Mal that is the issue I was pointing out. When you look at the zoomed in portion you can see exactly what is happening once you understand how this machine works. Instead of getting one normal breath the machine is forcing you to take 2 short quick breaths with a quick exhalation in between. If you tried to replicate that it would be fast almost hyperventilation type breathing.
With the autoset APAP you were using before everything relies on your spontaneous breathing. Your spontaneous breathing much of the night has strange oscillations that are not normal (have seen them in a few other members but not many and not near as often as yours appear). It is hard to know what causes that breath waveform but in effect what is happening is something is causing you to exhale when you should be inhaling and inhale when you should be exhaling. Random guesses on my part include stuff like diaphragm/chest muscles not in sync(asynchronous breathing), diaphragm spasming, amplified cardiogenic oscillations. This is why I was hoping a pulmonologist would review that and it frustrates me to no end how these guys refuse to even look at OSCAR data. Perhaps the only way to get them to review the data is to give them your CPAP SD card data and tell them they have to look at the flow rate data however they are capable of doing so if they aren't going to take your word for it. I also thought that your titration study would have commented on this but it doesn't seem that they did, I don't know if that is because it didn't happen or because the mode and settings they used were able to mitigate it. One thing that you could do that might help understand what is happening is setting up a camera to record yourself sleep and perhaps that would give some hints.
With the ST-A you have timing controls which are manipulating these strange oscillating breaths into the short breaths you see now. I highly recommend you review the clinical manual and what I post here to make sure you understand these controls as I believe they are very important in your situation. If there is something you do not understand about them please ask. The first control that comes into play is trigger sensitivity. The higher the trigger sensitivity the sooner the machine will start to supply pressure. Next is rise time, the shorter the rise time the faster the machine will build pressure. Then Timin which will hold pressure on for a set time (Timin) regardless of what your spontaneous effort during that time. Then you have cycle sensitivity, higher the sensitivity the sooner the machine will cycle back to lower pressure (EPAP). Then you have Timax, if you don't spontaneously start exhaling before Timax it will lower the pressure to try and force you to exhale.
The issue here is that these timing controls are started by spontaneous effort. When the first oscillation in your breath prematurely passes the 0 flow line then IPAP pressure is being triggered (due to trigger sensitivity), it is raising pressure faster than your old machine (because of rise time) and it is maintaining pressure for a minimum time (Timin) effectively forcing you to take a breath. If I had to guess most of the short breaths you posted end as soon or shortly after Timin is over then your body takes a quick exhale (because it didn't even get to finish exhalation process before getting pumped up again) and then the whole process repeats. During this whole time the machine is pretty much forcing you to breath at a rate that does not match your spontaneous effort and everything is out of sync and other than providing more ventilation I am not sure it is a better result than what you had on APAP.
Here I show this in pictorial version (what is happening and what I would like to try and make happen). This is one of your examples of breathing from APAP.
Trying to fight this with PS would make sense if they were flow limitations but I don't believe that they are because flow limitations restrict flow, not change its direction. Increasing PS could still work but if we are fighting chest or diaphragm etc I don't know that it will help do anything other than create more pressure and more ventilation in the same form of breathing.
If we could use timed breath mode in conjunction with iVAPS mode that is what I would have already proposed as I believe it might have merit with this situation. I do like the advantage of iVAPS over timed mode because your sleep study shows not only a comment about hypoventilation but also data of dropping O2 levels because of it. iVAPS allows us to use an oximeter to monitor O2 levels and increase target Va until desats are no longer present which was going to be my recommendation after we get the basics working properly.
And this isn't just based on my opinions(other than my opinion that you have a synchronicity problem). From the Resmed titration protocol.
Quote:Adjust TiControls and Synchrony features if • Chest wall movement is not in sync with mask pressure tracing
Quote:For SpO2 < 90% with all respiratory events eliminated: • Increase Target Va by 0.3 every ≥ 5 min until desaturations are resolved
Quote:When to adjust the trigger sensitivity threshold? The Medium (default) setting will be ideal for most patients. A Low (or Very Low) trigger sensitivity setting is recommended for the following conditions: • Cardiogenic oscillations and subsequent auto-triggering
I am sure trigger sensitivity should be decreased. I am not so sure about Ticontrols and rise time and feel we need to adjust trigger sensitivity first to figure out ideal settings.
Also if I haven't stated this yet, I have a hard time believing this breathing is due to sleep apnea or airway restriction. You do have other central apnea issues primarily during sleep transition, I'm not sure if they are related though.