RE: Resmed AirSense 10 have I set it up right ?
I won't be changing the ranges just yet need to get a more than 1 day in before I think of fine tuning my ranges.

In terms of the leaks could someone explain the results and the dreaded '24' ?
Post Reply Post Reply
#12
RE: Resmed AirSense 10 have I set it up right ?
the higher the leak rate, the more likely the machine will miss or wrongly characterize some of the apneas. sometime this results in unknown apneas (forest green). in addition, the higher the leakage, the more likely you have actually been awakened by blasts of air. finally, if the air is leaking out all over, the treatment is not as effective.

these results cause the "dread".

QAL
Dedicated to QALity sleep.
Post Reply Post Reply
#13
RE: Resmed AirSense 10 have I set it up right ?
Hi Simon,
Great first night. There will be up's and downs, keep your on on the "overall" and big picture as you go, it's what I try to do, and I'm further out than you at 34 nights, but your numbers "beat" mine (-: It gets very tempting to move things around too much in search of "better" but the truth is, we can't really know what will get us "better" until we have a few days to be sure that our one bad night was even related to anything we're doing.

I have the same machine as you do. I asked the Tech at the DME about the rainout I'd heard some folk have (water in hose that drips back down the hose into face) and she said "no way you'd get that with this setup" Because I have not only humidifier but also a heated hose. Well guess what I got for first two nights, pretty heavily? Rainout. She'd set the machine for temp/humidity and not Auto...well I switched it to auto and no more rainout. haven't seen it sense.

Don't know how that would affect your bits of water specks, doesn't sound like you had actualy rainout or anything much, but if you do, check your settings. I must say I do think my heated hose is wonderful, but I've got no experience without it.

Best wishes for the days to come.
The Manse Hen
Post Reply Post Reply
#14
RE: Resmed AirSense 10 have I set it up right ?
I will post my leak report in the morning but I guess what I am asking are my leaks high enough to worry about I don't know the range and scales very well so I don't know if they are high or not.

Sleepy head gave a warning and so did resmed's cloud service are my leaks on the high or low side ? I can try tighting the straps a little.
Post Reply Post Reply
#15
RE: Resmed AirSense 10 have I set it up right ?
(01-31-2015, 10:04 PM)Simon Wrote: I will post my leak report in the morning but I guess what I am asking are my leaks high enough to worry about I don't know the range and scales very well so I don't know if they are high or not.

Sleepy head gave a warning and so did resmed's cloud service are my leaks on the high or low side ? I can try tighting the straps a little.

Hi Simon,

Unintentional Leak (which doesn't include the designed-in vent rate of the mask) of up to 24 Liters per minute is considered acceptable.

If Unintentional Leak is greater than 30 L/min then your machine won't even try to determine whether the apneas were Central or Obstructive in type, and these apneas will be marked type Unknown.

If Unintentional Leak is greater than 40 L/min then your machine counts this as Large Leak.

I think Large Leak for more than 30% of the previous night results in Mr Frowny appearing on the machine's display.

As long as the CAI (Central Apnea Index, which is the average number of Central Apneas per hour of sleep) remains 2 or 3 or less, I suggest increasing EPR to 2 for a week, and then increasing EPR to 3.

You posted statistics for the minimum pressure (usually same as Start Pressure setting), the median pressure (the 50 percentile pressure you were at or lower for at least 50% of the night), the 90% pressure (the 90 percentile pressure you were at or lower for at least 90% of the night, and at or higher for at least 10% of the night), and the maximum pressure (the peak pressure reached that night).

The maximum pressure reached was 20 cm H2O (20 centimeters water column), which is the max your machine is able to produce.

If you are able to avoid sleeping on your back, this would likely reduce how much pressure you need, probably resulting in less Leak and less swallowed air and better therapy.

Take care,
--- Vaughn


The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies.  Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  ResMed AirSense 10 Respond (Autoset) dropped a foot radarrab 2 141 11-28-2024, 09:58 PM
Last Post: srlevine1
  ResMed AirSense 10 Respond Model and OSCAR upsman 8 3,191 11-28-2024, 07:50 PM
Last Post: SarcasticDave94
Question [Diagnosis] Comparing AHI Accuracy: ResMed AirSense 11 vs. In-Lab Sleep Study NightmaresOfTheBottom 2 132 11-28-2024, 07:08 PM
Last Post: Rickyricardo
  Resmed Airsense 11 Random SD Card Issues ineedsomesleep 10 495 11-27-2024, 09:23 PM
Last Post: Dgdoc
  Resmed Airsense 10 HELP Misteryams 3 187 11-26-2024, 08:54 PM
Last Post: Crimson Nape
  DC Jack for Resmed Airsense 10 HeadlessDeath 8 2,527 11-25-2024, 06:54 PM
Last Post: Puppy
  [Equipment] ResMed AirSense 11 compatible pulse oximeters timtaller 121 53,680 11-24-2024, 05:55 PM
Last Post: dgconder


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.