Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Need help on cpap settings
#1
Need help on cpap settings
        I have used resmed auto sense 10 for 2 years with no help on sleep quality. Wake up every 1-2 hours every night. Would you mind looking at my sleep report and see if there is any hope? I have enclosed one night for you to look at. Use resumes nasal pillow(n30I), settings on Cpap is 5 min. and 9 max on manual settings. Was set on these settings with 45 minute ramp time. Thanks
Post Reply Post Reply
#2
RE: Need help on cpap settings
Welcome to the forum, we will help you, we will get you better than you currently feel.  Getting you all the way may take another machine (ResMed VAuto, a BiLevel) but let's hold off on that thought until we see what we can do with your AutoSet.

First, the AutoSet is a good machine but it is not setup properly for your apnea.
The biggest issue that you have is your flow limits.  I suspect UARS, Upper Airway Resistance Syndrome, but without a sleep study I cannot prove it, at least not yet.

I would like you to change your settings

Set Mode = Auto (This is easier to optimize your settings, we may or may not suggest a fixed pressure again, don't know yet)
Set min Pressure = 7 (This will allow us to get maximum use and therapeutic value fro the EPR)
Set Max Pressure = 12 (Not too high as I suspect flow limits will drive pressure to the max )
Set EPR = 3, FullTime (EPR is our major tool to treat the Flow Limits which I suspect are worse that your numbers indicate)

This should lower the Flow Limits that you are currently seeing and (hopefully) eliminate the RERAs that you are currently experiencing.  RERAs are by definition a series of flow limitations ending with arousal.
I do expect to change these settings based on the results from using them.
Post Reply Post Reply
#3
RE: Need help on cpap settings
Thanks, I did do sleep study in 2017-18. I have a copy in you need it later. Made the changes to the autosense already, thanks.
Post Reply Post Reply
#4
RE: Need help on cpap settings
Your machine is currently set to a fixed pressure of 9.0

You need to switch to auto instead of fixed, and then start off with the settings suggested by Bonjour.
- They are not spelling/grammar errors.. I live in Australia, we do it differently Down Under  Big Grin -
Post Reply Post Reply
#5
RE: Need help on cpap settings
        Probably slept little better new settings. Still woke up every hour or two, but no problem with change in settings. Rera look better.
Post Reply Post Reply
#6
RE: Need help on cpap settings
I will leave the setting for the experts you have answering, however, your leak rate needs to be addressed.    Right click on the leak rate and go to dotted line and check mark the "leak rate upper threshold".  That will help you identify the amount the machine can accommodate before the leak is to big to adapt to. 

Check out the Collar link at the bottom in the signature.
Apnea (80-100%) 10 seconds, Hypopnea (50-80%) 10 seconds, Flow Limits (0-50%) not timed  Cervical Collar - Dealing w DME - Chart Organizing
Post Reply Post Reply
#7
RE: Need help on cpap settings
(09-09-2020, 09:05 AM)Beaumont1124 Wrote: Probably slept little better new settings. Still woke up every hour or two, but no problem with change in settings. Rera look better.

Please pay attention to my next observation:

Your machine is STILL showing that it is on CPAP mode, not on AUTO, which is what bonjour asked you to do.  You may need to use the manual to do this, but if you know how to do so, please get out of straight CPAP mode and switch the machine to AUTO.  Then, tomorrow, if the other settings suggested by bonjour are correct, you'll hopefully see a substantial diminution of your AHI.
Post Reply Post Reply
#8
RE: Need help on cpap settings
(09-09-2020, 04:02 PM)mesenteria Wrote:
(09-09-2020, 09:05 AM)Beaumont1124 Wrote: Probably slept little better new settings. Still woke up every hour or two, but no problem with change in settings. Rera look better.

Please pay attention to my next observation:

Your machine is STILL showing that it is on CPAP mode, not on AUTO, which is what bonjour asked you to do.  You may need to use the manual to do this, but if you know how to do so, please get out of straight CPAP mode and switch the machine to AUTO.  Then, tomorrow, if the other settings suggested by bonjour are correct, you'll hopefully see a substantial diminution of your AHI.

Actually, his machine is in Autoset mode, although there is a ramp starting at 5.0.  We can see the pressure adjusting and mostly staying near the maximum pressure setting of 12.0 due to high flow limitations.  The EPR is at 3.0 so and while there are some leaks, the results are improving with AHI at 2.84.  I'm not in this thread (until now), but I'm guessing Beaumont would greatly benefit from bilevel pressure, but will be moving to a minimum of 10.0 and maximum of 14.0 with EPR 3.

The Resmed SD data only reports the settings from before noon. If changes are made later in the day, the old settings show up, so we have to read the charts.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#9
RE: Need help on cpap settings
I'll go with that
Min = 10
Max = 14
EPR = 3 (current value)
and Turn Ramp off. If you need the ramp we will add it back in.
Post Reply Post Reply
#10
RE: Need help on cpap settings
I must not be looking at the right information, Sleeprider.  When I look at either of his two last panels, the index at the left margin says the mode set is 'CPAP' and it says the pressure is fixed at 9.0  Where else should I be looking?
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
Wink New CPAP User Seeking Advice on Pressure Settings wsswms 2 131 Yesterday, 12:55 PM
Last Post: Deborah K.
  Cpap settings help RobbieB 1 154 11-24-2024, 10:12 PM
Last Post: Deborah K.
  Help with CPAP settings - feeling better but still tired Jess2117 6 223 11-23-2024, 11:08 AM
Last Post: G. Szabo
  Cpap settings help RobbieB 2 161 11-22-2024, 08:31 PM
Last Post: Deborah K.
  [Treatment] Optimising CPAP settings Sleeper49 1 133 11-22-2024, 12:25 PM
Last Post: Old Steve
  [CPAP] Cpap settings help-newbie apoch 1 139 11-18-2024, 07:00 PM
Last Post: PeaceLoveAndPizza
  Weinmann / Löwenstein Prisma 20A Auto CPAP: APAP settings: standard vs. dynamic Costas P 8 13,568 11-18-2024, 02:52 PM