04-21-2019, 05:15 AM
(This post was last modified: 04-21-2019, 05:16 AM by Broomstick.)
RE: What settings are submitted in prescription for ResMed AirCurve 10 VAuto?
I'm in need of the same thing.
I already have a machine and supplies. I want to swap it out for the AirCurve VAuto.
1. Do I really need all those supplies to be listed on the prescription?
2. And, do I just need:
IPAP Max: ##
EPAP min: ##
PS: #
Or do I need:
EPAP: ##
IPAP: ##
EPAP min: ##
IPAP max: ##
Thank you.
RE: What settings are submitted in prescription for ResMed AirCurve 10 VAuto?
It is more like the first part you posted:
IPAP Max: ##
EPAP min: ##
PS: #
There is also a Ti Max and Ti Min that I think needs added.
There is a trigger of which mine is set at Medium. I am waiting for my copy of the clinician manual so I can determine more.
I'll post back when I figure out more.
RE: What settings are submitted in prescription for ResMed AirCurve 10 VAuto?
(04-22-2019, 09:28 PM)SleepFoo Wrote: It is more like the first part you posted:
IPAP Max: ##
EPAP min: ##
PS: #
There is also a Ti Max and Ti Min that I think needs added.
There is a trigger of which mine is set at Medium. I am waiting for my copy of the clinician manual so I can determine more.
I'll post back when I figure out more.
Thats what I thought. Ti Max and Ti Min should only be for BiPAP ST (timed with a puff of air every few seconds to treat centrals) if I understand.
RE: What settings are submitted in prescription for ResMed AirCurve 10 VAuto?
(04-23-2019, 09:51 PM)Broomstick Wrote: (04-22-2019, 09:28 PM)SleepFoo Wrote: It is more like the first part you posted:
IPAP Max: ##
EPAP min: ##
PS: #
There is also a Ti Max and Ti Min that I think needs added.
There is a trigger of which mine is set at Medium. I am waiting for my copy of the clinician manual so I can determine more.
I'll post back when I figure out more.
Thats what I thought. Ti Max and Ti Min should only be for BiPAP ST (timed with a puff of air every few seconds to treat centrals) if I understand.
Note that ResMed does not target a ST for Centrals
This info is from the ResMed Sleep Lab Titration Guide
- CPAP (continuous positive airway pressure) Fixed pressure delivered with optional expiratory pressure relief (EPR). It Treats OSA
- AutoSet/APAP (automatic positive airway pressure) Automatically adjusts pressure in response to flow limitation, snore and obstructive apneas. It treats OSA
- AutoSet for Her/APAP Automatically adjusts pressure in response to flow limitation, snore and obstructive apneas along with an increased sensitivity to each flow-limited breath, providing a more comfortable therapy for women. Increases sensitivity to each flow-limited breath, providing a more comfortable therapy for women (OK for men too). It Treats OSA
- VAuto Automatically adjusts pressure in response to flow limitation, snore and obstructive apneas; Pressure Support (PS) is fixed throughout the night and can be set by the clinician. It Treats OSA, non-compliant OSA
- S (Spontaneous) Senses when the patient is inhaling and exhaling, and supplies appropriate pressures accordingly. Both treatment pressures are preset: inspiration (IPAP) and expiration (EPAP). It treats Non-compliant OSA and COPD
- ST (Spontaneous/Timed) Augments any breaths initiated by the patient, but also supplies additional breaths if the breath rate falls below the clinician’s set “backup” respiratory rate. It Treats COPD, Neuromuscular disease (NMD), Obesity Hypoventilation Syndrome (OHS) and other respiratory conditions
- T (Timed) Supplies a clinician-set respiratory rate and inspiratory/expiratory time, regardless of patient effort. It Treats COPD, Neuromuscular disease (NMD), Obesity Hypoventilation Syndrome (OHS) and other respiratory conditions
- iVAPS (intelligent Volume-Assured Pressure Support) Maintains a preset target alveolar minute ventilation by monitoring delivered ventilation, adjusting the pressure support and automatically providing an intelligent backup breath. It Treats COPD, Neuromuscular disease (NMD), Obesity Hypoventilation Syndrome (OHS) and other respiratory conditions
- ASV (adaptive servo-ventilation) Targets the patient’s minute ventilation, continually learning the patient’s breathing pattern and instantly responding to any changes. It treats Central or mixed apneas, complex sleep apnea, Periodic Breathing (PB)
- ASVAuto Provides an ASV algorithm plus expiratory positive airway pressure (EPAP) that automatically responds on the patient’s next breath to flow limitation, snore and obstructive sleep apneas. It Treats Central or mixed apneas, complex sleep apnea, Periodic Breathing (PB)
- PAC (Pressure Assist Control, also known as Pressure Control) The inspiration time is preset in the PAC mode; there is no spontaneous/flow cycling. Inspiration can be triggered by the patient when respiratory rate is above a preset value, or delivered at a set time at the backup rate. It Treats Neuromuscular disease (NMD), pediatric patients