Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Any way to tackle the remaining CAs?(AutoSet for Her)
#11
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
(07-08-2020, 08:51 PM)bonjour Wrote: Sleeprider and I are on the same page.  The zoomed views validate that you have fairly extensive flow limitations and once again flow limits are much more extensive than the stats indicate.  I was going to suggest starting at EPR=2 because the flow limits are so extensive.  Follow SR's settings and let's see what happens.  Expected is a decrease in flow limits and an increase in Central Apneas, what we don't know is how much, CAI may not increase.

ResMed measures the 'flatness' of the breath and presents an index of 0 to 1 with higher being the more significant.

Should I also switch to AutoSet Standard?
Post Reply Post Reply
#12
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
Let's see what we get before we look at the ASV route.

and yes, Autoset standard

The Rule of thumb is that you must  fail CPAP, fail BiLevel without backup, fail BiLevel with Backup, before you get a chance at ASV.  
We are ASV proponents so don't worry,  we will send you down that path if we feel it is the correct one.
Post Reply Post Reply
#13
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
The difference between Autoset Standard and For Her is the rate at which pressure increases are applied. For Her reacts faster to the first flow limit, then slows down, while Standard will continue to respond.  In the interest of changing only EPR for now, feel free to stay with Autoset for her, and we can try the mode change later to see what affect it has. The idea of adding EPR is to prevent the flow limits from even being an issue. Your exhale pressure will be lower, but your inhale is assisted by the pressure support of an inhale pressure that is higher than exhale.

[Image: AutoSet_for_her.png]
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#14
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
(07-08-2020, 09:41 PM)Sleeprider Wrote: The difference between Autoset Standard and For Her is the rate at which pressure increases are applied. For Her reacts faster to the first flow limit, then slows down, while Standard will continue to respond.  In the interest of changing only EPR for now, feel free to stay with Autoset for her, and we can try the mode change later to see what affect it has.  The idea of adding EPR is to prevent the flow limits from even being an issue. Your exhale pressure will be lower, but your inhale is assisted by the pressure support of an inhale pressure that is higher than exhale.  

[Image: AutoSet_for_her.png]

I understand turning on EPR to tackle FL part. But I am still confused about the rationale regarding switching to AutoSet Standard.
Post Reply Post Reply
#15
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
It is pressure support that best treats flow limits, as you see the for her provides an initially quicker, but a slower and lessor response overall.

We are not against it if it works, but we are looking to attack the flow limits as hard as we can. Again, if you want to try the for her as a step in the process, neither one of us has an issue with what works, even if we haven't done it that way before.
Post Reply Post Reply
#16
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
(07-08-2020, 10:46 PM)bonjour Wrote: It is pressure support that best treats flow limits, as you see the for her provides an initially quicker, but a slower and lessor response overall.

We are not against it if it works, but we are looking to attack the flow limits as hard as we can.  Again, if you want to try the for her as a step in the process, neither one of us has an issue with what works, even if we haven't done it that way before.

I heeded you guys' advice––The flow limitations were reduced a lot, the CA remained about the same, and I feel better. Any further refinements can be suggest? How to reduce CAs or even further reduce flow limitations?


Attached Files Thumbnail(s)
       
Post Reply Post Reply
#17
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
Be aware you cannot treat CA with any PAP machine except an ASV. You can avoid some with your APAP.
Mask Primer

Positional Apnea

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#18
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
The numbers are still acceptable, yes the central apneas too.

I would like you to try EPR = 3 tonight and likely for several days.

Very important is how you feel, Which is better EPR=2 or EPR = 3?

We are looking for a balance between obstructive and central apneas where you feel best.
Post Reply Post Reply
#19
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
(07-09-2020, 09:40 AM)SarcasticDave94 Wrote: Be aware you cannot treat CA with any PAP machine except an ASV. You can avoid some with your APAP
Should I reduce mini pressure and increase EPR to 3?
Post Reply Post Reply
#20
RE: Any way to tackle the remaining CAs?(AutoSet for Her)
The only change to make is EPR setting. Keep everything else the same. The change to EPR 3 will make your minimum EPAP pressure only 5.0 cm (8.0/5.0), but if we see a need to increase the minimum pressure from 8 to 9.0, we can make that call after the first night.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  should I use cpap pressure over autoset Person 35 2,213 11-05-2024, 12:51 PM
Last Post: Dormeo
  Airsense 10 Autoset Too Much Pressure? MiRock72 14 881 10-16-2024, 08:48 PM
Last Post: MiRock72
  Larger reservoir for Air Sense 10 Autoset RonKMiller 33 25,550 10-12-2024, 07:07 AM
Last Post: Nosludhe
  [CPAP] New Airsense 10 Autoset - Help Optimize Settings Led Head 2 378 09-05-2024, 11:08 AM
Last Post: Led Head
  [Pressure] Resmed Airsense 10 Autoset Turpentine 6 497 08-28-2024, 05:41 PM
Last Post: G. Szabo
  Airsense 10 autoset not detecting breath? Camerongates79 12 832 07-11-2024, 12:04 AM
Last Post: HalfAsleep
  [CPAP] Autoset doesn't do much for me -- with one exception rrstv 3 409 06-17-2024, 09:20 PM
Last Post: Deborah K.


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.