(01-02-2017, 03:23 PM)mandali44 Wrote: I couldn't figure out how to change the vertical scale.
Hi mandali44,
I use a different program than SleepyHead and don't remember using SH how to set the vertical scale on the Flow. I would think the instructions would be in RobySue's great beginner's guide:
http://www.apneaboard.com/wiki/index.php...epyHead%20
You are having too many hypopneas. There may be a Central component to these hypopneas but I think they look predominantly obstructive in type. The stop & go jumpiness in the Flow waveform (intermittently jumping to zero and back) is an indication of intermittent obstruction.
ASV machines nearly always eliminate central hypopneas and central apneas, so your apnea and hypopnea events will usually be predominantly obstructive in type.
I suggest two things, increasing the Min EPAP by 2 (to 10) and increasing Min Pressure Support (Min PS) by 2 (to 2).
EPAP (Expiration Positive Airway Pressure) is the Mask Pressure during exhalation, and Pressure Support is the boost in pressure added while you are inhaling, helping you to inhale. Your pressure while inhaling (IPAP) is always EPAP + PS. Your machine very slowly adjusts EPAP throughout the night to minimize obstructive events and nearly instantly increases PS if you suddenly stop breathing or reduce breathing effort.
Setting the Min PS to 1 or 2 will let the machine always be doing for you at least a small portion of the work of breathing. When the machine needs to be doing all the work of breathing for you (such as during a central apnea) it will automatically quickly raise PS to around 10 or perhaps higher.
In general Apnea Board recommends making adjustments gradually, changing only one thing at a time and waiting a week or two or longer between adjustments, so the effect of each adjustment can be seen.
Increasing Min EPAP is more important, so I suggest the Min EPAP be increased to 9 for a week and then to 10 for a week. Then the Min PS increased to 1 for a week and then to 2 for another week. If you can ask your doctor to ask your Durable Medical Equipment (DME) provider to do it for you, that would have the benefit of involving your provider.
Some doctors do not take the time to look at the Flow waveforms (do not think it is necessary) and might not be experienced in interpreting Flow waveforms. I have no idea whether that may turn out to be the case for you, but if it does, in my view that would be a good reason to search for a more proactive doctor.
Take care,