Slowriter, I don't think "airflow resistance" can be measured non-invasively. Restricted airflow has certain characteristics in how rapidly the inspiratory flow increases or if it reaches a prolonged peak, and we recognize that as flow limitation. That is very effectively detected and used by Resmed to drive the auto pressure algorithm. Philips seems to use more of a volume to detect hypopnea and flow limitation. It does not seem to take into consideration "flatness" or that a longer inspiratory period is being used to get the same volume. As a result, flow restriction can develop into an event without the machine being able to make a proactive pressure adjustment. Until it becomes acute, flow limitation may not affect tidal volume or minute vent, which is how Philips flags flow reductions and classifies them as hypopnea or apnea. Philips does rely on snores as an indicator of flow restriction, but once snoring is going on, the flow limitation is already well advanced. This is my conjecture based on observed operating characteristics and the reading of patent rights documents.
I probably need to do some better documentation on Philips in the wiki, but the Wiki has this entry for Variable Breathing which is part of the Philips algorithm. I think the 4-minute moving average is far to slow to respond when restriction can build very quickly:
Variable Breathing - Philips Respironics variable breathing algorithm seeks to develop a peak inspiratory flow trend on a 4 minute moving average, and measure the deviation above or below that trend. The system is designed to identify variable breathing and to turn over control of the pressure support system, from the auto CPAP controller to the Variable Breathing controller. Without figures, we don't know what that controller response is, but it seems if VB is detected while the pressure is steady, the VB controller will maintain that pressure, but if pressure was increasing before VB controller took control, the pressure is reduced up to 2-cm. Similarly if pressure was decreasing ahead of VB detection, the VB controller will in increase pressure up to 2-cm. The duration of VB controller action is 5 minutes unless VB is still detected.
http://www.apneaboard.com/wiki/index.php..._Breathing
For a real-life idea of how the VB algorithm actually interferes with therapy, take a look at this recent post with an example graph
http://www.apneaboard.com/forums/Thread-...#pid364069