(04-28-2021, 07:06 PM)Geer1 Wrote: What is curious about that example you posted isn't the OA itself but that there were three other blockages/restrictions in same window. Those were obstructive in nature but because they are less than 10 seconds in length they don't get flagged as apnea or hypopnea.
You appear to be having more OA at these recent settings and that makes sense.
Your minimum pressure (EPAP) is what stints airway open to stop obstructive apnea from occurring. Initially you were using a fixed pressure of 13 cm with no EPR so your EPAP was constant at 13 cm. You followed directions and lowered min pressure to 9 cm and turned EPR on and to 3 cm which makes your minimum EPAP 6 cm. Because you also turned the machine onto auto it increases EPAP on its own but it does so reactively (after apnea or flow limitations have occurred). On the 27th your average EPAP was 9.42, still significantly lower than your original settings.
Hopefully that makes some sense. In short you should increase EPAP (by increasing min pressure) to try and treat these obstructions. Since min EPAP = min pressure - EPR my recommendation is a minimum pressure of 13 cm (10 cm EPAP). This is still lower than your original settings but hopefully will be enough to stint your airway open.
I would also increase the max pressure to 20 cm as the machine came close to/hit your max pressure setting of 15 cm both nights. If it starts going too high and is uncomfortable then you can reduce maximum but if the higher pressure doesn't bother you it might help resolve a few more events.
I had the settings as min 9 and max 16 as recommended earlier. I am not able to handle the max pressure as I end up choking and gasping for breath. I think 16 is just about the threshold for max that I am able to handle. I will increase the min pressure gradually and post the results.
Thank you,
Best Regards,
Kashyap