(01-18-2018, 12:28 PM)bonjour Wrote: That article also describes and contains links to the titration protocols that sleep clinics use.
Thank you. I have read the article in full as well as all of the linked pages and documents. I've read in full the latest AASM guidelines, the latest ResMed and Philips protocols, and a lot of the medical literature about APAP.
None of that provides a solid evidence base for adjusting APAP Min and Max values. I am therefore interested to hear expert opinion, ie what folks here tend to keep in mind as "rule of thumb" formulas based on initial APAP median and/or P90/P95 values.
One such opinion is set forth in the article you linked. For example, the article gives examples of Min = P90 - 2 for PR, or Min = Med -2 for ResMed.
Quote:A good rule of thumb is to keep your minimum pressure setting about 2-cm below your 90% pressure (PR) or Med Pressure (ResMed) or near the average if they are close. The Dreamstation tends to be conservative with raising pressure, and the algorithm favors returning to the minimum setting. This can often adequately treat OA, but leave you with excessive Hypopnea (H) events. The remedy is an increase in the minimum pressure setting. PR machine: If your 90% setting is 12 cmH2O Min is set to 10 cmH2O and Max Remains at 20 cmH2O ResMed machine: Use the Med value for pressure. If your Med setting is 11 cmH2O, set the minimum to 9 cmH2O and Max remains at 20 cmH2O.
But I am guessing that not everyone uses that approach. Others likely have different rule of thumb formulas they tend to keep in mind.
Anyone willing to share theirs? I'll share mine:
Given person with straightforward OSA (no CAs either on diagnostic study or while on APAP) has been on APAP for two weeks with a Min pressure of 4cm H2O and a Max pressure of 20cm H2O.
The two-week average of their daily median pressure = X
The two-week average of their daily 90th or 95th percentile pressure (P90/P95) = Y
My approach to initial adjustment would typically be as follows:
New Min = X-2 or 5, whichever is
higher
New Max = Y+2 or 20, whichever is
lower
That is regardless of whether they were using a ResMed or Respironics device and with the understanding that further adjustments would be made based on response.