Respironics System One A-Flex, blower won’t start, Beep Beep Alarm, “Service Required
#1
Respironics System One A-Flex, blower won’t start, Beep Beep Alarm, “Service Required

The Problem:
1.Blower won’t start, alarm sounds, see “Service Required” in display window
2.Sound asleep, blower stops, alarm sounds, see “Service Required” in display window

The Solution:
1. I did both of these things at the same time so, one or both solved the problem.
a. Took hose filter and intake filter off the system
b. Took memory card out and cleared the memory on the card
c. Put clean filters back on the system

CPAP has worked fine for a week now. THANK YOU TO THE PEOPLE WHO POST ON THIS BOARD. Even the ones who send you down the rabbit hole to look for Alice!
Post Reply Post Reply
#2
RE: Respironics System One A-Flex, blower won’t start, Beep Beep Alarm, “Service Required
Hi rae95131,
WELCOME! to the forum.!
It's good to hear that your machine is working now and I hope it continues.
Good luck to you as you continue your CPAP therapy.
trish6hundred
Post Reply Post Reply
#3
RE: Respironics System One A-Flex, blower won’t start, Beep Beep Alarm, “Service Required
Welcome to ApneaBoard rae95131 ! Glad you were able to fix your problem!
APNEABOARD - A great place to be if you're a hosehead!!  Rolleyes  

-------------------------------------------------------------------------------------------------
EVERY ACCOMPLISHMENT BEGINS WITH THE DECISION TO TRY!
Post Reply Post Reply
#4
RE: Respironics System One A-Flex, blower won’t start, Beep Beep Alarm, “Service Required
Hi everybody,

Ihave been successfully using system one Philips c flex for almost 2 years.
Lately, my machine started creating problems .
Last few nights, my cpap suddenly stopped for seconds then resumed back for few minutes then started beeping (alarm with blinking light) then totally stopped with the screen indicating SERVICE REQUIRED , I then unplugged the electrical cord then re-connected it then machined started , for few minutes though then stopped again!!!I changed the filters, removed memory card, changed my full mask, but problem persists.
I'm unable to use my machine as a result.

I seek your kind advice if anyone comes across same issue.
Post Reply Post Reply
#5
RE: Respironics System One A-Flex, blower won’t start, Beep Beep Alarm, “Service Required
Almost two yrs?.
Send it in to PR. PR machines have a two yr warrenty
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Respironics remstar Pro c-flex system one card won't download staduimjunkie 8 2,737 Today, 03:35 AM
Last Post: Hardhead
  UARS? Time to start therapy mpz12 1 153 11-25-2024, 10:21 PM
Last Post: gainerfull
Exclaimation [News] RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS Sleeprider 3,013 437,544 11-23-2024, 11:01 AM
Last Post: Tampa Jim
  Service Required Message jmack729 2 187 11-06-2024, 11:57 AM
Last Post: StratCat48
  NOTICE: Buddy system removed SuperSleeper 3 237 11-04-2024, 10:58 AM
Last Post: SuperSleeper
  Airsense 10 start leaking from bottom kerm007 0 176 10-14-2024, 07:52 AM
Last Post: kerm007
  Post Cervical Surgery start up DEJONES1954 1 227 10-05-2024, 11:31 AM
Last Post: Jay51


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.