REMstar error codes
REMstar PAP machines will present an error code on the screen when there is an error.
It will sometimes have audible beeping tones in various rhythms to indicate problems. Refer to the "see also" section of this page for more information regarding the tones.
If the machine only says "Service Required", it is usually a fatal error. This will happen with water damage that has caused a short circuit.
Below are the error codes for the REMstar xPAP machines. Please note: these codes are for older pre-2007 REMstar machines. For the newer PR System One machines, go to Philips-Respironics error codes.
E-0 - E-20
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-0 | NO_ERROR | Reserved | N/A |
E-1 | ERR_SOFTWARE | Reserved | Replace Main PCA |
E-2 | ERR_PROGRAM_CRC_FAILURE | Defective FLASH chip | Replace Main PCA |
E-3 | ERR_EXTERNAL_RAM_FAILURE | Defective RAM chip | Replace Main PCA |
E-4 | ERR_FIQ_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-5 | ERR_FIQ_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-6 | ERR_NEST_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-7 | ERR_NEST_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-8 | ERR_IRQ_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-9 | ERR_IRQ_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-10 | ERR_TMR_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-11 | ERR_TMR_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-12 | ERR_SVC_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-13 | ERR_SVC_STACK_UNDERFLOW | Defective RAM chip | Replace Main PCA |
E-14 | ERR_THREAD_STACK_OVERFLOW | Defective RAM chip | Replace Main PCA |
E-15 | ERR_ARM_UNDEFINED_INSTRUCTION_EXCEPTION | Defective FLASH chip | Replace Main PCA |
E-16 | ERR_ARM_SWI_EXCEPTION | Defective FLASH chip | Replace Main PCA |
E-17 | ERR_ARM_PREFETCH_EXCEPTION | Defective FLASH chip | Replace Main PCA |
E-18 | ERR_ARM_ABORT_EXCEPTION | Defective RAM chip | Replace Main PCA |
E-19 | ERR_ARM_RESERVED_EXCEPTION | Defective FLASH chip Defective microprocessor |
Replace Main PCA |
E-20 | ERR_AIC_DEFAULT_HANDLER | Defective FLASH chip | Replace Main PCA |
E-21 - E-40
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-21 | ERR_AIC_SPURIOUS_HANDLER | Defective FLASH or RAM chip | Replace Main PCA |
E-22 | ERR_CAL_DATA_CRC_FAILURE | Improper calibration Defective FLASH chip |
Recalibrate Replace Main PCA |
E-23 | ERR_INVALID_DRIFT_INDEX_AT_CAL | Improper calibration Occluded or restricted pressure tube Defective flow sensor |
Recalibrate Replace pressure tube Replace Main PCA |
E-24 | ERR_UNRECOGNIZED_CAL_VERSION_BY_SW | Improper calibration Defective FLASH chip |
Recalibrate Replace Main PCA |
E-25 | ERR_DRIFT_FLOW_TOO_HIGH | Occluded or restricted pressure tube Defective flow sensor |
Replace Flow Tube Replace Main PCA |
E-26 | ERR_UI_DISPLAY_FAILURE | Defective LCD driver chip | Replace Main PCA |
E-27 | ERR_DRIFT_PRESSURE_TOO_HIGH | Occluded or restricted pressure tube Defective pressure sensors |
Replace pressure tubing Replace Main PCA |
E-28 | ERR_CAL_DATA_BLANK | Improper calibration Defective FLASH chip |
Recalibrate Replace Main PCA |
E-29 | ERR_DRIFT_SECONDARY_PRESSURE_TOO_HIGH | Occluded or restricted pressure tube Defective pressure sensors |
Replace pressure tubing Replace Main PCA |
E-30 | ERR_INVALID_NVDATA_SCHEMA | Normal for first time power up Defective EEPROM |
Remove & restore power Replace Main PCA |
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-31 | ERR_INVALID_NVDATA_LOCK | Defective EEPROM | Replace Main PCA |
E-32 | ERR_INVALID_NVDATA_CHECKSUM | Normal for first time power up Defective EEPROM |
Remove & restore power Replace Main PCA |
E-33 | ERR_NVDATA_STORAGE_ERROR | Defective EEPROM | Replace Main PCA |
E-34 | ERR_NVDATA_QUEUE_FULL | Defective EEPROM | Replace Main PCA |
E-35 | ERR_NVDATA_READBACK_FAILURE | Defective EEPROM | Replace Main PCA |
E-36 | ERR_NVDATA_RANGE_FAILURE | Defective EEPROM | Replace Main PCA |
E-37 | ERR_CORRUPT_RTC_VALUE | Defective microprocessor | Replace Main PCA |
E-38 | ERR_RTC_NOT_TICKING | Defective microprocessor Defective crystal |
Replace Main PCA Replace Main PCA |
E-39 | ERR_UI_QUEUE_FULL | Defective PCA | Replace Main PCA |
E-40 | ERR_INVALID_BIST_TEST_CALL | Invalid RASP command | Reissue RASP command |
E-41 - E-60
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-41 | ERR_BIST_THREAD_LOCKED | Defective PCA | Replace Main PCA |
E-42 | ERR_ALARM_VOLTAGE_FAILURE | Defective power supply | Replace Main PCA |
E-43 | ERR_12VREF_FAILURE | Defective power supply | Replace Main PCA |
E-44 | ERR_5VREF_FAILURE | Defective power supply | Replace Main PCA |
E-45 | ERR_VBULK_FAILURE | Defective power supply | Replace Main PCA |
E-46 | ERR_NEG15VREF_FAILURE | Defective power supply | Replace Main PCA |
E-47 | ERR_FL12_SENSOR_RAILED | Occluded or restricted pressure tube Defective flow sensor |
Replace flow tube Replace Main PCA |
E-48 | ERR_PV2_SENSOR_RAILED | Occluded or restricted pressure tube Defective pressure sensor |
Replace pressure tube Replace Main PCA |
E-49 | ERR_FLASH_FAILURE | Defective FLASH chip | Replace Main PCA |
E-50 | ERR_UNRECOGNIZED_BOARD_REV | Needs new software Defective PCA |
Replace Main PCA Replace Main PCA |
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-51 | ERR_MOTOR_FAILURE | Defective motor drive chip | Replace Main PCA |
E-52 | ERR_BLOWER_RPM_FAILED_TOLERANCE | Stalled or locked blower | Replace Blower Replace Main PCA |
E-53 | ERR_BLOWER_OFF_CURRENT_HIGH | Stalled or locked motor Defective motor drive chip |
Replace Blower Replace Main PCA |
E-54 | ERR_BLOWER_ON_CURRENT_HIGH | Stalled or locked motor Defective motor drive chip |
Replace Blower Replace Main PCA |
E-55 | ERR_PRINTF_BUFFER_OVERFLOW | Defective PCA | Replace Main PCA |
E-56 | ERR_THERAPY_EVENT_QUEUE_FULL | Defective PCA | Replace Main PCA |
E-57 | ERR_KEYPAD_STUCK_KEY | Verify all keys protrude through top Defective Keypad Defective pad on PCA |
Align keys with top Replace Keypad Replace Main PCA |
E-58 | ERR_UNUSED_58 | Reserved | Reserved |
E-59 | ERR_UNUSED_59 | Reserved | Reserved |
E-60 | ERR_UNUSED_59 | Reserved | Reserved |
E-61 - E-80
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-61 | ERR_UNUSED_61 | Reserved | Reserved |
E-62 | ERR_UNUSED_62 | Reserved | Reserved |
E-63 | ERR_WATCHDOG_FAILURE | Defective microprocessor | Replace Main PCA |
E-64 | ERR_UNEXPECTED_WATCHDOG | Defective FLASH chip Defective RAM chip |
Replace Main PCA Replace Main PCA |
E-65 | ERR_UNUSED_65 | Reserved | Reserved |
E-66 | ERR_RTOS_INIT_FAILURE | Defective PCA | Replace Main PCA |
E-67 | ERR_SERIAL_QUEUE_FULL | Defective PCA | Replace Main PCA |
E-68 | ERR_SNORE_SAMPLING_THREAD_LOCKED | Defective PCA | Replace Main PCA |
E-69 | ERR_MAIN_EXECUTION_THREAD_LOCKED | Defective PCA | Replace Main PCA |
E-70 | ERR_INTERNAL_RAM_FAILURE | Defective microprocessor | Replace Main PCA |
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-71 | ERR_EVENT_LOG_QUEUE_FULL | Defective PCA | Replace Main PCA |
E-72 | ERR_TREND_DIV_BY_ZERO | Defective RAM chip | Replace Main PCA |
E-73 | ERR_ROLLOVER_CASE | Defective RAM chip | Replace Main PCA |
E-74 | ERR_WRONG_DISPLAY_TYPE | Defective PCA | Replace Main PCA |
E-75 | ERR_ALARM_CAPACITOR_FAILURE | Defective PCA | Replace Main PCA |
E-76 | ERR_OVER_PRESSURE | Occluded or restricted tube Defective PCA |
Replace Pressure tubing Replace Main PCA |
E-77 | ERR_UNDER_PRESSURE | Occluded or restricted tube Defective PCA |
Replace Pressure tubing Replace Main PCA |
E-78 | ERR_OVER_PRESSURE_REG | Occluded or restricted tube Defective PCA |
Replace Pressure tubing Replace Main PCA |
E-79 | ERR_UNDER_PRESSURE_REG | Occluded or restricted tube Defective PCA |
Replace Pressure tubing Replace Main PCA |
E-80 | ERR_ROTOR_LOCKED | Locked Motor Defective PCA |
Replace Motor Replace PCA |
E-80 - E-86
Error Code | Description | Probable Cause | Corrective Action |
---|---|---|---|
E-81 | ERR_MOTOR_EDGE_MISSED | Defective Motor Defective PCA |
Replace Motor Replace PCA |
E-82 | ERR_MOTOR_EDGE_EXTRA | Defective Motor Defective PCA |
Replace Motor Replace PCA |
E-83 | ERR_MOTOR_OVERSPEED | Defective Motor Defective PCA |
Replace Motor Replace PCA |
E-84 | ERR_MOTOR_BRAKE | Defective Motor Defective PCA |
Replace Motor Replace PCA |
E-85 | ERR_MOTOR_CURRENT | Defective Motor Defective PCA |
Replace Motor Replace PCA |
E-86 | ERR_INVALID_ERROR_CODE | Reserved | Reserved |
See Also
Philips-Respironics error codes (for newer, PR System One machines)
REMstar smart card error codes
For beeping errors (the machine will beep in patterns), see the PDF version of the user manual. Scroll to the "Device Alerts" section (page 12 of the manual, page 14 of the PDF).
http://respironicsremstars.respironics.com/PDF/REMstarAutoUserManual.pdf
Donate to Apnea Board