Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

LEAKS and their measurement.
#1
LEAKS and their measurement.
One of the biggest issues that users have to contend with is leaks. 
But what do we know about how leaks are measured?
I'm making an assumption there are sensors (pressure and flow) are in the body of the cpap machine.
The pressure sensor , after adjusting for pressure loss through the tubing, can monitor the mask pressure.

But how does the flow sensor "measure" the loss of air volume when there is a leak?

(My interest here is to understand how direct a measurement (of leakage) actually is. Seems like a lot of so-called measurements are actually inferences drawn from other measures. For example oxygen saturation inferred by measurement of color absorption, or sleep cycles inferred from "tone" of capillaries. )
Post Reply Post Reply
#2
RE: LEAKS and their measurement.
It's a direct reading from a single sensor on the machine's air intake (ResMed Airsense10). The raw data is filtered through an algorithm that attempts to find the "zero point" of flow into and out of the user's lungs inferred from the feedback of flow (and maybe also references pressure for this?) during inhale and exhale cycles. Anything greater than the expected value for the mask used (based on a simple FFM/nasal/pillows setting) is reported as a leak.

All told, it does an admirable job of estimating this from a limited amount of raw data, but it's pretty easily fooled. While awake, I have drawn in a long slow breath that was reported as a leak and a CA followed by a mysteriously large exhale.
Look, I'm an engineer, not a doctor! Please don't take my opinion as a substitute for medical advice.
Post Reply Post Reply
#3
RE: LEAKS and their measurement.
Engineer here as well (it's a cross we just have to bear!)

If I follow your reasoning, the leak calculation* is influenced by what the device infers the users tidal volume is (because the device doesn't directly measure tidal volume either).  The tidal volume is calculated by way of pressure differentials assuming constant supply of airflow. 

* I tend to say measurement value when there is a determined by an actual measurement; and I tend to say "calculated value" when the value is determined by an algorithm (aka "blackbox").
Post Reply Post Reply
#4
RE: LEAKS and their measurement.
Tidal volume is (I believe) calculated from the integration of deviation in real-time flow rates from the moving average. The moving average is where you see the red dotted line in OSCAR, which is straightened out for us in the report to make it easier to tell what's going on. When the real-time flow is up from that average, the machine attempts to determine if the flow is into the lungs of the user, or out the side of the mask (or in my case, more than likely right out my open mouth). When real-time flow is below that average, it gets assigned to exhale flow.

Tidal volume is the area between these curves and the baseline. An attempt appears to be made to keep the area above and below equal, which we can see on the charts of people with palatal prolapse. There is a normal inspiration curve, but flow is locked off to the mask and goes out the mouth due to the one-way valve action of their palate. This flow rate should technically be zero, on the red line, but the machine doesn't really know that, so assigns a portion of the mask venting as exhale flow to keep a balanced tidal volume.

Just some observations I've made and the explanations of what I think is behind them. If you want the real scoop, try calling ResMed to see if they'll feed you anything.
Look, I'm an engineer, not a doctor! Please don't take my opinion as a substitute for medical advice.
Post Reply Post Reply
#5
RE: LEAKS and their measurement.
Is it your understanding that the blower motor speed is variable? Such that the blower tries to maintain the command pressure?
Post Reply Post Reply
#6
RE: LEAKS and their measurement.
Yes, blower motors in ResMed machines are controlled to maintain pressure, not flow. Whatever flow results is a measured value, not a control value according to my understanding. I think Respironics machines use flow as a control parameter in addition to pressure, and I'm not sure about any other manufacturers.
Look, I'm an engineer, not a doctor! Please don't take my opinion as a substitute for medical advice.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Fighting Leaks and Other Challenges Brennie 10 428 8 hours ago
Last Post: SarcasticDave94
  Increasing AHI despite fewer leaks Aife 24 1,348 11-14-2024, 03:01 PM
Last Post: Jay51
  Cant manage mask leaks at all sfgiants123 5 376 11-09-2024, 10:49 PM
Last Post: Alfie
  Evora FFM Phantom Leaks psinderbrand 25 2,117 11-08-2024, 10:06 AM
Last Post: DCLT
  Why am I getting more leaks with the P10? ttsoldier 5 306 11-05-2024, 06:13 PM
Last Post: Deborah K.
  Aircurve 10 ASV Periodic Pressure Surges cause leaks scooterbill 2 161 11-04-2024, 04:31 PM
Last Post: SarcasticDave94
  Very high lpm leaks 59Bugeye 0 152 10-29-2024, 08:02 AM
Last Post: 59Bugeye


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.