How does Flow Limitation affect pressure
|
04-24-2016, 11:48 AM
How does Flow Limitation affect pressure
New to using APAP ResMed Autoset for her and trying to understand how pressure is increased. I know that OA's cause the pressure to increase but CA's have no affect.
Can someone explain how/if Flow Limits cause increase in pressure. I believe on Resmed units the shape of the inhalation over a sample time period determines Flow Limitation. Does this sampling cause both increase and decrease in pressure?
04-24-2016, 12:02 PM
RE: How does Flow Limitation affect pressure
The way I understand it is that an OA is a 100% flow limit, so when the CPAP starts seeing flow limits it starts expecting to see OA's. To prevent the start of OA's the CPAP starts raising pressure at the first sign of flow limits
04-24-2016, 12:05 PM
RE: How does Flow Limitation affect pressure
You can find Resmed's definition and reaction to flow limitation here: http://www.resmed.com/int/clinicians/com...nc=dealers
Read the linked page, and if you have further questions, post back. Quote:1. Flow Limitation
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
04-24-2016, 12:31 PM
RE: How does Flow Limitation affect pressure
The Autoset line probably has the same algorithm for managing pressure thru the Flow Limitation. My Flow rate has inhalation flat topping and the pressure shows little upticks (the inspir. time is flat). Assume the flow rate is being sampled but most of my inhalations are more flat than rounded so hard to see relationship to pressure adjustment. My pressure range is currently set to 7-11 (originally at 7-14) by DME). Would resetting the pressure back to 7-14 be recommended?
04-24-2016, 12:44 PM
RE: How does Flow Limitation affect pressure
(04-24-2016, 12:31 PM)Roy289 Wrote: The Autoset line probably has the same algorithm for managing pressure thru the Flow Limitation. My Flow rate has inhalation flat topping and the pressure shows little upticks (the inspir. time is flat). Assume the flow rate is being sampled but most of my inhalations are more flat than rounded so hard to see relationship to pressure adjustment. My pressure range is currently set to 7-11 (originally at 7-14) by DME). Would resetting the pressure back to 7-14 be recommended? What is your average and 95% pressure currently? If the machine is reaching its maximum pressure regularly, then you probably need a higher maximum pressure. Your question is pretty hard to answer without seeing the data to understand why a higher pressure may, or may not be needed. Generally, I don't recommend chasing flow limitation events unless they are disrupting your sleep or are precursors of OA or H events.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
04-24-2016, 01:19 PM
RE: How does Flow Limitation affect pressure
Statistics Channel Min Med 95% Max W-Avg: 9.51 5.00 9.90 11.00 11.00 Should I increase by increments of 1? Brought it down originally as was chasing mask leakage.
04-24-2016, 02:23 PM
RE: How does Flow Limitation affect pressure
My Flow Limitation graph provides a false positive to the Resmed and will cause the unit to go to .2cm below my programmed upper limit, regardless of how high the pressure setting. I've found that a straight pressure of 11 produces the lowest AHI and best sleep quality. Any pressure higher than 11 isn't beneficial and doesn't reduce the flow limitation graph. . . it just stays the same. If I had the money I'd like to try a P-R to see if "likes" me in this area.
Crimson Nape
Apnea Board Moderator Project Manager for OSCAR - Open Source CPAP Analysis Reporter www.ApneaBoard.com ___________________________________ Useful Links -or- When All Else Fails: The Guide to Understanding OSCAR OSCAR Chart Organization Attaching Images and Files on Apnea Board Apnea Helpful Tips INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
RE: How does Flow Limitation affect pressure
apneas are not flow limitations, and vice versa, just to clarify.
an apnea is a > 90% reduction in flow from the baseline for ten or more seconds. a hypopnea is a > 30% reduction in flow from baseline for ten or more seconds. a flow limitation does not necessarily correspond with either of the above, a flow limitation is a restriction in the airflow, making it harder to breath, which causes increased breathing effort, and can lead to arousals, however there does not have to be any reduction in total flow due to flow limitations, (thinking gulping a drink vs sipping it, you get the same amount of drink, just takes longer one way than the other)
04-24-2016, 04:03 PM
RE: How does Flow Limitation affect pressure
Palerider
Where does you definition of an Apnea come from? Is the formula use in SleepyHead or Rescan. I ask because I have been reading medical Journals and I find the definition varies by source (04-24-2016, 03:57 PM)palerider Wrote: apneas are not flow limitations, and vice versa, just to clarify.
2004-Bon Jovi
it'll take more than a doctor to prescribe a remedy Observations and recommendations communicated here are the perceptions of the writer and should not be misconstrued as medical advice.
04-24-2016, 08:56 PM
RE: How does Flow Limitation affect pressure
(04-24-2016, 04:03 PM)0rangebear Wrote: PaleriderSleepyHead does NOT have a "formula" for defining apneas or hypopneas. It reads the data that is scored by the machine and displays that data. Each manufacturer has their own algorithms for determining when to score an apnea or a hypopnea. They all are based on "reduction in airflow" from a moving baseline, although how to compute that moving baseline and long the "moving baseline window" is may vary from manufacturer to manufacturer. There's a fascinating white paper from the NHI at http://www.ncbi.nlm.nih.gov/pmc/articles/PMC4629962/ that contains a table that has some pretty extensive technical information about the criteria that Resmed, PR, and DeVilbass machines use to detect events. The table can be found at http://www.ncbi.nlm.nih.gov/pmc/articles...der-8-425/ The table indicates that Resmed machines score an apnea when the flow is reduced to less than 25% of baseline for at least 10 seconds whereas the PR machines score an apnea when the flow is reduced to less than 20% of baseline for at least 10 seconds. DeVilbass APAPs score apneas when the flow is reduced to less 10% of baseline for at least 10 seconds. Hypopnea definitions also vary from machine to machine:
The detection of flow limitations also vary by manufacturer: Resmed looks at each and every breath, PR does a 4-breath average, and DeVilbass uses an average of all breaths over a 12-second range. Another fascinating thing about the table is that it details how each of the three manufacturers have programmed their auto algorithms to respond to apneas, hypopneas, flow limitations, and snoring. Snoring is discussed in the paper itself, and I haven't dug into that part. (Tables are much easier to process quickly.) There's a lot of really interesting info in this paper for tech-geeks and data-nerds. Finally, it's also worth noting that there are two different ways that a lab can choose to score a hyponea on an in-lab sleep test according to the titration guidelines set up by the AASM, although the two ways are now based mainly on Medicare and some other insurance companies refusing to recognize the newer standards. Both ways to score hypopneas now seem to require a 30% drop in airflow for at least 10 second. The one that the AASM wants to use requires either a 3% drop in O2 OR an arousal. The one that Medicare insists on requires a 4% drop in O2. I wonder if PR has updated their algorithm for scoring Hs on the DreamStations. See http://www.aasmnet.org/articles.aspx?id=4132 and http://www.aasmnet.org/articles.aspx?id=3782 for details about the AASM rules for scoring hyponeas. |
« Next Oldest | Next Newest »
|
Possibly Related Threads... | |||||
Thread | Author | Replies | Views | Last Post | |
Expiratory Flow Limitation? | deebob | 0 | 59 |
11 hours ago Last Post: deebob |
|
Help Needed with CPAP Flow Chart Interpretation | HijolG | 19 | 1,014 |
11-23-2024, 03:25 PM Last Post: HijolG |
|
Flow Limitations? | gr3gg0 | 13 | 640 |
11-23-2024, 11:12 AM Last Post: SeePak |
|
[CPAP] Increased EPR reduces flow limits even when EPAP is simultaneously decreased | G. Szabo | 5 | 449 |
11-16-2024, 09:16 PM Last Post: SeePak |
|
EPR 1, hight flow limit, but lots of CA | throwaway1242 | 13 | 842 |
11-15-2024, 07:30 PM Last Post: OpalRose |
|
Low AHI, high flow limits | theswerve | 1 | 209 |
11-15-2024, 03:30 PM Last Post: PeaceLoveAndPizza |
|
Flow rate analysis | jake79 | 4 | 302 |
11-12-2024, 12:27 PM Last Post: jake79 |