AHI double with FFM than nasal mask
#1
Question 
AHI double with FFM than nasal mask
I am really curious about what is going on.  I had been using the Dreamwear nasal mask, and I've loved it.  But sometimes I need to breathe through my mouth here and there, so I recently got an Amara View FFM.  I used the Amara View 3 nights in a row, and then switched to the Dreamwear for 2 nights, and then switched to the Amara again last night. 

What I've found is that with the Amara View, my AHI is twice that with the Dreamwear.

Also, I found that my machine went into higher pressures with the Amara than with the Dreamwear. 

Is this normal?  Is it a personal thing for each person?  Or... is it something else?

Here are screenshots of my last 2 nights (1st is with Dreamwear, 2nd is with Amara View).  The previous 2 nights (where I also did 1 night Amara, 1 night Dreamwear) reflect this same pattern.

[Image: Tr1N3Nal.png]

[Image: DcXlJu6l.png]


I've seen you ask some people what some possible variants might be, so I'll put out some here for you.  This is at sea-level, same location, same sleep partner.  For the 2 nights prior to these 2 nights, I saw this same pattern, and I was at 4,000 ft altitude with no sleep partner.  Because of that consistency of what I can see as only the mask being the variant, I felt confident coming here and asking your thoughts on if masks make a difference in AHI.

P.S.  Having a sleep partner who interrupts my sleep is a bummer.  I kinda want to suggest that everyone (IRL) goes and gets a sleep study!
Post Reply Post Reply
#2
RE: AHI double with FFM than nasal mask
I noticed that changing between the Quattro Air and the Simplus resulted in a small increase in AHI and slightly higher pressures. These are both full face masks but the Simplus has a much larger volume than the Quattro. It would seem reasonable that a similar change would be evident when going between a nasal mask and an FFM.

In any case the increase in AHI is very small in absolute terms so I wouldn't worry too much about it. Your machine should be able to compensate.
Post Reply Post Reply
#3
RE: AHI double with FFM than nasal mask
I had the same experience. Even with low leak rates with the Amara View my AHI was averaging about 3. Immediately upon switching to the Dreamwear nasal mask my AHI dropped to under 1. My only theory is that the Amara was pushing slightly back on my jaw. In any event the Dreamwear is so much more pleasant to use even if the AHI hadn't dropped I would have stayed with the Dreamwear mask.
Post Reply Post Reply
#4
RE: AHI double with FFM than nasal mask
Hydrangea, the difference in pressure and AHI between your nasal and FFM are not that unusual. First, your pressure is very very low. Considering the volume difference between nasal pillows and a FFM, the mask pressure will be lower in a FFM, especially with EPR or Flex active. I would not hesitate to increase the minimum pressure 2-3 cm for FFM use. In your case, I would also increase the maximum pressure to 12. Also, are you changing the mask type or resistance setting when making the mask change? Finally, your AHI is under 2, even with FFM, so none of this is very urgent, but if you want similar results, you will need more minimum pressure when using the full face.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#5
RE: AHI double with FFM than nasal mask
My pressure ranged from 7-14 while using a resmed f10 FFM.

On a P10, I use a constant 8 pressure and have lower AHIs.
Post Reply Post Reply
#6
RE: AHI double with FFM than nasal mask
(02-07-2017, 10:45 AM)Sleeprider Wrote: Hydrangea, the difference in pressure and AHI between your nasal and FFM are not that unusual.  First, your pressure is very very low.  Considering the volume difference between nasal pillows and a FFM, the mask pressure will be lower in a FFM, especially with EPR or Flex active.  I would not hesitate to increase the minimum pressure 2-3 cm for FFM use.  In your case, I would also increase the maximum pressure to 12.  Also, are you changing the mask type or resistance setting when making the mask change?   Finally, your AHI is under 2, even with FFM, so none of this is very urgent, but if you want similar results, you will need more minimum pressure when using the full face.

Thank you all for your input.  I was in the midst of chaos from moving, selling houses, etc., so I apologize for not responding.  But your responses were all very helpful to me... and I've been pondering them ever since.

Sleeprider - I've not changed the mask type or resistance setting because (at least according to the PR info I've found) both the Amara View and the DreamWear are an x1 for the DreamStation.  Honestly, that does not make sense to me, as (like others mentioned here) the volume in the two seems to be quite different.  But, who knows?!  So, no, I've not changed that when switching masks.  Would you advise me to?  If so, what should I change it to, and for which mask?  I have my DreamStation set on x1.

Since reading your response to my thread, I did widen the range of my auto pressure (I changed it to 7-15).  Now, when I use the Amara View, the 95% pressure is between 12-15.  When I use the DreamWear, the 95% is between 7.5-9.

My follow-up questions are these:
- Should I widen the auto pressure any further? 
- Should I change the pressure settings each time I change from DreamWear to Amara View and vice versa?  (ie: when I use the DreamWear, set it at 6-11; when I use Amara View, set it at 10-17... or... something else?)
Post Reply Post Reply
#7
RE: AHI double with FFM than nasal mask
Well, it's been a month since my last response, and you've come a long way since. I always used non-resprionics masks, so left resistance at 0 and never thought much of it. Go with the recommendation or nothing, and I think it won't make much difference. Your experience of different pressure needs for nasal vs FFM isn't unusual. As long as your machine can keep up, no problem. If you have higher events with the FFM, then you know to increase minimum pressure by 2 or 3 cm.

Your pressure selections seem to make sense to me, and you've been around long enough to figure it out. The only question is why do you change masks?
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#8
RE: AHI double with FFM than nasal mask
(03-09-2017, 04:54 PM)Sleeprider Wrote:  The only question is why do you change masks?

Some nights I just cannot breathe through my nose. 

Sometimes my nose is just too stuffy, due to environmental things. But I'm learning that the humidity from the mask will often help things loosen up, and it's ok. 

One night I drank far too much alcohol to be able to keep control of my breathing and keeping my mouth closed, but I still wanted the therapy, so the FFM made that much easier. 

And as I've been experimenting with the various sizes of DreamWear cushions, I've realized I get a much better flow of air from the wider slits in the size Medium Wide. But the Medium fits my cheeks better.  So when I obtain some extra Medium cushions, I'm going to experiment with cutting a longer slit in them.

Thank you for the validation and the tips. I very much appreciate it.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Leaks and mask displacement for side-to-stomach sleepers. Tape down nasal pillows? Bostonsnow 13 1,805 12-15-2024, 05:16 AM
Last Post: 2SleepBetta
  [Equipment] F&P Nova Micro nasal pillows mask carnuntum 4 1,144 10-26-2024, 02:48 AM
Last Post: carnuntum
  Nasal pillows v nasal mask FLcracker 11 971 10-08-2024, 11:42 AM
Last Post: mogulman
Ohmy [Treatment] AHI skyrockets when going from double to single pillow?! Cpapaya 5 496 09-25-2024, 02:10 PM
Last Post: SeePak
  Dreamwear nasal pillows vs. full face mask arnoblalam 1 376 09-22-2024, 10:06 AM
Last Post: Nightynite
  Full Face Mask Issues vs Nasal Pillows Issues FozzyDaBear 8 598 09-17-2024, 03:58 PM
Last Post: FozzyDaBear
  Nasal sores and humidifier. ubuntootoo 9 880 09-07-2024, 06:25 PM
Last Post: Deborah K.


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.