FDA Warning
Date Issued: June 30, 2021
The U.S. Food and Drug Administration (FDA) is alerting people who use Philips Respironics ventilators, BiPAP, and CPAP machines and their health care providers that Philips Respironics has recalled certain devices (see table below) due to potential health risks.
The polyester-based polyurethane (PE-PUR) sound abatement foam, which is used to reduce sound and vibration in these affected devices, may break down and potentially enter the device’s air pathway. If this occurs, black debris from the foam or certain chemicals released into the device’s air pathway may be inhaled or swallowed by the person using the device.
If you use one of these affected devices (see table below), talk to your health care provider to decide on a suitable treatment for your condition and follow the recommendations listed below.
Philips Respironics is recalling the following affected devices manufactured between 2009 and April 26, 2021.
Recommendations for People Who Use Affected BiPAP or CPAP Machines and Caregivers
Talk to your health care provider to decide on a suitable treatment for your condition, which may include:
(1) Stopping use of your device
(2) Using another similar device that is not part of the recall
(3) Using alternative treatments for sleep apnea, such as positional therapy or oral appliances, which fit like a sports mouth guard or an orthodontic retainer.
(4) Initiating long term therapies for sleep apnea, such as losing weight, avoiding alcohol, stopping smoking, or, for moderate to severe sleep apnea, considering surgical options.
(5) Continuing to use your affected device, if your health care provider determines that the benefits outweigh the risks identified in the recall notification.
(6) Follow the manufacturer’s instructions and recommended cleaning and replacement guidelines for your CPAP machine and accessories. Ozone cleaners may worsen the breakdown of the foam, and there are other potential risks associated with the use of ozone and ultraviolet (UV) light products for cleaning CPAP machines and accessories.
(7) Register your device(s) on Philips Respironics’ recall to stay informed of updates from Philips Respironics regarding any new instructions or other corrective fixes, which the FDA is requiring.
(8) Report any problems with a device through the FDA’s MedWatch Voluntary Reporting Form.
[*** removed web-link ***]
PE-PUR Foam May Be Inhaled or Swallowed, Presenting a Potential Health Risk
Polyester-based polyurethane (PE-PUR) is a sound abatement foam used to reduce sound and vibration in these devices and other medical equipment. The PE-PUR foam in the affected Philips Respironics CPAP, BiPAP, and ventilator devices may:
(1) Break down (degrade) into particles which may enter the device’s air pathway and be inhaled or swallowed by the user
(2) Release certain chemicals into the device’s air pathway, which may be inhaled
These issues can result in serious injury, which can be life-threatening, cause permanent impairment, and require medical intervention to prevent permanent damage.
To date, Philips Respironics has received several complaints about the presence of black debris/particles within the device’s air pathway. Philips Respironics also has received reports of headache, upper airway irritation, cough, chest pressure, and sinus infection, which may be related to this issue, though the cause of the symptoms cannot be definitively linked.
The potential risks of particulate exposure include irritation to the skin, eye, and respiratory tract, inflammatory response, headache, asthma, and toxic or carcinogenic effects to organs, such as kidneys and liver.
The potential risks of exposure to chemicals released into the device’s air pathway from the PE-PUR foam include headache; dizziness; irritation in the eyes, nose, respiratory tract, and skin; hypersensitivity; nausea/vomiting; and toxic and carcinogenic effects.
The foam degradation may be exacerbated by high heat and high humidity environments, and by use of unapproved cleaning methods, such as ozone. To date, there have been no reports of death as a result of these issues.
FDA Actions
The FDA is working with Philips Respironics to evaluate the issue, the scope of the recall, and the most appropriate mitigation strategies, including corrective actions by the company.
[*** "Philips' "corrective actions" to-date refers to requesting user device registration details. ***]
The FDA is analyzing medical device reports (MDRs) related to the affected devices over the period of 2009-2021 for reports that could be related to this issue.
The FDA does not have evidence at this time that any other CPAP machines, BiPAP machines, or ventilators, from Philips or other manufacturers, are affected.
The FDA will continue to monitor supply and demand to assess availability of the affected devices and any potential shortages.
The FDA will continue to share updates with the public as we learn more.
Reporting Problems with Your Device
If you think you had a problem with a CPAP, BiPAP, or mechanical ventilator, the FDA encourages you to report the problem through the MedWatch Voluntary Reporting Form.
[*** web-link removed ***]