Inspiration and Expiration Time
#1
Insp.Time vs. Exp.Time (Ratio)
Quick Question. I have searched the board and have a simple question. What is a good Insp.Time vs. Exp.Time (Ratio) with CPAP therapy? Just trying to get an idea.
Post Reply Post Reply
#2
RE: Insp.Time vs. Exp.Time (Ratio)
According to Google:

Inverse Ratio Ventilation - StatPearls - NCBI Bookshelf
National Institutes of Health (.gov)  Apr 6, 2023
Quote:A typical I:E ratio for most situations would be 1:2. If we apply this ratio to the patient above, the 6-second breath cycle will break down to 2 seconds of inspiration and 4 seconds of expiration. Changing the I:E ratio to 1:3 will result in 1.5 seconds of inspiration and 4.5 seconds of expiration.

- Red
Crimson Nape
Apnea Board Moderator
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: Insp.Time vs. Exp.Time (Ratio)
(10-27-2023, 04:25 PM)Crimson Nape Wrote: According to Google:

Inverse Ratio Ventilation - StatPearls - NCBI Bookshelf
National Institutes of Health (.gov)  Apr 6, 2023

- Red

I saw that searching with Google, however, I rarely ever see OSCAR reports on this board with folks within that ratio.
Post Reply Post Reply
#4
RE: Insp.Time vs. Exp.Time (Ratio)
I just checked my averages for last night, and it equates to 1:1.63.

- Red
Crimson Nape
Apnea Board Moderator
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#5
RE: Insp.Time vs. Exp.Time (Ratio)
(10-27-2023, 05:21 PM)Crimson Nape Wrote: I just checked my averages for last night, and it equates to 1:1.63.

- Red

Ok thanks, that's around where I am as well.
Post Reply Post Reply
#6
RE: Insp.Time vs. Exp.Time (Ratio)
I am more like 1:1,1 no matter the settings on my CPAP. Don't feel like having an issue with a stable respiratory rate of between 13-15 (measurements by smart watch). My therapy is not working well yet, waiting for BiPAP to solve flow limitations (my AHI has always been < 1).

Interesting question I wanted to ask as well. Following.
Post Reply Post Reply
#7
RE: Insp.Time vs. Exp.Time (Ratio)
You cannot necessarily rely on the Oscar reported inspiratory and expiratory median times. These are actually reported by the machine in question and is invariably inaccurate in all Philips machines and may be inaccurate in Resmed devices, especially where expiratory flow approaches zero and/or in the presence of cardiogenic oscillation. I often ask members for zoomed images of the flow rate so I can confirm or refute the reported statistics. Most of the time in healthy individuals, the respiratory ratio falls into a normal range from 1:1 to 1:3, and I have observed the majority of I:E ratios are close to one to 1.5. Inspiration time that is confirmed to be longer than expiration time is normally attributable to flow limitation, and abnormally long expiration is often asthma or COPD.

Historic trivia: Originally Sleepyhead displayed an I:E ratio, but it was incorrectly calculated as Texpiration/Tinspirationx100 and produced a number like 65.0, which was off by a factor of 100, meaning inspiration was .65 the time of expiration. It was decided back in February 2019 that Oscar should not report the I:E ratio due to the inaccurate reporting of inspiration and expiration times by the machines, general confusion on the part of Oscar users and the fact the legacy calculation was simply wrong. Ultimately, Oscar is a "Reporter" of information provided by the devices and today, Oscar does not report the I:E ratio for this reason, and it has not been reported since Sleepyhead 1.1.0. Here is a legacy image of Sleepyhead:

[Image: attachment.php?aid=10192]
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#8
RE: Insp.Time vs. Exp.Time (Ratio)
Thanks for your explanation Sleeprider!
Post Reply Post Reply
#9
Inspiration and Expiration Time
Sorry for all the questions, but reading and looking at others OSCAR data in past posts, what is ideal inspiration and expiratory times. 

For example, I have seen some charts where inspiration time is something like 2.3s and expiratory time is 1.2s.

Then others, like mine, where inspiration time is 1.3s and expiratory time is 2.1s or something like that. Does it matter? Do the machines get this correct? Thanks for all this board does!
Post Reply Post Reply
#10
RE: Inspiration and Expiration Time
This seems to be an OK web site, and there's information there:

https://respelearning.scot/topic-1-anato...-breathing

Keep in mind that the machine counts the pause between breaths as exhalation time.

When I used a ResMed Airsense 10 Autoset, I got weird ratios of inhalation to exhalation (inhalation longer than exhalation). With the VAuto, all is well. It seems that the Airsense 10 got confused by the cardioballistic artifacts that show up as little oscillations between breaths on my flow rate graphs. For whatever reason, the VAuto isn't confused by the CB artifacts.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  [CPAP] Please help, first time analyzing my oscar data napnea 3 102 Yesterday, 02:38 PM
Last Post: Old Steve
  1 year on CPAP, still tired all the time joelin02 5 431 09-10-2024, 06:39 PM
Last Post: quiescence at last
  First Time Poster, need advice for new machine! kbjames 7 400 08-26-2024, 08:18 PM
Last Post: SarcasticDave94
  Trying Bilevel for first time - need help! sleepquality 4 331 08-18-2024, 12:38 PM
Last Post: sleepquality
  [Symptoms] Sleep paralysis every time I use my CPAP DoRaCh79 5 2,371 08-01-2024, 08:53 AM
Last Post: jaguarman9991
  [Diagnosis] First time seeing CSR on my carts ozbusa 7 297 07-31-2024, 04:10 PM
Last Post: ozbusa
  long time user , new problems spud45 9 437 07-18-2024, 09:01 PM
Last Post: Deborah K.


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.