Support Apnea Board & OSCAR  

Changes

Jump to: navigation, search

REMstar error codes

2,188 bytes added, 20:52, 14 April 2012
E-61 - E-80
|}
==E-61 - E-80==
{| class="wikitable"
|-
! scope="col" width="75" align="center" | Error Code !! scope="col" width="325" | Description !! scope="col" width="225" | Probable Cause !! scope="col" width="225" | Corrective Action
|-
| E-61 || ERR_UNUSED_61 || Reserved || Reserved
|-
| E-62 || ERR_UNUSED_62 || Reserved || Reserved
|-
| E-63 || ERR_WATCHDOG_FAILURE || Defective microprocessor || Replace Main PCA
|-
| E-64 || ERR_UNEXPECTED_WATCHDOG || Defective FLASH chip<br />Defective RAM chip || Replace Main PCA<br />Replace Main PCA
|-
| E-65 || ERR_UNUSED_65 || Reserved || Reserved
|-
| E-66 || ERR_RTOS_INIT_FAILURE || Defective PCA || Replace Main PCA
|-
| E-67 || ERR_SERIAL_QUEUE_FULL || Defective PCA || Replace Main PCA
|-
| E-68 || ERR_SNORE_SAMPLING_THREAD_LOCKED || Defective PCA || Replace Main PCA
|-
| E-69 || ERR_MAIN_EXECUTION_THREAD_LOCKED || Defective PCA || Replace Main PCA
|-
| E-70 || ERR_INTERNAL_RAM_FAILURE || Defective microprocessor || Replace Main PCA
|}
 
{| class="wikitable"
|-
! scope="col" width="75" align="center" | Error Code !! scope="col" width="325" | Description !! scope="col" width="225" | Probable Cause !! scope="col" width="225" | Corrective Action
|-
| E-71 || ERR_EVENT_LOG_QUEUE_FULL || Defective PCA || Replace Main PCA
|-
| E-72 || ERR_TREND_DIV_BY_ZERO || Defective RAM chip || Replace Main PCA
|-
| E-73 || ERR_ROLLOVER_CASE || Defective RAM chip || Replace Main PCA
|-
| E-74 || ERR_WRONG_DISPLAY_TYPE || Defective PCA || Replace Main PCA
|-
| E-75 || ERR_ALARM_CAPACITOR_FAILURE || Defective PCA || Replace Main PCA
|-
| E-76 || ERR_OVER_PRESSURE || Occluded or restricted tube<br />Defective PCA || Replace Pressure tubing<br />Replace Main PCA
|-
| E-77 || ERR_UNDER_PRESSURE || Occluded or restricted tube<br />Defective PCA || Replace Pressure tubing<br />Replace Main PCA
|-
| E-78 || ERR_OVER_PRESSURE_REG || Occluded or restricted tube<br />Defective PCA || Replace Pressure tubing<br />Replace Main PCA
|-
| E-79 || ERR_UNDER_PRESSURE_REG || Occluded or restricted tube<br />Defective PCA || Replace Pressure tubing<br />Replace Main PCA
|-
| E-80 || ERR_ROTOR_LOCKED || Locked Motor<br />Defective PCA || Replace Motor<br />Replace PCA
|}
==E-80 - E-86==
101
edits



Donate to Apnea Board  

Navigation menu