In the resmed machine, you can see the pressures rose faster and resulted in a median pressure of 10.4. There are some leaks in this data and again, I think your mask fit will be improved if the minimum pressure is increased to at least 7. The cluster of OA at 6:20 began before the machine could ramp up pressure, and that again points to higher minimum pressure.
Overall you are doing better in terms of AHI with the faster responding Resmed machine. Both units would treat you better with higher minimum pressure, and this is more important with the slower responding Philips Respironics. When judging comfort, you need to take this higher minimum pressure requirement for the Philps into account. The Resmed will prevent your events from a pressure of 7, I don't think the Philips will until the minimum is at 9.
Apnea Board Moderator
www.ApneaBoard.com
____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.