It doesn’t say CPAP auto it just says CPAP so I’m assuming it only goes up to set pressure. The other two modes are S and Vauto. When I have Vauto set the pressure lingers after my inhale if that makes sense so it sort of triggers central’s. In S mode I’m clueless. Ive been reading about low pressures being better for central’s but not sure what to set machine to anything helps thanks
Hello Guest,
Welcome to Apnea Board !As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address. Login or Create an Account
ResMed AirCurve 10 Vauto Help
|
10-13-2019, 10:57 AM
ResMed AirCurve 10 Vauto Help
Hello all, this is my first thread on apneaboard and would like to ask some questions about the AirCurve 10 vauto. I used the machine last night in hopes to resolve some of my central’s. I started in CPAP MODE and set pressure on 6.5
It doesn’t say CPAP auto it just says CPAP so I’m assuming it only goes up to set pressure. The other two modes are S and Vauto. When I have Vauto set the pressure lingers after my inhale if that makes sense so it sort of triggers central’s. In S mode I’m clueless. Ive been reading about low pressures being better for central’s but not sure what to set machine to anything helps thanks
RE: ResMed AirCurve 10 Vauto Help
I think it would help people help you if you could post some background data to help understand why you're asking these questions.
For example, your sleep study results? How about previous examples (on CPAP) of the "central" problem you note? FWIW, I see no reason not to use the VAuto mode. It's flexible enough to be configured however you need it.
10-13-2019, 12:25 PM
RE: ResMed AirCurve 10 Vauto Help
Welcome to the forum, we can and will help you.
Download OSCAR and post your Daily charts. Look in the Organization link in my signature for which charts to include, some are more important than others. Specifically what to do depends on what kind of events you have and the quantity and when they occur and what is occurring around them. The ROT is that for Central events we need to lower pressure and the differential in pressure that occurs both algorithmically and from the difference between EPAP and IPAP pressure that occurs on every breath. This assumes that the cause of centrals is from the increased washing out of CO2 from the blood that frequently occurs when CPAP is initiated. It would help us if you post (redacted) full copies of your sleep studies, not just the summaries.
Gideon - Project Manager Emeritus for OSCAR - Open Source CPAP Analysis Reporter
Download OSCAR OSCAR - The Guide New to Apnea? Helpful tips to ensure success Soft Cervical Collar Mask Primer Dealing with a DME Organize Charts Optimizing Therapy
10-13-2019, 02:13 PM
RE: ResMed AirCurve 10 Vauto Help
The Vauto is not capable of treating central apnea. A higher level bielvel device like the Aircurve 10 ASV is needed to perform the functions mentioned by Bonjour. With the Vauto, CPAP pressure is a fixed single pressure, but EPR (exhale pressure relief) is availablle to provide up to 3-cm of pressure reduction during exhale. VPAP S means a fixed bilevel pressure with separate IPAP and EPAP pressures, where the difference between IPAP - EPAP = Pressure Support (PS). Vauto has a minimum EPAP pressure and fixed PS, and will vary pressure up to the maximum pressure set. (EPAP - PS = IPAP).
In general, fixed low pressure works best for central apnea. Using bilevel pressure without a backup that triggers IPAP pressure just makes CA worse. The ASV has the backup that triggers IPAP and adaptively adjusts the pressure to ensure your minute vent remains stable.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
10-13-2019, 02:40 PM
RE: ResMed AirCurve 10 Vauto Help
Thank you everyone for replying so quickly. I’m trying to give my best with the equipment I have. Without an ash machine you’re saying that I can get somewhat close to ASV therapy by lowering the pressure support and enabling the trigger setting to very high so the pressure catches the slightest inhale? A fixed pressure like CPAP mode won’t be beneficial from what I understand.. what about easy breathing? Or the auto feature with a low pressure support and high trigger? Or shouldn’t I even bother
10-13-2019, 03:34 PM
RE: ResMed AirCurve 10 Vauto Help
Ben, if you have predominately central apnea, the ASV is your best option. The Vauto cannot treat CA because it relies on spontaneous respiratory effort to trigger IPAP, regardless of sensitivity. At this point, we really don't know what you're dealing with in terms of severity or past results. The Aircurve 10 Vauto is a fantastic machine for obstructive apnea as well as treating hypopnea and flow limitation.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
RE: ResMed AirCurve 10 Vauto Help
(10-13-2019, 02:40 PM)Ben024 Wrote: I’m trying to give my best with the equipment I have. Sure. Absolutely. An ASV or another machine takes time to get so why not use what you have to make each night until then better and learn something along the way. (10-13-2019, 02:40 PM)Ben024 Wrote: A fixed pressure like CPAP mode won’t be beneficial from what I understand.. what about easy breathing? Lets start here. Correct, CPAP mode is likely to have less potential value than all the other options. Easy Breathing is generally a good thing. I will just say a research paper comes to mind that says people would find better final settings faster by titrating (working to find your best setting) with Easy Breathing turned on early rather than consider it an add on for later, because Easy Breathing might change how your body responds to various settings EPAP, IPAP, PS and EPR settings. (10-13-2019, 02:40 PM)Ben024 Wrote: Without an ash machine you’re saying that I can get somewhat close to ASV therapy by lowering the pressure support and enabling the trigger setting to very high so the pressure catches the slightest inhale? First I will answer your question and then add a "but really you should do this instead." "I can get somewhat close to ASV therapy by lowering the pressure support and enabling the trigger setting to very high so the pressure catches the slightest inhale?" Nope. That is not it. And now the "... really you should do this instead." Actually, to Ben024 and EVERYONE else wondering if an ASV is right for them, to slowly, night by night replicate the ASV to see if that is what you need you should titrate: (Others board members: This approach is an on-the-fly developed draft, please reply with improvements.) Also, you may not need an ASV because sometimes CAs go away when another PAP solution is well aligned with a person's needs. The process below will help sort that out. So lets get to it. To Titrate to see what machine you really need while assuming in advance that you need an ASV. First methodically titrate EPAP up to find out how high it needs to be to remove Obstructive Apneas - OAs. (Just like you would an ASV, or what an ASV would sort of automatically nearly do for you.)
Everyone, please reply and improve. Time to Snooze. Zzzzzzz WillSleep A few key links Bonjour posted above Soft Cervical Collar Organize Charts Attaching Charts Dealing with a DME INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
10-14-2019, 07:47 AM
RE: ResMed AirCurve 10 Vauto Help
Will, I would rather see you work out your theory in a separate thread, not another users thread
If Ben were to follow your directions it is extremely likely that his Central Apnea would become profoundly worse. The issue is that you are counting on a signal from his body that has essentially been turned off, the signal to breathe. No CPAP or BiLevel without backup, without timed breathing, or a full on ventilator can compensate for the lack of single to breathe. By titrating PS as you have, especially with a susceptible user like Ben, you have greatly increased the efficiency of his breathing resulting in the increased washout of CO2 from his blood which decreases the drive to breathe causing severe Central Sleep Apnea.
Gideon - Project Manager Emeritus for OSCAR - Open Source CPAP Analysis Reporter
Download OSCAR OSCAR - The Guide New to Apnea? Helpful tips to ensure success Soft Cervical Collar Mask Primer Dealing with a DME Organize Charts Optimizing Therapy
10-14-2019, 12:50 PM
RE: ResMed AirCurve 10 Vauto Help
(10-14-2019, 07:47 AM)bonjour Wrote: If Ben were to follow your directions it is extremely likely that his Central Apnea would become profoundly worse. Hi Bonjour, Thanks for replying! I disagree with your conclusions. Because if he followed the steps I laid out he would start at very low pressures and post his results every day or so here where, from what I have seen in other threads, he would see what is happening at each step of his titration and adjust, you and others would coach him at each step of his titration and he would never enter the danger zones you are describing. The initial plan / framework above is just that .. a starting place and roadmap based on common tritation methods in the Clinical Tritation guides/manuals. There is an old "No battle plan survives contact with the enemy." Each day he studies and posts his OSCAR he and those coaching him will respond to facts being learned in those OSCAR reports and soon be completely ignoring and forgetting about what I wrote above, .. unless it actually ends up better for him than anything else he or we come up with. A few key thoughts on my mind:
Soft Cervical Collar Organize Charts Attaching Charts Dealing with a DME WillSleep INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT. |
« Next Oldest | Next Newest »
|
Possibly Related Threads... | |||||
Thread | Author | Replies | Views | Last Post | |
airsense 10/11 aircurve 10 heated tube only | eok361 | 0 | 110 |
11-19-2024, 11:30 PM Last Post: eok361 |
|
New Bi Level AC/11 VAUTO | skcampbell2 | 20 | 1,315 |
11-17-2024, 05:23 PM Last Post: skcampbell2 |
|
Help with ASVAuto ResMed AirCurve 10 settings (and esophageal pain) | RowG | 2 | 155 |
11-17-2024, 08:34 AM Last Post: RowG |
|
[Equipment] AirCurve 11 vs AirSense 11? | UKFan643 | 3 | 280 |
11-12-2024, 11:02 AM Last Post: sllama |
|
resmed as10, as11 aircurve 10 daylight savings time end | eok361 | 10 | 487 |
11-09-2024, 08:05 PM Last Post: specialed2000 |
|
[Mod Note] Resmed AirCurve 11 Users | Crimson Nape | 0 | 177 |
11-09-2024, 06:38 PM Last Post: Crimson Nape |
|
>> Attention: Resmed AirCurve 11 Users << | Crimson Nape | 2 | 262 |
11-09-2024, 01:25 PM Last Post: sauerkraut |