Hi everyone - new PAP user here. I notice that for large parts of my evening my wave form has sharp peak with a steep drop off at exhalation instead of the rounded tops that I've been told is the goal of therapy. What does this wave form show, and if it's abnormal what can I do to remedy it? This is from last night and I feel very rested.

   
   
Post Reply Post Reply
#2
RE: Flow Rate wave form has sharp peaks
This is actually a class II flow limitation that your machine is missing for some reason. Your inspiration flow rises sharply from zero then diminishes through the rest of the breath before declining to zero-flow (red line), and becoming expiration. We can quickly solve this by increasing minimum pressure from 6.0 to 7.0, leaving maximum pressure at 11.0 and adding EPR Full-Time at Setting 2 or 3.  Is this flattened or declining inspiration flow typical throughout the night? See the Flow Limitation wiki https://www.apneaboard.com/wiki/index.ph...limitation

[Image: attachment.php?aid=73823]
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: Flow Rate wave form has sharp peaks
Thanks for the quick reply.

Will setting EPR=2 at 7 cmH2O provide enough expiratory pressure support to prevent mouth leak? I've been struggling with that as a new user and from what I've been reading having high expiratory pressure will help give the soft palate proper support. Thanks again for your insight.

And yes, there are periods where I have perfect rounded wave forms. But these tend to dominate.
Post Reply Post Reply
#4
RE: Flow Rate wave form has sharp peaks
Although expiratory pressure will be a bit lower with EPR, it should help, rather than hurt any mouth leaks. With EPR on, pressure will rise during inspiration to provide support where your airway becomes more obstructed with higher inspiratory flow and effort. It should work well for you.

We can work on mouth leaks as a separate issue. If you sleep on your side, try pulling a corner of your pillow between your shoulder and jaw to physically help keep your jaw closed. You can practice speaking while under pressure, and sipping water through a straw. This teaches airway control. It may take a few tries to get it, but eventually the skill in preventing leaks while talking or sipping water can translate to the subconscious sleep state.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#5
RE: Flow Rate wave form has sharp peaks
Thank you! I'll change to APAP 7-11 with EPR=2 tonight and will report back with hopefully positive results.
Post Reply Post Reply
#6
RE: Flow Rate wave form has sharp peaks
(Yesterday, 10:24 AM)Sleeprider Wrote: Although expiratory pressure will be a bit lower with EPR, it should help, rather than hurt any mouth leaks. With EPR on, pressure will rise during inspiration to provide support where your airway becomes more obstructed with higher inspiratory flow and effort.  It should work well for you.

Reporting back after setting EPR=2 with APAP 7-11 per your recommendation. Wave forms look much better! Does this mean I'm a good candidate for bi-level? Or is bi-level only for those on higher pressures? I thought EPR was only a comfort setting; I didn't realize it could help in this way.

P.S. I'm stilling getting consistent flattened peak that the Resmed algorithm either responds to or doesn't. I'm assuming this means I need to titrate higher until this is eliminated.
Post Reply Post Reply
#7
RE: Flow Rate wave form has sharp peaks
EPR works in the same way as PS in a bilevel. You have not used the maximum EPR available to you yet. I can show you side-by-side images of the Airsense 10 and Aircurve 10 mask pressures, and the way pressure support is delivered is nearly identical except bilevel allows PS above 3.0 cm. The hardware on your Airsense 11 is identical to the Aircurve 11, and only the firmware is different. I think if we were able to dig deep into the actual logic, that we would find the CPAP is actually a bilevel with limited PS capability. So to answer your question, if you get good results with CPAP, there is no need to spend the money on bilevel. For any continued inspiratory peak flattening, just increase minimum pressure a bit and move to EPR 3.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  AirCurve 11 ASV Backup Rate? b.e.wilson 7 110 46 minutes ago
Last Post: b.e.wilson
  [Pressure] OA without Flow Limitation? lessgroggy 10 200 10 hours ago
Last Post: SeePak
Arrow AHI < 1.0 now, but still tired? Is it UARS, RERA, arousals or FL in your FR peaks? 2SleepBetta 66 17,982 Yesterday, 12:03 PM
Last Post: SeePak
  High leak rate from large nasal pillows? gohoos1998 2 106 01-05-2025, 08:14 PM
Last Post: SarcasticDave94
Question Help interpreting OSCAR Flow Rate Data w/ UARS Primarysecondary465 0 110 01-04-2025, 11:33 AM
Last Post: Primarysecondary465
  OSCAR data high respiratory rate amc237 4 229 01-03-2025, 12:33 PM
Last Post: Dormeo
  Flow Limits How to Upgrade CPAP to Bilevel Deborah K. 118 17,343 01-02-2025, 04:35 AM
Last Post: alimama


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.