RE: Looking for opinions
Due to high flow limits, you're going to want EPR 3 to get to work faster. Your min pressure of 5 prohibits this until your pressures pass 7. I would increase this min pressure to 7 to help correct this.
I would also consider turning off the Ramp, even though yours is on auto. Ramp removes almost all therapy until it transitions to therapy pressures.
Mask Primer
Positional Apnea
Attach OSCAR, etc.
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
RE: Looking for opinions
Thanks. I don't understand the point about EPR. By the time the apneas and flow limitations start, the exhalation pressure is above 7 anyhow. All that increasing the minimum pressure would do would be to raise the pressure in those few intervals where it currently drops below 7. I assume that the pressure is dropping because the machine figures higher pressure isn't needed. Am I wrong?
Not sure about the ramp. Some of it is while I am awake, and for the rest there aren't any apneas to treat in any of the data that I have looked at.
RE: Looking for opinions
Dave is right. You need the extra pressure both to make the EPR 3 work properly and to help overcome your high flow limits. You need to turn off ramp also, as that fights with your effort to overcome high flow limits.
Machine: ResMed AirCurve 10 Vauto
Mask: Bleep DreamPort Sleep Solution
RE: Looking for opinions
Flow limits are apnea just like Oa and H events. The Oa and H has to be 10 seconds or longer in duration. A flow limit no as long as the other events and Resmed mechines raise pressue as fast as they can to top them from becoming Oa and H events. Also if they pressure start up a little (like 7 instead of 5) it can stop the flow limits quicker and they may not be classified as an H event.
RE: Looking for opinions
Thanks all. I will try the recommendations tonight.
RE: Looking for opinions
I can't determine if the leaks are the cause of higher pressure and flow limits, or if flow limits required higher pressure and leaks result. Either way, your leaks needs attention.
There's another item maybe at play, let's get some feedback from others more knowledgeable about this. The For Her settings might be another source of hindered therapy.
Mask Primer
Positional Apnea
Attach OSCAR, etc.
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
RE: Looking for opinions
The Flow limits are bad and you are at the max of EPR 3. The Resmed bilevel might be what is needed to get them down but as Dave said the For Her mode may make some difference. I would suggest you change it to the normal setting to see if it helps..