to the motor.

FIGURE 5
[Image: attachment.php?aid=29]

REMstar Pro CPAP Machine Blower Assembly

In figure 6 you can see that the CPAP blower assembly is a fairly simple design. The motor makes a plastic "fan" spin really fast inside the blower assembly and this spinning fan generates air pressure and pushes the air out of the blower assembly and through the pressure outlet port. This particular REMstar Pro CPAP machine features C-Flex which is a technology which lowers the pressure setting when the CPAP user exhales. When exhalation is sensed the micro-processor sends a signal to the motor (in the form of a lower voltage) in order to briefly slow the motor down. Upon inhalation the voltage to the motor is increased and the motor speeds up to deliver a higher inhalation pressure. This happens with each breath you take - thousands of times per night.

FIGURE 6
[Image: attachment.php?aid=30]

That's all there is to a CPAP machine! Micro-processing technology and a motor hooked up to a fan. This should satisfy your curiosity about the inner workings of your CPAP machine. If you do ever take apart your CPAP machine, be sure not to plug it in. It'll work just fine, and that's the problem. There's a risk of electric shock, so be careful.

Note: this article includes some technical information copied from the REMstar Pro Service Manual.


Attached Files Thumbnail(s)
                           
SuperSleeper
Apnea Board Administrator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.


Post Reply Post Reply
#6
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
rayhapes,(time=1266741842) Wrote:Ok, now I don't feel quite so dumb.. I have soaked my Goodknight 420s. I did turn it back on before dry however. I have let it dry out and am getting error 15. could you post the error codes for this model please?
Hi ray, and welcome to Apnea Board!

I don't currently have the error codes for the Puritan Bennett machines - maybe someone who has them can post them here. Might want to check back from time to time and see.

SuperSleeper
Apnea Board Administrator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.


Post Reply Post Reply
#7
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
asmawi Wrote:my CPAP can not power on. already checked. supply fuse burnt. also power regulator burnt. but unfortunately i don't have the circuit diagram to check the power regulator part number. regulator completely burn. couldn't see the part number. anybody can help me to give the power regulator part number. my model #1009587 RemPlus w/C-Flex Intl.
attached herewith the damaged board...any help is very much appreciated...many thanks.

[Image: attachment.php?aid=32]


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#8
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
asmawi,(time=1280922427) Wrote:my CPAP can not power on. ... the power regulator part number. my model #1009587 RemPlus w/C-Flex Intl.
attached herewith the damaged board...any help is very much appreciated...many thanks.
Hi asmawi, welcome to Apnea Board!

We have a copy of the Service Manual for the Respironics REMstar Plus w/ Cflex machine

The service manual should list all the part numbers and also help you service the machine, I would think.

You can get this service manual and many other service & repair manuals (available for download) online, in our Private Files & Links Area (you'll need to sign in to get access to the forum), then find the thread that says "Links to Members Only Files".

SuperSleeper
Apnea Board Administrator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.


Post Reply Post Reply
#9
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
djharkavy Wrote:Is there any way to do a reset of the error codes. I am travelling and my Remstear Auto M-Series just started flashing the 'service required.'

It beeps and flashes the lights, but does not start up. I can have it fixed/replaced when I get home, but I would like to get it running for the next few days.

No error codes showing, and I don't know how to get them otherwise.
Post Reply Post Reply
#10
RE: Service Required Error Codes - Inner Workings of a CPAP machine [copied]
claudeg Wrote:
SuperSleeper,(time=1266768276) Wrote:
rayhapes,(time=1266741842) Wrote:Ok, now I don't feel quite so dumb.. I have soaked my Goodknight 420s. I did turn it back on before dry however. I have let it dry out and am getting error 15. could you post the error codes for this model please?
Hi ray, and welcome to Apnea Board!

I don't currently have the error codes for the Puritan Bennett machines - maybe someone who has them can post them here. Might want to check back from time to time and see.
Looking at the clinician manual for the Goodknight 420g, the only error codes that are explained are "In 1" (hose or mask disconnected) and "In 3" (power supply voltage dropped). In the troubleshooting guide section, the cause for "Er XX" is "The device has detected an operating fault" and the recommended action is "Contact the technical support". That's French for "You're up the creek, monsieur." Either that, or "Muck in the U-joint."

Funny how it's always muck in the U-joint.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Is it safe to use a refurbished/used CPAP machine purchased from an end user online? jake79 31 772 11-15-2024, 10:25 PM
Last Post: Deborah K.
  Best Backup CPAP Machine Options? Lucid 14 556 11-14-2024, 10:21 AM
Last Post: jake79
  [Equipment] Looking to buy my own CPAP machine mrjeswan 6 544 11-11-2024, 06:45 AM
Last Post: tundranapper
  Service Required Message jmack729 2 170 11-06-2024, 11:57 AM
Last Post: StratCat48
  Resmed Airsense 10 Error 006 lanecito 2 268 11-03-2024, 06:24 PM
Last Post: PeaceLoveAndPizza
Exclaimation CPAP Machine Choices - read this before you accept a new machine SuperSleeper 305 283,472 10-28-2024, 07:30 PM
Last Post: ZacharySimon
Idea Generator Suggestions For CPAP Machine? edenmaker 11 659 10-21-2024, 06:34 PM
Last Post: edenmaker


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.