I am new to using an older Remstar Pro C-Flex device, try DIY rather than normal channel.
The device seems undamaged and cycles through the set up screens as expected.
I want to start out at low pressure and assess the impact.
The settings menu allows me to increment "pressure" from 4 cm to 20 cm.
However, when I exit and start the machine at either 4 cm or 20 cm, I can detect no difference in the airflow (it seems high).
Anyone have experience or suggestions?
Should there normally be a noticeable difference in the airflow?
RE: Air Pressure
There should be a difference. Does it have a ramp feature that is turned on? If so turn off the ramp and you should notice a difference.
Dave
05-24-2017, 02:10 PM
(This post was last modified: 05-24-2017, 02:11 PM by TASmart.)
RE: Air Pressure
There should not be much difference. Open flow your machine is trying to hold a pressure and since there is not any stopping the flow, it maximizes it's output at either setting, blowing like crazy trying to get to the set pressure. The only flow that your machine is designed to have is any leak amount plus the volume of your inhalation. Flow is no the same as pressure. If you are feeling flow you have a leak problem.