Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Where do I begin? [AirCurve 10 ST to Aircurve 10 ASV ]
#21
RE: Where do I begin? [Help with AirCurve 10 ST]
Here are the nights results with Gideon’s recommended setting of EPAP @ 13.0


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#22
RE: Where do I begin? [Help with AirCurve 10 ST]
Any info if you felt better with this? Better comfort, rest, so on?
Mask Primer

Positional Apnea

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#23
RE: Where do I begin? [Help with AirCurve 10 ST]
This is so much better than your PS of 2 from earlier.
It is very important to note how you feel as Dave mentioed above.
See the cluster of events at about 0730 and the smaller one at about 0650? These normally indicate a positional apnea but because of the impovement I see from before I'd like to reduce EPAP by 1 again going to PS of 5 to see if we get additional improvement first.
Starting with the PS4 data and gping forward I'd like to see a couple of 5 minute views including at least one with UA in them
Post Reply Post Reply
#24
RE: Where do I begin? [Help with AirCurve 10 ST]
Mode: ST
IPAP: 17.0
EPAP: 13.0
Resp Rate: 10.0
Ti Max: 4.0
Ti Min: 0.3
Rise Time: Min
Trigger: Med
Cycle: Med
Mask: ResMed Mirage Quattro
Ramp Time: Off
Climate Ctrl: Auto
Temp Tube: 80 F
Humidity Level: 4

Did not post of how I felt because the night before was a terrible night. However, feelings from last night were better rested but, still woke with headache and brain fog. I will reduce the EPAP by 1 for a PS of 5 and see how that works. I am struggling with the machine going from IPAP to EPAP and vice versa until I can drift off. It feels as though timing is a bit off. The machine gives IPAP while just before I am ready to inhale and gives EPAP while I am still inhaling, rather annoying.


Attached Files Thumbnail(s)
           
Post Reply Post Reply
#25
RE: Where do I begin? [Help with AirCurve 10 ST]
The purpose of the Trigger setting is to adjust the initiation of the switch from exhaling to inhaling.  
Cycle is the transition from Inhale to exhale.  
One change at a time, But I think you can do this while relaxed but awake.  
Your short segments are heavily flow-limited.  So I'll ask do you have any, other than apnea, pulmonary or breathing conditions?
Go to PS 5 by Increasing the max by 2 and the min by 1.

I think you want to try Cycle at low.


   
Post Reply Post Reply
#26
RE: Where do I begin? [Help with AirCurve 10 ST]
Not aware of anything other than apnea. Back in 1994 I was diagnosed with 10 apneas and hypopneas observed in sleep study. They were characterized as: 1 central apnea, 2 mixed apneas, and 7 hypopneas and that started me on CPAP at 8 cm of water with nasal mask. Before I left CPAP, on Oct 15, 2021, I was on an auto set with pressures at 12 and 18 cm of water with an averaged 14~15 cm of water thru the night with full face mask.
 
Go to PS 5 by Increasing the max by 2 and the min by 1.
Do you mean IPAP=17.0 EPAP=12.0 or did you mean IPAP=19 EPAP=11.0
I will try Cycle at low for tonight.
Post Reply Post Reply
#27
RE: Where do I begin? [Help with AirCurve 10 ST]
You are at
IPAP=17
EPAP=13
PS=4

go to
IPAP=19
EPAP=14
PS=5
Post Reply Post Reply
#28
RE: Where do I begin? [Help with AirCurve 10 ST]
(10-29-2021, 10:32 PM)Bird Brain Wrote: The first night I was bombarded with constant IPAP/EPAP pressures. Felt like pressure oscillations could not get a relaxed breath or sleep. I have been unsuccessful with his prescription. 

To be able to get some sleep, I changed the settings as follows.
Mode: ST
IPAP: 16.0
EPAP: 14.0
Resp Rate: 8.0
Ti Max: 4.0
Ti Min: 0.3
Rise Time: Min
Trigger: Med
Cycle: Med
Mask: ResMed Mirage Quattro
Ramp Time: Off
Climate Ctrl: Auto
Temp Tube: 80 F
Humidity Level: 4

If I try to raise the Resp Rate above 8.0 it feels like the machine wants me to exhale before I finish inhaling or it wants me to inhale before I am finished exhaling so it is left on 8.0
 

I have not finished reading everyone else's response , so forgive me if I am repeating some of what others have suggested (its actually a good thing if they have).
I am not sure why the machine even needs to know your "Respiratory rate" because it is clearly out of synch with your normal RR at 8, so if there is a way to shut this parameter down, it would be worth a try.
As for prescription EPAP pressure of 16 which you later changed to 14, in my opinion this is way too high. I am several years younger than you and have no other physiological issues other than CSA and I can't tolerate an EPAP pressure above the bare minimum of only 4 because of the conscious effort that I must put into exhalation at pressures higher than 4. High EPAP pressures are akin to sticking my head out the window of a speeding car at 200 MPH, trying to exhale...It is difficult and interferes with my sleep....So I am wondering if you are open to dropping your EPAP setting to a minimum and see how it impacts your sleep and then raise EPPAP (slowly, perhaps 1 CMH2O per night) if your AHI number is high.
Ipap of 16 for a bilevel machine is in my opinion also way too high, so dropping it down to 12 and seeing how you feel should help you with a better sleep. Of course if lowering IPAP makes you feel like you are gasping for more air at inhalation, then you can increase it by 1 at a time until you find your own sweet spot.
Post Reply Post Reply
#29
RE: Where do I begin? [Help with AirCurve 10 ST]
Thanks Gideon and S. Manz for input. I will try Gideon's recommendation tonight but I am interested in S. Manz suggestions also. I am so wanting to get a good night’s rest I think I am getting desperate.
 
S. Manz the Doc changed through the DME as the following
Mode: ST
IPAP: 18.0
EPAP: 4.0
Resp Rate: 10.0
Ti Max: 3.0
Ti Min: 0.3
Rise Time: Min
Trigger: Med
Cycle: Low
Mask: ResMed Mirage Quattro
Ramp Time: Off
Climate Ctrl: Auto
Temp Tube: 80 F
Humidity Level: 4
 
The change between IPAP and EPAP was terrible it felt like a hammer each time the pressures would change. I tried to stay with these settings till I would hopefully pass out but no luck. That is when I had a terrible night, worst night since on BIPAP.
 
I change the settings as follows and give it a try.
 
Mode: ST
IPAP: 19.0
EPAP: 14.0
Resp Rate: 10.0
Ti Max: 4.0
Ti Min: 0.3
Rise Time: Min
Trigger: Med
Cycle: Low
Mask: ResMed Mirage Quattro
Ramp Time: Off
Climate Ctrl: Auto
Temp Tube: 80 F
Humidity Level: 4
 
Sorry Resp. Rate has been @ 10.0 since Docs setting change 11/04/2021 and I have not changed it back.
Post Reply Post Reply
#30
RE: Where do I begin? [Help with AirCurve 10 ST]
Ok so I read a bit more about BIPAP machines and can now see why there is a "Respiratory Rate" setting. The machine internal algorithm uses this RR to calculate when to raise pressure (for IPAP or inhalation) and then drop pressure (for EPAP or exhalation) which is kind of strange because everyone's RR changes while they are asleep .
Normally people take between 12 to 15 breaths per minute which can drop as low as 6 or 8 for some (such as myself) during sleep so how a BIPAP machine compensates for this change in rate is still a mystery to me ( I use a ASV machine which automatically monitors my respiratory rate and adjusts), so I am still not sure how to deal with this setting on your machine.

As for EPAP pressure, I see that your doctor dropped it to the lowest setting at 4  but then you didn't like it because the IPAP setting was set too high and the rate of change in pressure with each breath, kept you up ( "felt like a hammer").
This is why I think that lowering the IPAP pressure as well as the EPAP is a good idea to try.
Let's say EPAP at 4 and IPAP at only 8 to start and see if your sleeping improves.
For now if there is a setting that simply stops the BILEVEL feature and sets up "AUTO" feature instead, it would be worth a try,
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Algorithm change in the ResMed AirCurve 10 ASV terp1984 14 926 3 hours ago
Last Post: Sleeprider
  Aircurve 11 vAuto issues Sphinxx729 7 166 Yesterday, 01:08 PM
Last Post: Sphinxx729
  Aircurve VAuto PS...pros & cons WW4B 13 610 12-26-2024, 12:28 PM
Last Post: Sleeprider
  Aircurve 10 VAuto settings question dec3169 6 248 12-24-2024, 01:25 PM
Last Post: Deborah K.
  AirCurve ASV vs Vauto (is ASV king of the hill?) Carnal 3 226 12-23-2024, 12:27 PM
Last Post: Deborah K.
  Air leaks with Aircurve 11 ASV for CA chrishark 13 638 12-20-2024, 08:25 AM
Last Post: Sleeprider
  [Treatment] BIPAP Settings - ResMed AirCurve 11 Vauto dcbauman 20 983 12-17-2024, 03:42 PM
Last Post: Sleeprider


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.