[Treatment] Still fatigued, history, are these CAs real?
#71
RE: Still fatigued, history, are these CAs real?
My bad.
Post Reply Post Reply
#72
RE: Still fatigued, history, are these CAs real?
The high sensitivity for trigger is something I did a couple years ago. I had similar central events flagged and that Change works great to clear it up. You have a relatively long expiration time so leave cycle sensitivity at medium. I think as you become accustomed to the Vauto you’re really going to like it. Hopefully no smoke odor this time.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#73
RE: Still fatigued, history, are these CAs real?
(08-05-2020, 02:45 PM)Sleeprider Wrote: The high sensitivity for trigger is something I did a couple years ago. I had similar central events flagged and that Change works great to clear it up. You have a relatively long expiration time so leave cycle sensitivity at medium. I think as you become accustomed to the Vauto you’re really going to like it. Hopefully no smoke odor this time.

Definitely no unexpected odor this time!

I reduced pressure a little bit to 5.8/4/9.8, so still at fixed pressure.  Looks like flow limits might have increased a bit over 6/4/10 but it is difficult to know for sure with just one night of each setting.

I tried high sensitivity last night.  In statistics, the one thing that stood out was the I:E ratio -- 0.58 instead of the 0.93 that is typical.  I guess that is a result of setting trigger sensitivity to High?  I noticed a difference in breathing at the beginning of the night but immediately became used to it.

As I look over the flows for the night, I am struck by how a flow limit of just 0.02 is accompanied by a flattened-top wave pattern.  Still have some waxing and waning but less than before I started EERS testing.

Are "waxing and waning" and "oscillatory breathing" and "CO2 inspired breathing" all synonyms for one another?
Useful links
Download OSCAR (current version is 1.5.1)
Best way to organize charts
How to attach charts to your post

Apnea Board Monitors are members who help oversee the smooth functioning of the Board. They are also members of the Advisory Committee which helps shape Apnea Board's rules & policies. Membership in the Advisory Members group does not imply medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply
#74
RE: Still fatigued, history, are these CAs real?
IMHO yes.  "Waxing and waning" and "oscillatory breathing" are a description of the waveform whereas "CO2 inspired/induced breathing" refer to a cause of that breathing pattern.
Post Reply Post Reply
#75
RE: Still fatigued, history, are these CAs real?
I hate the way OSCAR calculates I:E ratio. It's based on I/E, but if you look at ResScan I:E ratio is expressed as 1 to xx for inspiration to expiration and this is accomplished by the product of E/I. I just don't like this perpetuation of the Mark Watkins error. That said, the longer expiration time is very normal and should allow better ventilation. Any change to tidal volume or minute vent?
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#76
RE: Still fatigued, history, are these CAs real?
Tidal volume and minute vent were about the same as before I started the EERS experiments: 460 and 6.13 were values for the last two nights on the VAuto.

I see that OSCAR calculates an I:E ratio that it shows on the Overview page.  Does it show this anywhere else?  What machines does it show this for anyway?  I don't see it for the AutoSet but do see it for the VAuto.

UPDATE: Never mind.  I found your post, with a list of machines that report this, and discussion from early 2019.  What has changed in OSCAR about handling I:E ratio since then?
Useful links
Download OSCAR (current version is 1.5.1)
Best way to organize charts
How to attach charts to your post

Apnea Board Monitors are members who help oversee the smooth functioning of the Board. They are also members of the Advisory Committee which helps shape Apnea Board's rules & policies. Membership in the Advisory Members group does not imply medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply
#77
RE: Still fatigued, history, are these CAs real?
We need to flag this as an OSCAR calc, anything that OSCAR itself calculates, perhaps a supercase "O"
Post Reply Post Reply
#78
RE: Still fatigued, history, are these CAs real?
(08-07-2020, 08:15 AM)Sleeprider Wrote: I hate the way OSCAR calculates I:E ratio.  It's based on I/E, but if you look at ResScan I:E ratio is expressed as 1 to xx for inspiration to expiration and this is accomplished by the product of E/I.  I just don't like this perpetuation of the Mark Watkins error.  That said, the longer expiration time is very normal and should allow better ventilation.  Any change to tidal volume or minute vent?

It looks like this is Resmed-specific and imported from the raw Resmed data. How is it being displayed for you?

I've seen I/E used by pulmonologists (though usually notated as Ti/Te) along with Ti/TTot.

https://www.pftforum.com/blog/what-does-...cise-mean/

If it's being labeled "I:E" we should display it as "1:N" (even if Resmed is only recording/reporting N).

If we want to keep displaying "N," we should change the label to E/I.
Post Reply Post Reply
#79
RE: Still fatigued, history, are these CAs real?
ok, I now have six days of data with VAuto 6-10, PS 4, high sensitivity for trigger.  Maybe a slight improvement over the AutoSet, but still not feeling as refreshed as I would like.

I am attaching screenshots for the last two days, links to videos of 6-minute through the entire days, and spreadsheet.

What should I try next?
  • Very high sensitivity?
  • Increase PS to reduce flow limits further?
  • ?
Thanks!

                   

https://www.dropbox.com/s/9cr0ut3ur06v3t...7.mp4?dl=0

https://www.dropbox.com/s/dlvqp76j2izs60...0.mp4?dl=0
Useful links
Download OSCAR (current version is 1.5.1)
Best way to organize charts
How to attach charts to your post

Apnea Board Monitors are members who help oversee the smooth functioning of the Board. They are also members of the Advisory Committee which helps shape Apnea Board's rules & policies. Membership in the Advisory Members group does not imply medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply
#80
RE: Still fatigued, history, are these CAs real?
I don't see a lot of optimization opportunities remaining in this data, and it may be time to look at factors external to PAP therapy as sources for "not feeling refreshed" or fatigue. We can see therapy is fragmented into several sessions, with primary sleep separated by a 2:15 hour break on 8/11 and almost 2-hours on 8/12. Everyone has different sleep habits, but routine breaks of over 2-hours in a night would make anyone not feel particularly refreshed the next morning.

I had a really crappy sleep last night myself. I don't know what time I woke up, but I laid in bed until dawn without sleeping, and really felt it. For me it was a combination of getting that damn Shingrex vaccination and too much wine at dinner. I'll do some make-up tonight. The point is, be aware of your sleep hygiene and that it is going to impact how your feel. If you're not aware of being disrupted by your therapy, you probably aren't. It sure doesn't show in the data.

Feel free to make small changes in minimum pressure to deal with the OA events. I would stick to small changes of 0.4 or less. Changing pressure support (IPAP) may improve some ventilation, but I don't see much flow limitation at PS 4.0. Again, try small changes in EPAP first while maintaining PS, then try small changes in PS. I don't think your settings are the root problem, but it's always okay to try small changes.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  OSCAR Data - Low AHI but still not feeling great, and are these real OA events? Santino 38 1,841 10-16-2024, 10:55 AM
Last Post: SeePak
  Extremely fatigued despite good numbers major maser 3 315 09-19-2024, 04:46 PM
Last Post: Crimson Nape
  Months of CPAP use and feeling more fatigued senseisaitama 4 353 09-15-2024, 11:43 PM
Last Post: SarcasticDave94
  Still fatigued with CPAP - looking for advice Zallomallo 7 392 08-31-2024, 10:50 PM
Last Post: Phaleronic
  How do real users with card to cloud send data to real sleep medical providers? silverchief2 2 326 06-01-2024, 08:11 PM
Last Post: SleepyDawg
  Is this real CSR or periodic breathing? Should I be worried? Jonkier 10 891 05-24-2024, 07:57 AM
Last Post: Sleeprider
Sad 7 Years of Pap Therapy and Still Utterly Fatigued. Top doctors Unhelpful kalensk 2 507 04-30-2024, 11:09 AM
Last Post: ButtonNoseBarbie


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.