Aircurve 10 ASV
I received my AirCurve 10 ASV and in the first attempt to use it, the exhale back-off response is very slow in comparison to my old machine (Phillips SysOne)
For example with my old machine, the inhalation pressure backs off immediately as I start to exhale, but with the Air Curve it takes 1-2 seconds, which when it happens over and over I start to feel like I'm suffocating.
Is this long response time normal for this device?
P.S. I know I can adjust the ramp, but presumably this will wake me up once that period is done.
Thanks for any help.
4 hours ago
(This post was last modified: 4 hours ago by SarcasticDave94.)
RE: Aircurve 10 ASV
I found Ramp to be a hindrance while using the ASV for 2 years. The ResMed ASV has the following settings:
Mode: CPAP ASV ASV Auto
CPAP a static single pressure
EPAP either as single pressure in ASV or a Min and Max range in ASV Auto
PS a range in both ASV and ASV Auto, must use a range minimum of 5
IPAP is the sum of EPAP and PS at the moment, will always be a range, isn't manually set, caps at 25 Max
Then there's the common Ramp, humidity, heated hose, smart start, etc.
Note there is zero manual timing controls. No Ti Min and Max, no Trigger, cycle, etc.
Given this, yes you'll want to post OSCAR data so we can guide on what setting edits might help.
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
3 hours ago
(This post was last modified: 3 hours ago by kbjames. Edited 1 time in total.)
RE: Aircurve 10 ASV
Thanks for the responses all.
I think at a minimum i'll need to check the settings and set it to ASVAuto, as I believe its currently in standard ASV.
Right now EPAP is set to 15 and it feels too strong. PS Min is 6, PS Max 9. Any recommendations on what range I should try first in ASV Auto mode?
Its also fairly loud when i'm inhaling, any recommendations there to make it quieter?
P.S. I'm not setup with OSCAR yet, but that is on my to do list. Been busy otherwise, so trying to get this stuff going.
Thanks
2 hours ago
(This post was last modified: 2 hours ago by SarcasticDave94. Edited 2 times in total.
Edit Reason: Edit
)
RE: Aircurve 10 ASV
The EPAP and PS you have are pretty much opposite what is typical.
Lacking ASV Auto mode is going to hinder, because that EPAP is static. 15 EPAP, is that what you need in some way?
PS Min 6 is also too high for the typical need.
Is it possible these settings were like a copy and paste from your older Phillip Respironics System One?
What EPAP do you require?
Consider this edit if you want...
ASV Auto mode
EPAP 8-15
PS 3-15
IPAP will not be seen within settings edit area, but will be the sum of EPAP and PS. Example, my suggested edit makes the IPAP range 11-30. Note the ResMed ASV Max IPAP is 25 though.
Cheat codes for clinical menu
Hold both Home and click the Dial together for 5 seconds.
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.