Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

advice on setting up an auto-bipap
#11
RE: advice on setting up an auto-bipap
I may be doing something wrong, but I set the numbers as suggested:

Min EPAP = 10 = your old min pressure setting
Max IPAP = 17 = your fixed IPAP pressure
Max PS = 4 = the difference between your fixed IPAP and EPAP settings.
Min PS = 2 = a wild guess on my part

and got these results according to Sleepyhead
EPAP min 10 and max 15, with med of 10 and 95% of 14.80
IPAP min 12 and max 17 with med of 12.5 and 95% of 16.8
PS min 2 and max 4 and med of 2 and 95% of 4

Is this in fact correct as you'd expect? AHI is down to 1.69 but this is one night's results.

thanks again for all your help.
Channel Min Med 95% Max
EPAPExpiratory Pressure (cmH20)
W-Avg: 10.91 10.00 10.00 14.80 15.00
IPAPInspiratory Pressure (cmH20)
W-Avg: 13.38 12.00 12.50 16.80 17.00
PSPressure Support (cmH20)
W-Avg: 2.47 2.00 2.00 4.00 4.00
Minute Vent.Minute Ventilation (L/min)
W-Avg: 7.43 0.12 7.25 10.25 20.25
Resp. RateRespiratory Rate (breaths/min)
W-Avg: 12.55 7.79 12.40 14.60 17.40
LeaksLeak Rate (L/min)
W-Avg: 4.61 0.00 0.00 17.00 53.00
Total LeaksTotal Leak Rate (L/min)
W-Avg: 36.27 27.00 34.00 50.00 89.00
Post Reply Post Reply
#12
RE: advice on setting up an auto-bipap
(04-20-2014, 12:34 AM)bob simons Wrote: I may be doing something wrong, but I set the numbers as suggested:

Min EPAP = 10 = your old min pressure setting
Max IPAP = 17 = your fixed IPAP pressure
Max PS = 4 = the difference between your fixed IPAP and EPAP settings.
Min PS = 2 = a wild guess on my part

and got these results according to Sleepyhead
EPAP min 10 and max 15, with med of 10 and 95% of 14.80
IPAP min 12 and max 17 with med of 12.5 and 95% of 16.8
PS min 2 and max 4 and med of 2 and 95% of 4

Is this in fact correct as you'd expect?
Those numbers make perfect sense given the settings and the fact that you had been titrated at 17/13.

If allowing the EPAP to go to almost 15 is causing problems with air in your stomach, you could bump the min PS = 3. Chances are that won't increase the AHI

Quote: AHI is down to 1.69 but this is one night's results.
That's good news. Let's hope the AHI stays low.

Post Reply Post Reply
#13
RE: advice on setting up an auto-bipap
Question.  I have a 750P Respironics. I get my best AHI with fixed PS, and worst with variable.  I don't know what to change to toggle?  Please?
Post Reply Post Reply
#14
RE: advice on setting up an auto-bipap
(10-21-2017, 04:06 PM)Aardvark08 Wrote: Question.  I have a 750P Respironics. I get my best AHI with fixed PS, and worst with variable.  I don't know what to change to toggle?  Please?

Hmmm... How were you able to determine that you get a better AHI with a fixed PS if you don't know how to set the machine at a fixed PS?

I don't know, either, but I would guess that you just set the min PS and the max PS at the same value. Perhaps someone will come along and tell us for sure.
Sleepster

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#15
RE: advice on setting up an auto-bipap
Right, that would be logical.  But mine is an early 750P and does not allow the setting of minimum and maximum PS.  Max only, and the lowest setting is 3.  I only see the difference in the Sleepyhead data.  Thanks!
Post Reply Post Reply
#16
RE: advice on setting up an auto-bipap
(10-21-2017, 10:12 PM)Aardvark08 Wrote: I only see the difference in the Sleepyhead data. 

What is it that you see? There's some particular value of the PS that gives you the lowest AHI?
Sleepster

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#17
RE: advice on setting up an auto-bipap
Thanks.  I note that sometimes the Sleepyhead Machine Settings shows variable PS and at other times fixed PS while in auto mode.  While my profile shows a Resmed Airsense 10, I have had such bad luck with it (don't know exactly why) I thought I would go back to my 2009 Repironics 750 bipap  to do a bit of titration.  My best numbers seem to come on that machine with äuto bipap and FIXED PS.  I don't know how you can have a FIXED PS on this machine unless the setting of minimum and maximum EPAP are the same, which of course is not possible on my machine.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  AC11 humidity control setting Shamus54 8 275 11-20-2024, 05:50 AM
Last Post: Shamus54
  [Equipment] Need advice for battery gennadyd 10 425 11-13-2024, 07:16 PM
Last Post: Polo 1955
  [CPAP] Need some advice before I quit SloppySteve 6 353 11-01-2024, 11:40 AM
Last Post: super7pilot
  Airsense 11 Auto - What is auto-off supposed to do? Schnoozer 3 261 10-31-2024, 07:33 PM
Last Post: Warnick13
  Just looking for some advice bambam1096 3 228 10-27-2024, 12:00 AM
Last Post: PollCat
Smile Looking for some advice NeilFromTheRiver 4 333 10-25-2024, 02:45 PM
Last Post: NeilFromTheRiver
  re setting password on OSCAR needed sleepingbetterinFl 3 217 10-14-2024, 11:26 PM
Last Post: sleepingbetterinFl


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.